What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 23, 2014 12:09 pm

Hello all, happy holidays.

Last night I was having a few wake-ups a few hours after falling asleep. I decided to throw my pulse oximeter on. I found something of surprise in my readouts from the oximeter. I see that it twice recorded my oxygen going down to 80%, which is something I've never seen before. I would like to get your honest thoughts on whether or not this looks legit or if it looks more like artifact. I notice my pulse spikes at each occurrence. I do not understand how my O2 could get this low while using the ASV and I've never seen it before.

Also, looking at my data from my ASV, my AHI was 2.0 all comprised of hypopnea. I've attached 2 screenshots from my pulse oximeter readout report.

Image

Image

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: What's going on with my spO2 dips? ASV.

Post by SGearhart » Tue Dec 23, 2014 1:55 pm

I've had a few like what you show. Mine have even dipped to 0. I've traced it down to a sudden movement. It could be just from changing sleeping position.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's going on with my spO2 dips? ASV.

Post by palerider » Tue Dec 23, 2014 4:31 pm

icipher wrote: I would like to get your honest thoughts on whether or not this looks legit or if it looks more like artifact.
it's impossible for your spo2 to drop 15 points in a few seconds and then bounce right back up to the spot it was before the glitch.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Demerit

Re: What's going on with my spO2 dips? ASV.

Post by Demerit » Tue Dec 23, 2014 4:56 pm

It's artifact. Nothing to worry about.

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Tue Dec 23, 2014 5:54 pm

Thanks for the replies. Basically the same thing I thought with the very sharp, steep decline and rise.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: What's going on with my spO2 dips? ASV.

Post by HoseCrusher » Sat Dec 27, 2014 12:45 am

I am not sure it is an artifact. The time is over a couple of minutes and the drop in O2 is accompanied by a spike in pulse rate. The pulse went from around 50 to around 80 over the same time frame. Also the % artifact is quite low and there are no gray sections indicating a loss of contact.

With all that said it is a rather abrupt change. Keep checking for a few more nights and see if it goes away. If it does then it may not mean anything.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: What's going on with my spO2 dips? ASV.

Post by Sludge » Sat Dec 27, 2014 4:23 am

HoseCrusher wrote:I am not sure it is an artifact. The time is over a couple of minutes and the drop in O2 is accompanied by a spike in pulse rate.
Which gives additional support to the artifact theory. During artifact, both values will tend to get screwy. If only one was screwy and other was perfectly stable, then you might be thinking hmm.

A more scientific analysis of oximetry values can be made in non-Walmart oximeters, where the pleth wave can be examined for quality, as well as a more precise measurement (i.e., oximetry waveform values 0.1% instead of 1.0%).
You Kids Have Fun!!

mecheng
Posts: 28
Joined: Mon Jun 30, 2014 1:40 pm
Location: United Kingdom

Re: What's going on with my spO2 dips? ASV.

Post by mecheng » Sat Dec 27, 2014 10:37 am

I quite agree with the partially dislodged artifact probability. It is quite common particularly with clip on finger type of oximeter when the curvature of the clip is not a good match to the particular digit chosen. Tape can sometimes help but may also act as a pivot point.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Sleepyhead and CMS50L pulse/ox and SPO2 Assistant
Philips Respironics Remstar Auto with A flex and humidifier. Pilaro Q mask / Swift FX mask
Wife as above. Respironics Wisp Nasal mask
Using Sleepyhead and Encore Basic software and CMS 50L pulse ox. Withing's pulse O2 and Sleep as Android for sound.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's going on with my spO2 dips? ASV.

Post by palerider » Sat Dec 27, 2014 2:05 pm

HoseCrusher wrote:I am not sure it is an artifact. The time is over a couple of minutes
the time is closer to 30 seconds, possibly less, (the tick marks across the bottom are five minute intervals, a 'couple minutes' would be very broad)

plus, you'll note that the device itself didn't flag an event (the bars across the top and bottom) in the data stream there is a flag for "data good", I'd be willing to bet that it's not set there.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: What's going on with my spO2 dips? ASV.

Post by archangle » Sat Dec 27, 2014 3:33 pm

I think measurement error/artifact is the most likely explanation. Zoom in closer on one of the spikes if you can, and it will be probably be easier to figure out.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: What's going on with my spO2 dips? ASV.

Post by HoseCrusher » Sat Dec 27, 2014 5:05 pm

palerider wrote:
HoseCrusher wrote:I am not sure it is an artifact. The time is over a couple of minutes
the time is closer to 30 seconds, possibly less, (the tick marks across the bottom are five minute intervals, a 'couple minutes' would be very broad)

plus, you'll note that the device itself didn't flag an event (the bars across the top and bottom) in the data stream there is a flag for "data good", I'd be willing to bet that it's not set there.


Let's put some context around this...

I put on my pulse oximeter and relaxed until my pulse rate was in the 50's and my O2 was in the upper 90's. I then exhaled and held my breath.

My O2 went from 97 - 88 in 27 seconds during which my pulse went from 53 - 77. When I couldn't hold my breath any longer I took some deep breaths and my O2 went from 88 - 98 in 18 seconds and my pulse went from 77 - 50.

That's a 9 point drop in O2 and a 24 BMP increase in pulse in 27 seconds and the rebound involved a 10 point increase in O2 and a 27 BPM drop in pulse in 18 seconds. The whole event took place over 53 seconds.

After a few minutes relaxing I tried it again. This time I recorded a 8 point drop in O2 along with a 24 BPM increase in pulse in 38 seconds and the rebound had a 9 point increase in O2 along with a 28 point drop in BPM in 15 seconds. This event ran 62 seconds.

The third time was less of a drop in O2. I saw a 6 point drop in O2 and a 18 BPM increase in pulse in 13 seconds and the rebound had a 7 point increase in O2 and a 24 BPM reduction in pulse in 18 seconds. The total event time was 48 seconds.

When you run a test like this with your oximeter what do you see?

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
bwexler
Posts: 1581
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: What's going on with my spO2 dips? ASV.

Post by bwexler » Sat Dec 27, 2014 8:22 pm

What brand and model of oximeter are you using? Do you have Sleepy head software? It will provide a nicer more adjustable graph integrated with your APAP. If you have one of the CBS oximeters it will also show profusion index which helps to identify artifacts.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: What's going on with my spO2 dips? ASV.

Post by icipher » Sun Dec 28, 2014 8:51 pm

bwexler wrote:What brand and model of oximeter are you using? Do you have Sleepy head software? It will provide a nicer more adjustable graph integrated with your APAP. If you have one of the CBS oximeters it will also show profusion index which helps to identify artifacts.
Hello,

I have the Pulse Oximeter CMS-50D Plus from Cooper medical, if that helps. I do not think it will integrate into Encore.

User avatar
bwexler
Posts: 1581
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: What's going on with my spO2 dips? ASV.

Post by bwexler » Sun Dec 28, 2014 9:44 pm

It definitely will not integrate with Encore.
You might want to download SleepyHead software, it is free. Compared to Encore it is like going from an old black and white TV to a 50" Hi Def Flat Screen. SH does integrate the Oximeter data and the APAP data.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
Too tall
Posts: 297
Joined: Sat Dec 14, 2013 10:58 pm

Re: What's going on with my spO2 dips? ASV.

Post by Too tall » Sun Dec 28, 2014 11:27 pm

If your sleepyhead software showed an OSA event at that time, then it's not an artifact. It's hard to tell with that large scale. It looks like a sudden drop but it could be over a longer period.