ASV Help understanding the meaning of RAMP
ASV Help understanding the meaning of RAMP
Had my first night on ASV last night in in general it was awful, but that is another story. My main question after talking to my DME is whether I truly understand what RAMP really is. I noticed that all the menus on my machine did not list RAMP and the clinical menus (tricky little devils aren't we?) did not list it as well. That confused me. My DME told me that is because my min EPAP is already set as low as the machine can go i.e. 4cm H20 therefore it cannot go below this for a RAMP.
So - does that mean that Ramp only affects the min EPAP and none of the other pressures? The user manual for the machine just says that the "pressure" is reduced for a period of time but does not designate what type of pressure they are taking about.
My question then is, what pressure does the RAMP feature effect on my Res ASV machine and is she right that it only effects the min EPAP.
Learning in leaps and bounds, but tired as hell....
So - does that mean that Ramp only affects the min EPAP and none of the other pressures? The user manual for the machine just says that the "pressure" is reduced for a period of time but does not designate what type of pressure they are taking about.
My question then is, what pressure does the RAMP feature effect on my Res ASV machine and is she right that it only effects the min EPAP.
Learning in leaps and bounds, but tired as hell....
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software
Re: ASV Help understanding the meaning of RAMP
It is a comfort feature, and will start you at a lower EPAP and IPAP pressure, then slowly bring you up to the minimum EPAP and IPAP. There is a minimum pressure support, which is the minimum difference between the EPAP and the IPAP. So, for example, if your minimum EPAP is 7 and your minimum pressure support is 4, your starting IPAP would be 11. If you used ramp and you started out at 4 EPAP, then your IPAP would start at 8 (4 + 4).
But since you are already starting at a minimum EPAP of 4, you are correct in that the ramp does nothing for you.
And yes, ASV's are a real PIA at first. It DOES get better, but may take some adjustments as well as adjusting to it. By the way, if you scream into a full face mask, the ASV goes nuts!
But since you are already starting at a minimum EPAP of 4, you are correct in that the ramp does nothing for you.
And yes, ASV's are a real PIA at first. It DOES get better, but may take some adjustments as well as adjusting to it. By the way, if you scream into a full face mask, the ASV goes nuts!
Re: ASV Help understanding the meaning of RAMP
Thanks for that explanation. Hmmmmmm what do you mean by "go nuts"?!?!?!!?
It is already ruining my life; how on earth could it get worse? (wink).
It is already ruining my life; how on earth could it get worse? (wink).
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software
Re: ASV Help understanding the meaning of RAMP
If the min EPAP = 4.0, then the RAMP is disabled on the PR machines because the EPAP cannot be lowered below 4cm.photonic wrote:Had my first night on ASV last night in in general it was awful, but that is another story. My main question after talking to my DME is whether I truly understand what RAMP really is. I noticed that all the menus on my machine did not list RAMP and the clinical menus (tricky little devils aren't we?) did not list it as well. That confused me. My DME told me that is because my min EPAP is already set as low as the machine can go i.e. 4cm H20 therefore it cannot go below this for a RAMP.
If your minimum EPAP is greater than 4cm, then the RAMP is enabled. The starting ramp EPAP is set in the clinical menu. The starting ramp IPAP = starting ramp EPAP + min PS. To make this clearer, let's suppose your settings were:So - does that mean that Ramp only affects the min EPAP and none of the other pressures? The user manual for the machine just says that the "pressure" is reduced for a period of time but does not designate what type of pressure they are taking about.
- min EPAP = 7
max IPAP = 20
min PS = 3
max PS = 8
ramp start EPAP = 4 (I'm not sure what the exact name of the starting ramp EPAP setting is, but it's close to this.)
If your min EPAP > 4, then RAMP would affect both the starting EPAP and starting IPAP pressure. But because min EPAP = 4, there's no way of using a RAMP.My question then is, what pressure does the RAMP feature effect on my Res ASV machine and is she right that it only effects the min EPAP.
Curiosity:
You've said your min EPAP = 4cm.
What are your min PS, max PS, and max IPAP settings?
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: ASV Help understanding the meaning of RAMP
Robysue,
Min EPAP 4
Max EPAP 15
Max Pressure 25
Min PS 0
Max PS 20
Any education on understanding these values would be much appreciated. Thanks for taking an interest!
Min EPAP 4
Max EPAP 15
Max Pressure 25
Min PS 0
Max PS 20
Any education on understanding these values would be much appreciated. Thanks for taking an interest!
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software
Re: ASV Help understanding the meaning of RAMP
so they gave you a default out of the box setup, one would have thought you'd get something more tailored to you after all the doctors and techs and sleep studies and all that... as opposed to the 'bought it on craigslist and set it to default' option that I hadphotonic wrote:Robysue,
Min EPAP 4
Max EPAP 15
Max Pressure 25
Min PS 0
Max PS 20
Any education on understanding these values would be much appreciated. Thanks for taking an interest!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ASV Help understanding the meaning of RAMP
You start the night off with:photonic wrote:Robysue,
Min EPAP 4
Max EPAP 15
Max Pressure 25
Min PS 0
Max PS 20
Any education on understanding these values would be much appreciated. Thanks for taking an interest!
- EPAP = Min EPAP = 4cm
- IPAP = Min EPAP + Min PS = 4 + 0 = 4cm
- Min EPAP <= Current EPAP <= Max EPAP all night, so 4 <= Current EPAP <= 15 all night long.
- Current EPAP + min PS <= Current IPAP <= smaller of (Current EPAP + Max PS OR 25)
- Current EPAP + 0 <= Current IPAP <= smaller of (Current EPAP + 20, 25)
- Current EPAP <= IPAP <= 25
- 4 <= EPAP <= 15 all night long
- 4 <= IPAP <= 25 all night long
- 0 <= IPAP - EPAP <= 20 all night long.
- EPAP = 4 = Min EPAP, IPAP = 4, 5, 6, 7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24 = Current EPAP + Max PS
- EPAP = 5, IPAP = 5, 6, 7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 6, IPAP = 6, 7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 7, IPAP = 7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 8, IPAP = 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 9, IPAP = 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 10, IPAP = 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 11, IPAP = 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 12, IPAP = 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 13, IPAP = 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 14, IPAP = 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
- EPAP = 15 = Max EPAP, IPAP = 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25 = max Pressure
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: ASV Help understanding the meaning of RAMP
Just a gentle reminder to get and keep a copy of your sleep studies.
In there are the min EPAP and IPAP values, etc. that you need to compare with the actual machine settings.
Hang in there. It will all be good.
In there are the min EPAP and IPAP values, etc. that you need to compare with the actual machine settings.
Hang in there. It will all be good.
_________________
| Mask: Oracle HC452 Oral CPAP Mask |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: ASV Help understanding the meaning of RAMP
It's good to see that I am not the only one to dumb to understand the details of ASV.
Please keep asking dumb questions, so I can keep learning.
I have learned more about ASV in the past week from relevant posts then in the 9 months I have had the machine.
Please keep asking dumb questions, so I can keep learning.
I have learned more about ASV in the past week from relevant posts then in the 9 months I have had the machine.
_________________
| Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear |
| Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i |
- Uncle Flapp
- Posts: 247
- Joined: Thu Nov 14, 2013 2:24 pm
- Location: Arizona
Re: ASV Help understanding the meaning of RAMP
My min EPAP starts only at 6 so I don't use the ramp to fall asleep but sometimes snoring will trigger the machine to blast off like a rocket to 25/15 and the higher pressures often wake me up. When I do, the quickest way to reset is hit the ramp button, change positions, and get back to blissful dreamland.
I used to have claustrophobia, fight the rhythm of the machine, and be lucky to make the 4-hour compliance. It just takes time and persistence. Stick with it and you'll get there.
- Flappy
I used to have claustrophobia, fight the rhythm of the machine, and be lucky to make the 4-hour compliance. It just takes time and persistence. Stick with it and you'll get there.
- Flappy
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84 |