Post by Ken Pulse Oximeter » Sun Jul 06, 2014 9:03 pm

Trying unsuccessfully to upload my data from wrist unit CMS50F - downloaded manual and programs from internet - get hung up where it says "On the main menu interface, short press the button to move the menu choice bar to "Upload", then long press the button to choose whether to transmit the data to the computer for displaying and analyzing or not". Is the main menu interface on the wrist devise or on the SpO2 software? What button? I've tried short pressing everything I can see on both and nothing happens. This is the blue faced Pulse Oximeter. Any enlightenment would be most welcome. Ken

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS50F

Post by palerider » Sun Jul 06, 2014 10:52 pm

Ken Pulse Oximeter wrote:Trying unsuccessfully to upload my data from wrist unit CMS50F - downloaded manual and programs from internet - get hung up where it says "On the main menu interface, short press the button to move the menu choice bar to "Upload", then long press the button to choose whether to transmit the data to the computer for displaying and analyzing or not". Is the main menu interface on the wrist devise or on the SpO2 software? What button? I've tried short pressing everything I can see on both and nothing happens. This is the blue faced Pulse Oximeter. Any enlightenment would be most welcome. Ken
it sounds like they're talking about the oximeter,

I have a 50I and it just has to be on for the software to find it and dl data, so, that's different.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
wm_hess
Posts: 334
Joined: Sat Oct 16, 2010 2:39 am

Re: CMS50F

Post by wm_hess » Mon Jul 07, 2014 12:19 am

Hi I have the old style (white face) model. For me this is what I do.

Attach USB cord to computer and oximeter
Turn oximeter on
Start up SPO2
Click on the New Sessions button (magnifying glass) on the bottom of SPO2 Review
Fill in what info I want on the name/weight area.
Make sure the connection is showing a com port (com3 for me)
Press ok. SPO2 Review shows a box stating Device Connected, Waiting for Data
On the CMS50 press and hold the white long button on the bottom of the face of the oximeter for about 2 seconds.
A new screen appears that has brightness at the top and near the bottom the word upload. tap the long white button on the bottom of the CMS50 screen and the menu bar moves down 1 entry at a time. When you get upload highlighted, press and hold the long white button again and the information is uploaded to SPO2 Review.

Please remember this is for the old style of the unit, so I cannot tell you if it will work with you, but it's probably similar.

-Bill

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: CMS50F

Post by JDS74 » Mon Jul 07, 2014 6:46 am

Short Press means just press and release.
Long Press means press and hold for a while as suggested 2 seconds is about right.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.