What's going on with my spO2 dips? ASV.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Mon Dec 29, 2014 2:05 pm

bwexler wrote:It definitely will not integrate with Encore.
You might want to download SleepyHead software, it is free. Compared to Encore it is like going from an old black and white TV to a 50" Hi Def Flat Screen. SH does integrate the Oximeter data and the APAP data.
I don't think sleephead integrates with my phillips ASV machine.. Encore seems a little limited, wish I could use sleepyhead.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's going on with my spO2 dips? ASV.

Post by palerider » Mon Dec 29, 2014 3:16 pm

icipher wrote:
bwexler wrote:It definitely will not integrate with Encore.
You might want to download SleepyHead software, it is free. Compared to Encore it is like going from an old black and white TV to a 50" Hi Def Flat Screen. SH does integrate the Oximeter data and the APAP data.
I don't think sleephead integrates with my phillips ASV machine.. Encore seems a little limited, wish I could use sleepyhead.
are you using the 0.9.8-1 version?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Mon Dec 29, 2014 3:20 pm

palerider wrote:
icipher wrote:
bwexler wrote:It definitely will not integrate with Encore.
You might want to download SleepyHead software, it is free. Compared to Encore it is like going from an old black and white TV to a 50" Hi Def Flat Screen. SH does integrate the Oximeter data and the APAP data.
I don't think sleephead integrates with my phillips ASV machine.. Encore seems a little limited, wish I could use sleepyhead.
are you using the 0.9.8-1 version?
I believe I am using this:

Encore Pro 2.12 64 bit

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's going on with my spO2 dips? ASV.

Post by palerider » Mon Dec 29, 2014 6:11 pm

icipher wrote:
palerider wrote:
icipher wrote:don't think sleephead integrates with my phillips ASV machine.. Encore seems a little limited, wish I could use sleepyhead.
are you using the 0.9.8-1 version?
I believe I am using this:

Encore Pro 2.12 64 bit
we were discussing sleepyhead.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Mon Dec 29, 2014 8:15 pm

palerider wrote:
icipher wrote:
palerider wrote:
icipher wrote:don't think sleephead integrates with my phillips ASV machine.. Encore seems a little limited, wish I could use sleepyhead.
are you using the 0.9.8-1 version?
I believe I am using this:

Encore Pro 2.12 64 bit
we were discussing sleepyhead.
ah, my bad.

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 30, 2014 8:29 am

So I used my pulse OX all night last night. Here are my results, with yet ANOTHER instance where it shows my 02 going below 88% and my pulse spiking well above 100.

What's going on? Artifact again?

Image

Image

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's going on with my spO2 dips? ASV.

Post by palerider » Tue Dec 30, 2014 10:49 am

icipher wrote:So I used my pulse OX all night last night. Here are my results, with yet ANOTHER instance where it shows my 02 going below 88% and my pulse spiking well above 100.

What's going on? Artifact again?
import the data into sleepyhead and see what it looks like with your cpap data.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 30, 2014 11:03 am

palerider wrote:
icipher wrote:So I used my pulse OX all night last night. Here are my results, with yet ANOTHER instance where it shows my 02 going below 88% and my pulse spiking well above 100.

What's going on? Artifact again?
import the data into sleepyhead and see what it looks like with your cpap data.
My ASV is not compatible with sleepyhead. looking at my data in encore last night, my AHI was 1.4 all comprised of hypopneas.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's going on with my spO2 dips? ASV.

Post by palerider » Tue Dec 30, 2014 11:56 am

icipher wrote:
palerider wrote:
icipher wrote:So I used my pulse OX all night last night. Here are my results, with yet ANOTHER instance where it shows my 02 going below 88% and my pulse spiking well above 100.

What's going on? Artifact again?
import the data into sleepyhead and see what it looks like with your cpap data.
My ASV is not compatible with sleepyhead. looking at my data in encore last night, my AHI was 1.4 all comprised of hypopneas.
I ask again what I asked in the other thread, which you didn't answer, have you tried the current version of sleepyhead?

if you can't use SH, then at least compare any events between the two reports at the time of whatever that was, if you don't have any breathing anomalies, that should tell you something

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 30, 2014 12:01 pm

palerider wrote:
icipher wrote:
palerider wrote:
icipher wrote:So I used my pulse OX all night last night. Here are my results, with yet ANOTHER instance where it shows my 02 going below 88% and my pulse spiking well above 100.

What's going on? Artifact again?
import the data into sleepyhead and see what it looks like with your cpap data.
My ASV is not compatible with sleepyhead. looking at my data in encore last night, my AHI was 1.4 all comprised of hypopneas.
I ask again what I asked in the other thread, which you didn't answer, have you tried the current version of sleepyhead?

if you can't use SH, then at least compare any events between the two reports at the time of whatever that was, if you don't have any breathing anomalies, that should tell you something
I can certainly try downloading and installing sleepyhead, perhaps it has become compatible over the past year. Unfortunately, encore doesn't really do a good job of marking times for when events occur.

I see nothing in encore report that shows any long lasting cessation of breath, at least that I can see. no apneas or PB, just hypopneas which are quickly fixed when the machine kicks in.

Thanks.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: What's going on with my spO2 dips? ASV.

Post by HoseCrusher » Tue Dec 30, 2014 1:41 pm

That looks more like something like a heart flutter. The heart looses efficiency in pumping, then O2 drops. The heart resets and everything goes back to normal.

An obstruction in breathing is usually the other way around. O2 drops, then the heart rate increases to build the O2 levels back up.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 30, 2014 1:47 pm

HoseCrusher wrote:That looks more like something like a heart flutter. The heart looses efficiency in pumping, then O2 drops. The heart resets and everything goes back to normal.

An obstruction in breathing is usually the other way around. O2 drops, then the heart rate increases to build the O2 levels back up.

Hmm, that's not reassuring to hear. I sent this to my DME, but whenever I send them snapshots of possible concerns, they just say everything is normal.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: What's going on with my spO2 dips? ASV.

Post by HoseCrusher » Tue Dec 30, 2014 2:00 pm

If you want to discuss it you need to do so with someone that knows how to read oximetry data. A cardiologist would be a good choice. Your general practice doctor may also have some experience with this.

I wouldn't expect your DME to understand anything beyond the general basics.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 30, 2014 2:03 pm

HoseCrusher wrote:If you want to discuss it you need to do so with someone that knows how to read oximetry data. A cardiologist would be a good choice. Your general practice doctor may also have some experience with this.

I wouldn't expect your DME to understand anything beyond the general basics.
By DME, i mean my sleep board certified sleep doctor. Do you feel fairly confident that this is some sort of heart flutter or irregular rhythm, at least in the snapshot I provided?

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: What's going on with my spO2 dips? ASV.

Post by HoseCrusher » Tue Dec 30, 2014 2:44 pm

Let's put this into perspective...

People experience irregular heart rhythms all the time. Most people don't track their pulse rate so they go unnoticed until they occur frequently during the day.

I gave you my "best guess." While I have had a lot of experience reading oximetry data I am not an expert so just take it as a "best guess."

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...