small child profile light... but seriously folks, 30 something intentional and 20something excess... no good, but not horrible, no?Morbius wrote:So I have a question.
How the heck did Total Leak get to ~56 LPM?
No mask does that, even at 14.0 cmH2O:
http://www.healthcare.philips.com/pwc_h ... 111027.pdf
Help! Hubby's settings failed him, stop breathing short time
Re: Help! Hubby's settings failed him, stop breathing short time
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Help! Hubby's settings failed him, stop breathing short time
From ~20 the day before? With 1.0 cmH2O additional pressure?


Re: Help! Hubby's settings failed him, stop breathing short time
He wears a chin strap but I note he has his mouth open just a tiny bit but the machine has never shown big leak so I thought it was okay,
Re: Help! Hubby's settings failed him, stop breathing short time
most of the night looks like the worst of the night before, and the periods of very low flow variation coincide with the highest of leak...Morbius wrote:From ~20 the day before? With 1.0 cmH2O additional pressure?
machine having problems registering flow variation when the leak's that bad, perhaps?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Help! Hubby's settings failed him, stop breathing short time
I don't understand the question. Would you restate what you have told me? I really want to understand how to help him. Thanks so much for looking at this.
Re: Help! Hubby's settings failed him, stop breathing short time
Just to avoid the confusion of multiple threads -- this discussion has moved to a new thread: viewtopic/t106001/Morbius-would-you-take-a-look.html
_________________
| Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle |
| Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura |
Re: Help! Hubby's settings failed him, stop breathing short time
"IMHO", and overall, the Patient Leak is OK. The Total Leak is what appears to be screwy:palerider wrote:most of the night looks like the worst of the night before, and the periods of very low flow variation coincide with the highest of leak...
machine having problems registering flow variation when the leak's that bad, perhaps?


AAMOF, the worst leaks (which still aren't horrible) do not exhibit the Low Flow Behavior.
If you want, upload the whole file to Dropbox, and breath-by-breath research can be used to try to figure out what the issue is.
Re: Help! Hubby's settings failed him, stop breathing short time
Actually, the pressure increase was 1.5 cmH2O, but that should have increased Intentional Leak by only ~2 LPM.Morbius wrote:With 1.0 cmH2O additional pressure?
Re: Help! Hubby's settings failed him, stop breathing short time
it might be important to note that the total leak is the only data reported from that machine, the 'patient leak' is a sleepyhead calculated value, and it has no sanity checking in this version, if you have an added unintentional leak it won't show up in the 'leak' line. SH takes the total leak and does math stuff to calculate a baseline. but if you never get a decent seal, it won't be able to compute a valid baseline. I've had discussions with jedimark about adding in some sanity checks, but those haven't been included yet as of the 098-1 version. therefore it's best not to put a lot of stock in the 'leak' line on a respironics machine trace.Morbius wrote:"IMHO", and overall, the Patient Leak is OK. The Total Leak is what appears to be screwy:palerider wrote:most of the night looks like the worst of the night before, and the periods of very low flow variation coincide with the highest of leak...
machine having problems registering flow variation when the leak's that bad, perhaps?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Help! Hubby's settings failed him, stop breathing short time
We can figure that out manually by finding out is what mask was being used and calculate it using the previously posted reference chart. Also noting any mask changes. That 56 LPM looks stable, and if you subtract the ~35 LPM of typical Intentional Leak, the remaining 21 LPM is not particular toxic (ResMed gives you to 24 LPM).palerider wrote:...the 'patient leak' is a sleepyhead calculated value...
Consequently, one still must believe that the low flow rates are accurate....
Re: Help! Hubby's settings failed him, stop breathing short time
...but obviously need to determine if they require emergency intervention or if they are benign (and clearly you don't want to guess wrong).
Anyway, another explanation could be a dramatic decrease in Dead Space. If the Dead Space zones are flushed, then the Tidal Volume will also decrease as breathing becomes highly efficient...
Anyway, another explanation could be a dramatic decrease in Dead Space. If the Dead Space zones are flushed, then the Tidal Volume will also decrease as breathing becomes highly efficient...
Re: Help! Hubby's settings failed him, stop breathing short time
But just in case somebody is thinking "Wow, what a great idea, I'm going to invent a system that shoots oxygen in at 60 LPM and do wonderful things!", don't bother, you're too late:
http://www.rtmagazine.com/2013/09/high- ... ifference/
http://www.rtmagazine.com/2013/09/high- ... ifference/
Re: Help! Hubby's settings failed him, stop breathing short time
that's what I'd noticed, but as to what it means, your guess is far better than mine.Morbius wrote:...and although data is limited, Low Flow Behavior is predictable. When there is a stable Unintentional Leak of ~21 LPM, LFB appears:
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.