RemStar Pro M Series won't come on when I push the button

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
TX Girl
Posts: 2
Joined: Tue Jul 22, 2014 9:48 am

RemStar Pro M Series won't come on when I push the button

Post by TX Girl » Tue Jul 22, 2014 9:52 am

I have a 2009 RemStar Pro M Series that won't always come on when I push the button, I sometimes have to push it 4 or 5 times and listen for a beep before it will start. Any suggestions would be greatly appreciated. Thanks!

User avatar
kteague
Posts: 7793
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: RemStar Pro M Series won't come on when I push the button

Post by kteague » Tue Jul 22, 2014 10:22 am

Hopefully someone familiar with this will weigh in soon. I have the same machine and been expecting to have the same problem, as the machine I have is a replacement for one that had a malfunction of the on/off switch. Read at that time it was a known problem. It began with sporadic issues before it got to the point where repeated pushing wouldn't result in it coming on. My question for those familiar is does this machine have an auto on/off setting that can be used to bypass the switch?

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: RemStar Pro M Series won't come on when I push the button

Post by JDS74 » Tue Jul 22, 2014 11:16 am

The RemStar Pro M Series provider manual makes reference to Auto Off only.
Auto On is mentioned in the glossary but not in the details of the settings.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: RemStar Pro M Series won't come on when I push the button

Post by Krelvin » Tue Jul 22, 2014 11:36 am

I never use the button to turn on my M-Series... I just put the mask on and start breathing. I only use the button to turn it off.

Not sure if there a program option for that or not.

The only quirk of the machine is when I unplug it from the wall. When I plug it back in, the unit will not work. I have to remove the small power cable that plugs into the back (round connector) and then plug it back in. Then the machine works. I don't move it often but this happens even when we lose power.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
Craig-Tx
Posts: 232
Joined: Fri Dec 22, 2006 6:56 am
Location: Fort Worth, TX

Re: RemStar Pro M Series won't come on when I push the button

Post by Craig-Tx » Tue Jul 22, 2014 12:15 pm

I currently use an M-Series Auto. The ONLY time I use the ON/OFF button is to turn it off. To turn it on, I simply put on the mask and take a couple of deep breaths.
Auto-On most certainly works. I'm not certain if it is simply there or if it has to be enabled through the provider menu. I do know I have the AUTO-OFF option disabled, so it is not tied to that.

-craig

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started PAP on 1/16/07. - Typical 90% pressure ~ 10 cm H20

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: RemStar Pro M Series won't come on when I push the button

Post by JDS74 » Tue Jul 22, 2014 12:48 pm

Yes, the M Series Auto has Auto On and Auto Off.
The M series Pro doesn't - at least the provider manual manual says not.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: RemStar Pro M Series won't come on when I push the button

Post by palerider » Tue Jul 22, 2014 12:55 pm

JDS74 wrote:Yes, the M Series Auto has Auto On and Auto Off.
The M series Pro doesn't - at least the provider manual manual says not.
my pro does.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: RemStar Pro M Series won't come on when I push the button

Post by JDS74 » Tue Jul 22, 2014 1:54 pm

Could you provide directions to the OP so that they can make the settings on their machine work properly?
Your machine is a Pro and not the Auto - correct?

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: RemStar Pro M Series won't come on when I push the button

Post by palerider » Tue Jul 22, 2014 2:42 pm

JDS74 wrote:Could you provide directions to the OP so that they can make the settings on their machine work properly?
Your machine is a Pro and not the Auto - correct?
yeah, my first auto was a 560 that I got in march, and have since replaced.

it doesn't *have* a setting (I just checked the machine... it has mask alert, and auto off, but if you wheeze on the hoze a bit, it'll start up on it's own.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Guest

Re: RemStar Pro M Series won't come on when I push the button

Post by Guest » Tue Jul 22, 2014 6:42 pm

me thinks it is a safety feature built into the machines so they will self-start in the event there is a power failure
not something you can turn off or on
if it doesnt start from breathing a few breaths you likely have a mask leak

User avatar
Craig-Tx
Posts: 232
Joined: Fri Dec 22, 2006 6:56 am
Location: Fort Worth, TX

Re: RemStar Pro M Series won't come on when I push the button

Post by Craig-Tx » Tue Jul 22, 2014 6:52 pm

Just checked my M-series Auto (different machine, I know, but similar) and it has the auto-off option, but auto-on just works. There is not an option in the clinition menu.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started PAP on 1/16/07. - Typical 90% pressure ~ 10 cm H20

Guest

Re: RemStar Pro M Series won't come on when I push the button

Post by Guest » Tue Jul 22, 2014 7:33 pm

Craig-Tx wrote:Just checked my M-series Auto (different machine, I know, but similar) and it has the auto-off option, but auto-on just works. There is not an option in the clinition menu.
zactly you cant shut it off

TX Girl
Posts: 2
Joined: Tue Jul 22, 2014 9:48 am

Re: RemStar Pro M Series won't come on when I push the button

Post by TX Girl » Thu Jul 31, 2014 2:49 pm

Thank you everyone for your comments and suggestions. I will try mine and see if it comes on when I put the mask on. I'll let you know.