Can't sleep on right side
Can't sleep on right side
For some reason if I turn on my right side to sleep, the pressure leaps dramatically and the mask starts hooting and honking until I turn back on left side (which is my usual side)
This lack of being able to change sides is a MAJOR discomfort.
Ordinarily I turn on right side to get ...back... to sleep if I'm restless or fed up with left side.
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Upper moderate apnea.
Presently on titration (test) machine Resmed S9 for the evaluation week.
Have swapped masks trying for best fit. Start pressure is 8 cmh20.)
This lack of being able to change sides is a MAJOR discomfort.
Ordinarily I turn on right side to get ...back... to sleep if I'm restless or fed up with left side.
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Upper moderate apnea.
Presently on titration (test) machine Resmed S9 for the evaluation week.
Have swapped masks trying for best fit. Start pressure is 8 cmh20.)
_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
Additional Comments: CPAP is A-Flex and not C-Flex. Not using humidifier. Pressure 11 cm |
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
Re: Can't sleep on right side
I answered in your other thread, but the fitting tips in my blog will give you some ideas on how to adjust a mask for side sleeping (hint: LOOSEN the straps slightly on the right side, tighten slightly on the left side, and use the pillow to "snuggle" your face into to help with little leaks that spring up). Make sure your hose isn't kinked or otherwise in a bad position when you lie on that side. A hose management system helps.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Can't sleep on right side
I...think.... I'm trying that.
But the pressure is rising SO high (compared to left side) that it honks and blows and tries to lift off.
Hmmm.... maybe I haven't tried loosening the right side.
I'll do that.
I've done up the sides till I think I'm gonna squoosh my head. I'll try loosening.
Maybe super glue.
But the pressure is rising SO high (compared to left side) that it honks and blows and tries to lift off.
Hmmm.... maybe I haven't tried loosening the right side.
I'll do that.
I've done up the sides till I think I'm gonna squoosh my head. I'll try loosening.
Maybe super glue.
_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
Additional Comments: CPAP is A-Flex and not C-Flex. Not using humidifier. Pressure 11 cm |
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Can't sleep on right side
Maybe try Lansinoh on the mask where it touches your face.
A tiny bit is fairly sticky--like gasket sealer.
They sell it near the baby/nursing stuff.
(yes, it is nipple goo)
A tiny bit is fairly sticky--like gasket sealer.
They sell it near the baby/nursing stuff.
(yes, it is nipple goo)
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Can't sleep on right side
Do you use a hose hanger so that when you turn over there is no tugging on the mask?
Is this happening towards the end of ramp time (if you are using the ramp feature)?
Are you letting the mask be pushed by your pillow?
Are you using the mask fit feature so that your mask is fitted for leaks at a higher pressure?
Even though your current mask and machine are temporary, could you fill out the equipment profile? That would give folks a little more information.
Do you have a copy of your sleep study? Can you get one?
It contains valuable information for you.
Hang in there, it will get better.
Is this happening towards the end of ramp time (if you are using the ramp feature)?
Are you letting the mask be pushed by your pillow?
Are you using the mask fit feature so that your mask is fitted for leaks at a higher pressure?
Even though your current mask and machine are temporary, could you fill out the equipment profile? That would give folks a little more information.
Do you have a copy of your sleep study? Can you get one?
It contains valuable information for you.
Hang in there, it will get better.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Can't sleep on right side
NO! And that's why I'm not ordering my machine til I get some info.Do you have a copy of your sleep study?
It contains valuable information for you.
I better be able to get it.
_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
Additional Comments: CPAP is A-Flex and not C-Flex. Not using humidifier. Pressure 11 cm |
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
Re: Can't sleep on right side
You can ask for your sleep study now. Your Dr. should give it to you. (Actually they are required to if you ask but some aren't cooperative)
Sounds like your machine might be fully data capable. If you want to see what is going on you can run software to see your nightly sleep results.
Just let everyone know and you'll have plenty of people that will try to help you install the software.
Sounds like your machine might be fully data capable. If you want to see what is going on you can run software to see your nightly sleep results.
Just let everyone know and you'll have plenty of people that will try to help you install the software.
_________________
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Straight CPAP @ 12 cmH2O, EPR 2, Sleepyhead Software v0.9.8-1 |
Re: Can't sleep on right side
I doubt I could get the study yet.
There is still one more night to complete the 7 day study.
Just getting ideas of what to ask for
There is still one more night to complete the 7 day study.
Just getting ideas of what to ask for
_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
Additional Comments: CPAP is A-Flex and not C-Flex. Not using humidifier. Pressure 11 cm |
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
Re: Can't sleep on right side
I guess I misread what you wrote. I thought you were saying you were doing your titration study. I see you said you were using a titration machine, a Resmed S9.
_________________
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Straight CPAP @ 12 cmH2O, EPR 2, Sleepyhead Software v0.9.8-1 |
Re: Can't sleep on right side
I keep suggesting that! nobody's taken me up on it!Napnea wrote: Maybe super glue.
unless they accidentially superglued their fingers to their face and can't type anymore... *ponder*
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Can't sleep on right side
Not funny! I superglued my fingers last week doing some dumb thing or other, and after applying every conceivable thing in the house, I ended up using sandpaper. It worked! So maybe I no longer have prints and can rob a bank!
Re: Can't sleep on right side
Go to the medical secretaries office at your Kaiser facility and request a copy of your sleep study and the narrative report. Those should be done since you were referred on for the titration study. Sometimes they will give it to you on a disk. You may have to pay a small fee (a few dollars).Napnea wrote:I doubt I could get the study yet.
There is still one more night to complete the 7 day study.
Just getting ideas of what to ask for
You have a right to it by law, but navigating Kaiser's system is sometimes challenging. You probably haven't even met the sleep doctor (and probably never will), but be nice to the techs in the sleep lab--they are your new best friends. They're really the ones running the show in the sleep department, the doctor is only there to sign off on what they do to comply with state and federal law. If you're nice to them, and you need something, they will help you. Your primary care physician won't be much help because he/she does not work in that department.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Can't sleep on right side
You didn't misread. It was my thinking that the titration part IS the sleep study.I guess I misread what you wrote. I thought you were saying you were doing your titration study. I see you said you were using a titration machine, a Resmed S9.
It looks like I am premature on my questions in this forum.
_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
Additional Comments: CPAP is A-Flex and not C-Flex. Not using humidifier. Pressure 11 cm |
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
Re: Can't sleep on right side
You are now ahead in this game. Good luck, EarlNapnea wrote:.... It looks like I am premature on my questions in this forum.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PAPCap, 3M Microfoam Surgical Tape, PoliGrip Strip, APAP 12.0 - 14.0 cm |
Re: Can't sleep on right side
Boy, that is the truth. My primary guy deferred to the sleep lab for a previous question about apnea (???)navigating Kaiser's system is sometimes challenging. You probably haven't even met the sleep doctor (and probably never will), but be nice to the techs in the sleep lab--they are your new best friends. They're really the ones running the show in the sleep department, the doctor is only there to sign off on what they do to comply with state and federal law. If you're nice to them, and you need something, they will help you. Your primary care physician won't be much help because he/she does not work in that department.
And that info is good.
However my primary doc is now "out" til 8/25 so getting the sleep test (if I'm asking for the right thing) won't be coming from him til then.
Oh well.
_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
Additional Comments: CPAP is A-Flex and not C-Flex. Not using humidifier. Pressure 11 cm |
(Male, 74 yrs old, 6', 250 lbs down from 354 a few years back
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.
but plateaued against further loss anything short of starving.
Starting pressure 8.0 cm ramps up to 11 during night.
Not using humidifier.