Did I STOP Breathing "AGAIN"

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by JDS74 » Sun Aug 17, 2014 5:58 pm

Although I prefer PhotoBucket, getting to detailed graphs on Flickr is quite easy, at least in Windowas.
1. Click on the image.
2. Right click on the Flickr image.
3. Select Full size (way on the right)
There you are.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by igdoc » Tue Aug 19, 2014 3:12 am

Could be a blockage of the flow sensor. However my first thought on this is that your tubing or mask blocked off. It happened suddenly and corrected suddenly. Is it possible that the tubing kinked somewhere? The machine reactions and Sleepyhead do look to be working correctly. I presume you would have ended up breathing through your mouth whilst the blockage occurred. Does this seem to fit?
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Sludge » Tue Aug 19, 2014 5:19 am

Cowboy Casey wrote:...well I feel the same so maybe not dead yet...
When you are, can I have your boat?

TIA.
You Kids Have Fun!!

User avatar
Cowboy Casey
Posts: 154
Joined: Wed Jun 04, 2014 10:08 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Cowboy Casey » Tue Aug 19, 2014 11:58 am

Ian, I thought about that and I am just not positive that did not happen... it could have, but what if it did not?? crazzy to think that but maybe...

igdoc wrote:Could be a blockage of the flow sensor. However my first thought on this is that your tubing or mask blocked off. It happened suddenly and corrected suddenly. Is it possible that the tubing kinked somewhere? The machine reactions and Sleepyhead do look to be working correctly. I presume you would have ended up breathing through your mouth whilst the blockage occurred. Does this seem to fit?
Ian
PR System One REMstar 60 Series BiPAP Auto SV Advanced
https://www.cpap.com/productpage/pr-60-series-bipap-autosv-advanced.html
Pilairo Q Nasal Pillow CPAP Mask with Headgear
https://www.cpap.com/productpage/fp-pilairo-q-nasal-pillow-cpap-mask.html

User avatar
Cowboy Casey
Posts: 154
Joined: Wed Jun 04, 2014 10:08 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Cowboy Casey » Tue Aug 19, 2014 12:01 pm

TIA, LOL, its for sale You get about 20K in free stereo and equipment..
Sludge wrote:
Cowboy Casey wrote:...well I feel the same so maybe not dead yet...
When you are, can I have your boat?

TIA.
PR System One REMstar 60 Series BiPAP Auto SV Advanced
https://www.cpap.com/productpage/pr-60-series-bipap-autosv-advanced.html
Pilairo Q Nasal Pillow CPAP Mask with Headgear
https://www.cpap.com/productpage/fp-pilairo-q-nasal-pillow-cpap-mask.html

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by igdoc » Wed Aug 20, 2014 2:51 am

Cowboy Casey wrote:Ian, I thought about that and I am just not positive that did not happen... it could have, but what if it did not?? crazzy to think that but maybe...
If it did not, sorry to say but you would not be here. The tidal volumes reported during those two hours are less than the physiological dead space of around 175 mls. So you would have been getting zero oxygen to your lungs from your machine and mask during that time. Is the extra tubing or some part of your Pilairo mask susceptible to getting kinked or squashed if you lie with your face down on your pillow for example?
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG

User avatar
Cowboy Casey
Posts: 154
Joined: Wed Jun 04, 2014 10:08 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Cowboy Casey » Wed Aug 20, 2014 7:27 am

I have it hooked above my head on the wall so it is suspended down to me, there is just enough slack so the mask does not pull off my face... I tried to kink/smash the hose last night and could not recreat the same as that night...

I wonder if the machine just got lost for awhile?? Anything can happen...

It nice to now that it did react and do what it was suppose to do and I am still here, thats a great thing... I sent the data to my nurse and doctor and they are looking at it... we will see what they come up with...

Thanks to every one for looking over this, I will let everyone know what and if they come up with anything...
PR System One REMstar 60 Series BiPAP Auto SV Advanced
https://www.cpap.com/productpage/pr-60-series-bipap-autosv-advanced.html
Pilairo Q Nasal Pillow CPAP Mask with Headgear
https://www.cpap.com/productpage/fp-pilairo-q-nasal-pillow-cpap-mask.html

Paralel
Posts: 216
Joined: Thu Aug 14, 2014 7:49 pm

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Paralel » Wed Aug 20, 2014 3:57 pm

Agree with the above, if that was real, you'd be dead. Either machine malfunction or data corruption, take your pick.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by archangle » Wed Aug 20, 2014 5:54 pm

igdoc wrote:Could be a blockage of the flow sensor. However my first thought on this is that your tubing or mask blocked off. It happened suddenly and corrected suddenly. Is it possible that the tubing kinked somewhere? The machine reactions and Sleepyhead do look to be working correctly. I presume you would have ended up breathing through your mouth whilst the blockage occurred. Does this seem to fit?
Ian
Look at the leak rate. It stays pretty much the same throughout the event. That strongly suggests:

The flow rate sensor is not plugged and is functioning normally.
The average pressure in the mask was the same before and during the event.
The hose is not plugged or kinked, unless it's plugged or kinked between the exhaust port and your lungs.
There's not some sort of leak event causing the problem.
The exhaust vent isn't blocked.

The only possible mechanical blockage explanation is that the pillows are somehow kinked between the elbow and your nose, or your nose is mostly clogged or blocked. I fooled around with my Swift FX, and couldn't find a reasonable way to mostly block off the airflow at the pillows. Casey, play around with your Pilario Q and see if you can get it to collapse or fold somehow such that airflow is blocked right at the pillows.


BTW, this is a good example of why it's so stupid to throw away the total leak data the way ResMed does. Not only does it verify the flow sensor is working, it verifies the hose isn't kinked, the exhale vent isn't plugged or blocked, etc.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Sludge » Thu Aug 21, 2014 3:31 am

I like the "occluded pillows/nasal passages" argument.

A closer examination shows this was not necessarily a sudden event. Inspiratory flow is reduced and pressures have been climbing prior to the onset of machine breaths:

Image

That last disrupted inspiration could be the completion of the occlusion and conversion to mouth breathing (which, of course, the machine would be unable to recognize).
You Kids Have Fun!!

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Sludge » Thu Aug 21, 2014 5:21 am

Of course, the question that begs to be asked is:

"If CC was totally mouth-breathing, that is essentially untreated OSA. Why is there no evidence of that?"

Well, I would submit that CC's OSA is not that severe and/or easily treatable, as witnessed by only sporadic events on low-level therapy:

Image
Image
Image

So if the above scenario was registered during lateral sleeping or SWS, there may not have been any events.
You Kids Have Fun!!

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Sludge » Thu Aug 21, 2014 6:47 am

BTW can you post an image of the return to normal flow in a ~5 minute window? Looks like the situation corrected during a SWJ.

Also the center area.

TIA.
You Kids Have Fun!!

User avatar
Cowboy Casey
Posts: 154
Joined: Wed Jun 04, 2014 10:08 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by Cowboy Casey » Thu Aug 21, 2014 8:28 am

Sludge wrote:BTW can you post an image of the return to normal flow in a ~5 minute window? Looks like the situation corrected during a SWJ.

Also the center area.

TIA.
I will tonight, Sorry I have not been on as much... working 14 hour days and when I get home I eat and go to sleep...

Casey
PR System One REMstar 60 Series BiPAP Auto SV Advanced
https://www.cpap.com/productpage/pr-60-series-bipap-autosv-advanced.html
Pilairo Q Nasal Pillow CPAP Mask with Headgear
https://www.cpap.com/productpage/fp-pilairo-q-nasal-pillow-cpap-mask.html

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by archangle » Thu Aug 21, 2014 1:13 pm

Sludge wrote:I like the "occluded pillows/nasal passages" argument.
Sludge, do you think the reported tidal volume/minute vent numbers are below the possible levels? i.e. would he be dead if all the data was really correct? Can we assume that there's either a data glitch or he's somehow breathing "around" the mask? His tidal volume appears to be below 170 for quite some time and minute vent is around 1 for a while?

Casey, can you zoom in on about 2:05 to 2:25 for this graph below?

Image

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: Did I STOP Breathing almost 2 hours with ASV????

Post by igdoc » Thu Aug 21, 2014 2:20 pm

Archangle, excellent point about the constant leak rates putting any mechanical blockage beyond the vent ports. Not familiar with the Pilairo but I wonder if the pillows could displace and impact into the skin. It is pretty extreme and sudden for nasal blockage. Thank goodness for mouth breathing!
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG