SLEEPYHEAD NOT DOWNLOADING

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
jwerley
Posts: 383
Joined: Thu Mar 29, 2012 7:31 pm
Location: SOCAL

SLEEPYHEAD NOT DOWNLOADING

Post by jwerley » Fri Jul 18, 2014 10:30 am

I am having problems downloading Sleepyhead onto my new laptop HP windows 8.1. Is it not compatible, or is it me , not knowing what I am doing. I am trying to download the 9.6 newest version, and it just doesn't do anything.

I am not the sharpest knife in the drawer when it comes to technical things, but I have not had problems downloading other things so far.[/s

I will appreciate any help.....Thank you!

Janice

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ASV Adapt 36037/Ruby Chin Strap/Chin-Ups/ Sleepyhead
"He who understands you is greater kin to you than your own brother. For even your own kindred may neither understand you nor know your true worth."
Kahlil Gibran

User avatar
Pugsy
Posts: 65307
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SLEEPYHEAD NOT DOWNLOADING

Post by Pugsy » Fri Jul 18, 2014 11:20 am

Where are you trying to download from?
http://sleepfiles.com/SH/
I just tried it and it was slow to start but eventually it did start to download.
It is compatible with Win 8.1 so that isn't the issue.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: SLEEPYHEAD NOT DOWNLOADING

Post by robysue » Fri Jul 18, 2014 12:22 pm

jwerley wrote:I am having problems downloading Sleepyhead onto my new laptop HP windows 8.1. Is it not compatible, or is it me , not knowing what I am doing. I am trying to download the 9.6 newest version, and it just doesn't do anything
Are you sure you are clicking on the RIGHT download link? The OpenGL version is ONLY available on the Macs; the BrokenGL version will work on your Window 8.1 machine EVEN IF your machine tries to tell you that you need the OpenGL version.

Here's a picture of the download links at http://sleepfiles.com/SH/ with the appropriate links both Windows and Macs clearly identified:
Image

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
jwerley
Posts: 383
Joined: Thu Mar 29, 2012 7:31 pm
Location: SOCAL

Re: SLEEPYHEAD NOT DOWNLOADING

Post by jwerley » Fri Jul 18, 2014 12:59 pm

I have downloaded:

--> Latest Windows version SleepyHead v0.9.6 [mirror

All I get are the files, the actual home page does not open. Also, there is no sleepyhead icon?? I must be doing something wrong, I just don't know what. the file says" zip file"

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ASV Adapt 36037/Ruby Chin Strap/Chin-Ups/ Sleepyhead
"He who understands you is greater kin to you than your own brother. For even your own kindred may neither understand you nor know your true worth."
Kahlil Gibran

User avatar
RogerSC
Posts: 1917
Joined: Sat Jul 21, 2012 1:11 pm

Re: SLEEPYHEAD NOT DOWNLOADING

Post by RogerSC » Fri Jul 18, 2014 1:24 pm

Here's how I installed this:

1. Backed up my current SleepyHead data folder just for safety.

2. Moved the old SleepyHead folder in Programs(x86) to SleepyHead.0.9.3.

3. Created a new SleepyHead folder in Programs(x86) and dumped all the extracted junk from the .zip folder into it.

4. Ran the .exe file, and seems to be working (after re-importing all the data from SleepyHead's own backup).

Keep in mind that this isn't a installation package, just a pile of files. When I created a shortcut to SleepyHead.exe (in Program Files) on my desktop, this created a cute little sheep icon that I use to invoke it.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: ResMed AirFit N30i mask
Last edited by RogerSC on Fri Jul 18, 2014 2:21 pm, edited 2 times in total.

User avatar
jwerley
Posts: 383
Joined: Thu Mar 29, 2012 7:31 pm
Location: SOCAL

Re: SLEEPYHEAD NOT DOWNLOADING

Post by jwerley » Fri Jul 18, 2014 1:36 pm

Thanks Roger....I am not that computer savvy....I have a computer man coming next Tues. to help me take out hard drives and transfer files from my old computers to the new one. I will have him set up sleepyhead for me, as I have been trying for quite a while now and am starting to get frustrated.

Janice

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ASV Adapt 36037/Ruby Chin Strap/Chin-Ups/ Sleepyhead
"He who understands you is greater kin to you than your own brother. For even your own kindred may neither understand you nor know your true worth."
Kahlil Gibran

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: SLEEPYHEAD NOT DOWNLOADING

Post by robysue » Fri Jul 18, 2014 1:54 pm

jwerley wrote:I have downloaded:

--> Latest Windows version SleepyHead v0.9.6 [mirror

All I get are the files, the actual home page does not open. Also, there is no sleepyhead icon?? I must be doing something wrong, I just don't know what. the file says" zip file"
There is no installer for SleepyHead 0.9.6.

You need to double click the *.zip file to unzip it. Then go into the unzipped folder and look for the SleepyHead.exe file. Right click on that file and select "pin to task bar" or "pin to start menu" so that you don't have to go looking for the SleepyHead.exe file to start SleepyHead.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
RogerSC
Posts: 1917
Joined: Sat Jul 21, 2012 1:11 pm

Re: SLEEPYHEAD NOT DOWNLOADING

Post by RogerSC » Fri Jul 18, 2014 2:22 pm

jwerley wrote:Thanks Roger....I am not that computer savvy....I have a computer man coming next Tues. to help me take out hard drives and transfer files from my old computers to the new one. I will have him set up sleepyhead for me, as I have been trying for quite a while now and am starting to get frustrated. :?

Janice
Understood, sounds like a good way to go. Enjoy using the new version, it does have a lot of improvements.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: ResMed AirFit N30i mask

User avatar
ShelaghDB
Posts: 233
Joined: Mon Dec 23, 2013 8:30 pm

Re: SLEEPYHEAD NOT DOWNLOADING

Post by ShelaghDB » Fri Jul 18, 2014 2:31 pm

Not sure what CPAP machines you are using but as much as I like Sleepyhead and have it installed on 2 of my MacBooks, I am an apple person but like to keep my fingers in Windows as well so i recently purchased a nice Samsung laptop as it will work best with my Samsung S5 phone whilst my MacBooks work best with my iPads and iPhones..............so I was surprised when I downloaded the RESMED software and threw it on the Samsung that I quite liked it.

Mind you, I don't like Windows 8 and went back to Win 7 so its on that OS but its a nice software...........i don't seem to notice too many people using it but so far I am very happy with it as well.
So perhaps its an option if you just happen to find Sleepyhead to difficult to install..?

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Prescription: 13 Pressure
 5k 27" Retina iMac,
 MacBookPro 10.9.5 13" 2.53Ghz + 10.9.5 13" 2.4Ghz
 iPad 4 16Gig + 3G  iPhone 5 + One + One 64gig CyanogenMod Version
 Apple TV + Airport Extreme

User avatar
jwerley
Posts: 383
Joined: Thu Mar 29, 2012 7:31 pm
Location: SOCAL

Re: SLEEPYHEAD NOT DOWNLOADING

Post by jwerley » Fri Jul 18, 2014 3:06 pm

WOO HOO! I stuck with it and finally got it. Do not ask me how.....
It feels good when you are so dumb...then the lightbulb comes on!!

Than you for your help everyone!

Janice

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ASV Adapt 36037/Ruby Chin Strap/Chin-Ups/ Sleepyhead
"He who understands you is greater kin to you than your own brother. For even your own kindred may neither understand you nor know your true worth."
Kahlil Gibran

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SLEEPYHEAD NOT DOWNLOADING

Post by palerider » Fri Jul 18, 2014 5:58 pm

jwerley wrote:WOO HOO! I stuck with it and finally got it. Do not ask me how.....
It feels good when you are so dumb...then the lightbulb comes on!!

Than you for your help everyone!

Janice
now, remember to always slide the little lock on the sd card to 'lock' before you stick it in your winblows puter, cuz winblows likes to leave 'tracks' on the card that the S9 doesn't like.

then remember to unlock the card before putting it in the vpap, or it'll grumble at you on it's screen

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jwerley
Posts: 383
Joined: Thu Mar 29, 2012 7:31 pm
Location: SOCAL

Re: SLEEPYHEAD NOT DOWNLOADING

Post by jwerley » Fri Jul 18, 2014 6:08 pm

Thanks....I have already learned that one.... microsoft does not improve our lives with this edition, but I guess we will get used to it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ASV Adapt 36037/Ruby Chin Strap/Chin-Ups/ Sleepyhead
"He who understands you is greater kin to you than your own brother. For even your own kindred may neither understand you nor know your true worth."
Kahlil Gibran

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SLEEPYHEAD NOT DOWNLOADING

Post by palerider » Fri Jul 18, 2014 6:43 pm

jwerley wrote:Thanks....I have already learned that one.... microsoft does not improve our lives with this edition, but I guess we will get used to it.
I just use linux

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.