Hi,
I cannot get my M-series Heated Humidified DOM unit to run on anything except a 115 volt plug at an RVV park or house power. I just got a new 12 volt cable and it will not work, I have a 400 watt converter and it will not work, the CPAP will not start ( the blower will not start although the light lites) I will check all the voltages next as the battery might be low.
Thanks
Carolyn
Power Problems
-
Guest
Re: Power Problems
I was going to try to help but I don't understand....Carolyn Bossard wrote:Hi,
I cannot get my M-series Heated Humidified DOM unit to run on anything except a 115 volt plug at an RVV park or house power. I just got a new 12 volt cable and it will not work, I have a 400 watt converter and it will not work, the CPAP will not start ( the blower will not start although the light lites) I will check all the voltages next as the battery might be low.
Thanks
Carolyn
What is a DOM unit?
Where did you get this 12V cable?
I do know some of the older (1st generation) M series humidifiers are not designed to work on 12V. You can still use it in passover.
Have you checked the voltages yet?
Re: Power Problems
Carolyn Bossard wrote:Hi,
I cannot get my M-series Heated Humidified DOM unit to run on anything except a 115 volt plug at an RVV park or house power. I just got a new 12 volt cable and it will not work, I have a 400 watt converter and it will not work, the CPAP will not start ( the blower will not start although the light lites) I will check all the voltages next as the battery might be low.
Thanks
Carolyn

Is this label similar to the one on the back of your humidifier?
If so, then look on page 17 of the user manual for your machine where it discusses options for battery power.
Here is a link to the manual if you don't have one.NOTE: When the DC Power Cord is used, the humidifier will not heat up unless it is also plugged into an inverter. The blue LED will flash when the heater plate does not heat up. In this mode, the humidifier acts as a passover, or non-heated, humidifier.
M Series Humidifier Manual
Hope this helps.
_________________
| Mask: Oracle HC452 Oral CPAP Mask |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Power Problems
There are two flavors of M Series humidifiers. One runs off of wall power, one runs off of DC from the power supply brick. If there are two cords coming from the brick to the CPAP/humidfier, it's the AC version. The DC version has a cord going from the brick to the humidifier, and then a jumper plug over to the blower unit.
I'm not sure this explains Carolyn's problems, but it may help avoid some confusion.
I'm not sure this explains Carolyn's problems, but it may help avoid some confusion.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.