Snores and Snorts?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Luthie2006
Posts: 363
Joined: Thu May 29, 2014 6:32 am

Snores and Snorts?

Post by Luthie2006 » Wed Jul 09, 2014 9:39 am

After trying several nasal masks and nasal pillows, I decided on the SWIFT FX pillows. (at least mask du jour). Last night I caught myself sleeping in a "snore/snort" mode and it jolted me and woke me up. I thought using an autopap machine, that it is supposed to catch these before the snorts/snores happen? I thought perhaps I would have a high AHI this morning, but the machine this morning said 0.8. And also, my DME stated that if the treatment works, you are not supposed to wake up to go to the bathroom at all in the middle of the night if you stopped drinking liquids at 8pm? I still get up at least once in the middle of the night for bathroom break, instead of 3 times prior to CPAP. Does anyone know why there are still snorts/snores that do not get resolved with the air? My pressure goes up to 14.8, leaks 16. Thank you.

User avatar
Pugsy
Posts: 65326
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Snores and Snorts?

Post by Pugsy » Wed Jul 09, 2014 9:46 am

The machine in auto mode doesn't always have the ability to prevent every little possible collapse of the airway in a time fashion because of the minimum setting and the time it takes to go from the minimum pressure to where the pressure can hold open the airway better.
If it is too much of an annoyance the usual remedy is simply a little more minimum pressure.

Regarding the nocturia....not all needing to pee during the night is related to OSA and if not related to OSA then the best AHI in the world isn't going to fix it.

What does your Flow limitation and snore graph look like?
Not everything is covered by the AHI number.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Luthie2006
Posts: 363
Joined: Thu May 29, 2014 6:32 am

Re: Snores and Snorts?

Post by Luthie2006 » Wed Jul 09, 2014 9:51 am

Thanks Pugsy; my pressure is set at 10-20 on Autopap. (no ramping now) Do you think that I should set the minimum to 11?
Can you please tell me how to do the graph and flow limitation and snore graph? I am not too computer savvy, so if there is a way to explain it to me, I appreciate it. Do I take the chip out of the REs Med S9 and do something with that? Thanks Pugsy! You are better than the sleep doc!

User avatar
Pugsy
Posts: 65326
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Snores and Snorts?

Post by Pugsy » Wed Jul 09, 2014 10:24 am

To get the graphs you have to use the software...like SleepyHead and you have to remove the SD card and put it in your computer and download the data.

I don't know that I would necessarily increase the minimum from 10 to 11 without evaluating the software reports.
I don't like to make changes without seeing what I am dealing with first.
And are we talking about 1 or 2 "snorts" or 10 to 20 snorts with awakenings?

If you need help with SleepyHead process someone else will have to step in as I am leaving the house shortly. Grandson is here for a visit and we have things to do today. Won't be back until late tonight.
I have some slightly outdated SH tutorials on how to use SH in my signature line to get you started though.

It sounds to me like SleepyHead might be easier for you to manage and get the snore or Flow limitation graphs than ResScan might be and that's why I would recommend it for you right now.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Luthie2006
Posts: 363
Joined: Thu May 29, 2014 6:32 am

Re: Snores and Snorts?

Post by Luthie2006 » Wed Jul 09, 2014 10:54 am

Thanks Pugsy: You have a wonderful day. I am at work now anyways, and can't do anything until later perhaps or even tomorrow evening!
Thanks, and let's keep in touch. Have a great day!

Luthie2006
Posts: 363
Joined: Thu May 29, 2014 6:32 am

Re: Snores and Snorts?

Post by Luthie2006 » Wed Jul 09, 2014 4:27 pm

Hello Pugsy, well I did receive the report from my doctor from the chip on the Res Med S9. Is there a place that I can attach it to you? I don't have a scanner at home but I have it in a PDF attachment.
I really would love for you to see this graph from the sleep doc from my Res Med. Thanks Pugsy.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Snores and Snorts?

Post by palerider » Wed Jul 09, 2014 5:28 pm

Luthie2006 wrote:Hello Pugsy, well I did receive the report from my doctor from the chip on the Res Med S9. Is there a place that I can attach it to you? I don't have a scanner at home but I have it in a PDF attachment.
I really would love for you to see this graph from the sleep doc from my Res Med. Thanks Pugsy.
robysue is working on a beginners guide to sleepyhead over on 'another board':
http://www.apneaboard.com/forums/Thread ... Discussion

take a look at it, it should help you get sleepyhead for your computer, and you can start making your own neat graphs and monitor your progress better!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Snores and Snorts?

Post by library lady » Wed Jul 09, 2014 7:11 pm

@Luthie, you cannot copy it directly into your post, you need to link to it from a photo host such as Photobucket or Tinypic; there are some others, but I'm not sure of the names.. these two are free. I use Tinypic, so these directions are for it, the others are probably quite similar.

To attach your document:
1. save it as a jpg on your computer (use snipping tool or a screenshot)
2. Upload the image and save it
3. follow the steps to: 1.image, 2.resize in dropdown box to Message Board
4. copy the code for Message Board to Word or other word processor
5 Create your post here at cpaptalk.com, pasting the code in where you want it.
Resizing for Message Board use is important because otherwise your complete image might not show up.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.