S9 SD data card
S9 SD data card
The power went out and when it came back on it gave me a error message invalid SD card . It asked to delete all data and if you check no it still says invalid SD card . I assume that I need to reformat the card. Can someone give me advice. I have res can software
Thanks
Thanks
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: S9 SD data card
The power outage occurred at precisely the time the machine was writing to the card. When that happened, some crucial piece of data was damaged. Then, when the machine came back on, the firmware prigram noticed the damage and wants to fix it.
Say yes and you'll be good to go. There will be some data loss but mostly things will be OK.
Pugsy and others who are more familiar with Resmed data will be able to explain exactly what happens when a ResNed card is reformatted.
Say yes and you'll be good to go. There will be some data loss but mostly things will be OK.
Pugsy and others who are more familiar with Resmed data will be able to explain exactly what happens when a ResNed card is reformatted.
_________________
| Mask: Oracle HC452 Oral CPAP Mask |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: S9 SD data card
let it do what it wants to do, it'll all be fine,halb wrote:The power went out and when it came back on it gave me a error message invalid SD card . It asked to delete all data and if you check no it still says invalid SD card . I assume that I need to reformat the card. Can someone give me advice. I have res can software
Thanks
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: S9 SD data card
The problem is rescan also says there is an error and can not download. (There is after i said yes to delete data)
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: S9 SD data card
I left the card out one damn night and lost that data set. Annoying there is no memory on board to store it.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: S9 SD data card
all of the good data is written to the card directly, the compliance data and stats are stored in the machine.halb wrote:The problem is rescan also says there is an error and can not download. (There is after i said yes to delete data)
if you have to reformat, you'll end up losing whatever was on the card at the time, but not any compliance data.
but, if you don't let the machine reformat, then it won't use the card, and you won't get any new data on it...
best thing is to download the data frequently into sleepyhead, which will archive it for you.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: S9 SD data card
Good advice, I had heard the machine stored one night's detail - and I had heard wrong.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: S9 SD data card
check page 25 of the escape/escape auto/elite/autoset clinicians manuals for the data recorded, and where stored, and for how long.englandsf wrote:Good advice, I had heard the machine stored one night's detail - and I had heard wrong.
I've pasted it into other msgs on here before, but it doesn't paste well, and it's a pain to format
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: S9 SD data card
Unplug the machine, and count to ten before plugging it back in.
You may still have to clear the card, but stuff happens--this could be a one-off.
You may still have to clear the card, but stuff happens--this could be a one-off.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |