Dog Slobber,
I want to thank you for taking the time to draw that graph that shows so clearly that sampling at a lower resolution (1/second) still shows a good approximation of the correct peaks and troughs of the hi res mask pressure graph. Clearly SleepHQ is either doing something weird with the data to create their mask pressure graph OR there's a bug in their software.
In either case, Wanderlustralia refuses to admit there's a problem with a graph that is quite obviously wrong in the SleepHQ presentation of the data. And that does call into question what other stuff might be buggy.
When I look at my Resmed AirCurve 10 VAuto data, all of the other graphs, on a cursory look in my data seem to be OK, although they only draw the IPAP pressure graph for the pressure graph instead of both the EPAP and IPAP curves.
But when I look at my PR DreamStation BiPAP Auto data, however, I have a serious problem with the way SleepHQ presents the pressure data: SleepHQ draws
one pressure graph that is neither the EPAP graph nor the IPAP graph that is shown in both Encore Pro (the PR software) and SleepyHead. Oscar puts in a third graph that looks remarkably like the Pressure graph in SleepHQ. Personally, I would call the way SleepHQ draws the pressure curve for PR BiPAPs a
bug that ought to be fixed.
Here are the IPAP and EPAP pressure curves for Oct.29, a night that I used the DreamStation, as shown in SleepyHead. (The graph in Encore Pro is similar.)
Here are the IPAP and EPAP pressure curves, along with a mystery "pressure" curve that is not present in Encore, as shown in Oscar:

I asked about that mystery curve
in this thread. Nobody gave a definitive answer, but the consensus was that it must be some kind of a moving average mask pressure value. (The Oscar documentation at ApneaBoard says this is just lo res sampling of the mask pressure data, but they haven't really really explained where that data comes from; when I look at my husband's PR System One APAP data, Oscar shows the pressure set curve and an EPAP curve that appears to be derived from the Pressure set curve and some kind of mystery data in the same way SleepyHead does.)
And here is the so-called pressure curve as drawn by SleepHQ:

Given that this is
BiPAP data and the PS is allowed to from 3 to 5 during the night, only drawing the mystery "pressure" curve is badly distorting the pressure data. Looking at this curve, you'd think that my pressure never reached 7cm, which is my
minimum IPAP pressure and so the pressure went to (at least) 7cm on each and every inhalation that I took over the course of the night.
Looking at the Statistics part of the SleepHQ data for this day is also misleading:

It correctly identifies the median, 95%, and maximum EPAP pressures. But missing are the median, 95%, and maximum IPAP pressures, which were 7cm, 9cm, and 9cm respectively on this night. The median, 95%, and maximum pressure levels shown in SleepHQ have nothing to do with my machine's therapeutic settings.
I could see the case where a newbie or average PR BiPAP Auto user might look at the pressure information in SleepHQ and start to wonder whether there is something wrong with their BiPAP since the pressure is never reaching the typical IPAP levels that would be expected given the clinical settings.
Finally, it should be clear that the PR pressure curves are most definitely
not hi-res. They represent the pressure settings for IPAP and EPAP, not the actual pressure at the mask. And PR machines take their own sweet time in adjusting the pressure settings when they decide to increase either the IPAP or the EPAP or both.