AHI still erratic, time to wingding?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
College3girls
Posts: 95
Joined: Wed Feb 05, 2014 6:46 pm
Location: Upstate NY

AHI still erratic, time to wingding?

Post by College3girls » Fri Jun 06, 2014 10:05 pm

I'm posting Jun 1-4 data for Mom (theresmp). When her AHI is good, it's great. When it's bad, it's bad. No middle ground. No consistency.
She can't tell is she is sleeping on her back. She's been using a blanket roll behind her thinking that would keep her from rolling supine, but the bed is so big (King), I'm not sure that is working.
Pugsy and Robysue, you have been commenting on this whole mess. Should I make changes to her settings, and if so, what should they be? I have my APAP all figured out for me, but I don't know enough about BIPAP to know where to start making changes. Pugsy had mentioned increasing the EPAP, I think. There is no way I could get video of Mom sleeping. She lives an hour away from me, and I've had to struggle to get to her as often as I have in the past few weeks to download sleepyhead on my computer so I could post for her. She will be able to make setting changes. I showed her how to access her clinical settings, so I could call her with changes.
As always, thanks for the input.
Image
Image
Image
Image
Image
Image
Image
Image

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Assisting Mom, on Resmed BIPAP with climate control; w/Amahara View FF
Ready for a good night's sleep.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: AHI still erratic, time to wingding?

Post by robysue » Sat Jun 07, 2014 12:26 am

College3girls,

Here's what I notice in the data you've posted:

Even on the good nights, there seem to be some thick clusters of events associated with some extremely steep rises in pressure. Most of the events are scored as OAs, but in the longest, most persistent clusters on the bad nights, there are a fair number of CAs mixed in with the OAs once the pressures are above IPAP = 21; EPAP = 17 (I'll use IPAP/EPAP in the rest of this post.) It's almost as if once the pressure gets above 21/17 your mom's breathing is at risk of becoming highly unstable. It doesn't happen every time the pressure gets that high, but it when it does, your mom has a doozy of a cluster that can last for 30 minutes to an hour or more. The two clusters on the night of June 1 are examples of what I mean. The thirty minute cluster towards the end of the night on June 3 is another example. But the really bad example is that two hour cluster on June 4.

Another thing that stands out in your mom's data: At the start of the worst of her clusters, there is always a sharp rise of as much as 6 cm in pressure in a relatively short period of time. For example, at the beginning of that nasty cluster on June 2, there's a total pressure increase of 8.5 cm, and that increase includes a vertical rise in pressure of about 6.5 cm that takes place in maybe 3-4 minutes right at the beginning of the cluster. Maybe that sharp pressure increase causes her to blow off too much CO2, which can then trigger a cycle of alternating hypoventilation and hyperventiation. Or perhaps that sharp rise in pressure is simply causing your mom to arouse and move around and then she winds up on her back and that's what keeps the cluster going for so long.

You'll notice that during the times when the pressure is highest (above 21/17), the respiratory rate is also jumping all over the place. Is her minute ventilation and tidal volume also jumping all over during the really bad clusters? It would be really useful to see a zoomed in version of some of these clusters of events. And by zoomed in, I mean zoomed in far enough to see the individual breaths more clearly.

And when these clusters occur, the timing of the beginning of the events still looks to me like they're likely starting in REM sleep, but for some reason they're persisting long after a REM cycle ought to be over. One possible explanation is that the really sharp increase in pressure over a very short period of time is what's causing the destabilization in your mom's breathing. Maybe that sharp pressure increase causes her to blow off too much CO2, which can then trigger a cycle of alternating hypoventilation and hyperventiation, and it's a CO2 undershoot/overshoot cycle that continues to feed the cluster for a very long time once it starts.

So here's what I'd be inclined to think about this were my data:

1) I'd start by bumping the min EPAP up. It looks like your mom's current min EPAP = 11 and PS = 4, so her min pressures (after the ramp goes off) are 15/11. And that's just not high enough to prevent the nasty clusters from getting started in the first place. Except for the best night, her median EPAP is running around 13.5-14.5 cm and her 95% EPAP is a lot higher. So she probably needs at least 2 cm more of EPAP pressure, but making a 2cm jump in min EPAP is a pretty big change. I'd start by bumping min EPAP up to 12 and leave it there for 4-7 days gathering more data. If the data is still all over the place, I'd bump EPAP up to 13 and leave it there for 4-7 days gathering more data.

2) But at the same time, I really think there's something strange going on when your mom is dealing with pressures above 21/17. I'm still worried that some of those OAs scored during the longest clusters when the pressure is at or above 21/17 for 30 minutes or more might just be mis-scored central events. Certainly in the longest of these types of clusters, there's a large number of CAs mixed in with the OAs. And so I'm wondering if your mom's CO2 levels start to go a bit crazy once the pressure gets that high. So I would lower the max IPAP down quite a bit from the current max IPAP = 25. How far to lower it? Well, that nasty cluster on June 4 really gets going when the pressure is above 23/19, but it keeps going even after the pressure dips as low as 22/18. Personally, I'd try lowering the max IPAP all the way down to 21 for a few days and seeing if the "high pressure clusters" get better or worse. If you see real improvement in the number and length of the "high pressure clusters", then you'll know that your mom is sensitive to extremely high pressure, and that the max IPAP does need to be set below the default of 25 cm. On the other hand, if the "high pressure clusters" get worse, then you'll know that those clusters probably ARE mainly obstructive and your mom needs the higher pressure (and probably needs an even higher EPAP in the long run.)

3) By raising the min EPAP and lowering the max IPAP at the same time, this may also reduce the tendency of your mom's VPAP to increase the pressures quite drastically over a very short period of time. And that in turn may also help stabilize her breathing pattern.

So here's what I'd propose as an idea to try for the next week or so:
  • Increase min EPAP to 13
  • Decrease max IPAP to 21
  • Keep PS = 4
  • Keep the ramp settings the same
These settings will allow EPAP to range from 13 up to 17 and IPAP to range from 17 up to 21. Even though the max pressure allowed will be less than the current settings, the higher min EPAP will do two things: It will make it a bit harder for a nasty cluster to start in the first place and once a cluster does start, the machine won't have to increase the pressure quite so far to get to the needed pressure.

And capping the IPAP and limiting the maximum pressure increase to a total of 4cm should help give your mom's breathing a chance to stabilize after a more modest pressure increase once a REM-induce cluster of events does occur. And capping the pressures at 21/17 may help prevent the super long clusters from happening if they are the result of unstable breathing due to some kind of a CO2 overshoot/undershoot problem that develops when when the pressures stay above 21/17 for long periods of time.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Todzo
Posts: 2014
Joined: Tue Apr 24, 2012 8:51 pm
Location: Washington State U.S.A.

Re: AHI still erratic, time to wingding?

Post by Todzo » Sat Jun 07, 2014 2:49 am

Talk with your doctors about [1,2].

[1]: Gilmartin G, McGeehan B, Vigneault K, Daly RW, Manento M, Weiss JW, Thomas RJ.
Treatment of positive airway pressure treatment-associated respiratory instability with enhanced expiratory rebreathing space (EERS).
Source: J Clin Sleep Med. 2010 Dec 15;6(6):529-38. Division of Pulmonary, Critical Care and Sleep Medicine, Beth Israel Deaconess Medical Center, Boston, MA, USA.
Link: http://www.ncbi.nlm.nih.gov/pubmed/21206741

[2]: Dynamic CO2 therapy in periodic breathing: a modeling study to determine optimal timing and dosage regimes
Yoseph Mebrate, Keith Willson, Charlotte H. Manisty, Resham Baruah, Jamil Mayet, Alun D. Hughes, Kim H. Parker and Darrel P. Francis
J Appl Physiol 107:696-706, 2009. First published 23 July 2009; doi: 10.1152/japplphysiol.90308.2008
Link: http://www.ncbi.nlm.nih.gov/pubmed/19628721
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: AHI still erratic, time to wingding?

Post by Sludge » Sat Jun 07, 2014 2:56 am

College3girls wrote:AHI still erratic, time to wingding?
It's dial wingin', not wingdingin'!!
You Kids Have Fun!!

User avatar
Todzo
Posts: 2014
Joined: Tue Apr 24, 2012 8:51 pm
Location: Washington State U.S.A.

Re: AHI still erratic, time to wingding?

Post by Todzo » Sat Jun 07, 2014 6:47 am

amazing
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!

User avatar
College3girls
Posts: 95
Joined: Wed Feb 05, 2014 6:46 pm
Location: Upstate NY

Re: AHI still erratic, time to wingding?

Post by College3girls » Sat Jun 07, 2014 7:52 pm

Thank you Robysue. I am calling Mom tonight to have her change her settings. I will plan a trip to her in a week to download her SD card to my computer again. She is not tech savy enough to download to photobucket or be able to post anything, which is why I've been figuring out how to do it to the best of my ability. I've just begun to figure out how to post the pictures so it doesn't take me hours to do it.

That said, I will try to pinpoint the date to show individual breaths for you. What graphs should I be sure to include in one shot to compare?

So wingding is the wrong term. I know I'm not at all an expert in this. I've only been on APAP myself since February. What I do know is Mom needs someone's help with this, because for over a year now, her treatment is not effective and all the medical community cares about is compliance. I want my Mom to have many more good years. As far as speaking to her sleep Dr., that's a joke. She's never met the actual Dr that signs the prescription. She's only seen a FNP, who first declared her therapy with an APAP auto was effective, with a consistent AHI of 14-22. She's had zero good days for a year with that. If it weren't for the fact that I was diagnosed with sleep apnea, with a much lower AHI before treatment than what Mom had with treatment, I wouldn't have known any better either.

Robysue, should I pm you with the data in a week or continue to post on the forum? Again, I can't thank you enough.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Assisting Mom, on Resmed BIPAP with climate control; w/Amahara View FF
Ready for a good night's sleep.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: AHI still erratic, time to wingding?

Post by robysue » Sat Jun 07, 2014 8:33 pm

College3girls wrote: That said, I will try to pinpoint the date to show individual breaths for you. What graphs should I be sure to include in one shot to compare?
I'd like to see:
  • The events table
  • The zoomed in wave flow graph
  • The zoomed in pressure graph
  • The zoomed in flow limitation graph
  • The zoomed in respiratory rate graph
  • The zoomed in tidal volume graph
  • The zoomed in leak graph is nice but optional
Note that all the graphs except the events table will all zoom in at the exact same time to the exact same level when you select a chunk of any of them to zoom in on.
Robysue, should I pm you with the data in a week or continue to post on the forum? Again, I can't thank you enough.
Whichever. If you'd rather take this off the main forum, that's fine with me. But if you think others might gain from reading the thread, post to the forum. If I don't get back send me a PM telling me where the post is.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI still erratic, time to wingding?

Post by palerider » Sat Jun 07, 2014 9:11 pm

College3girls wrote:Thank you Robysue. I am calling Mom tonight to have her change her settings. I will plan a trip to her in a week to download her SD card to my computer again. She is not tech savy enough to download to photobucket or be able to post anything, which is why I've been figuring out how to do it to the best of my ability.
do you think you could set up a batch file for her to zip up the contents of the sd card, and email them to you?

then you can unzip the files and import them into your sleepyhead...

and/or email the file to robysue

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sun Jun 08, 2014 2:14 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
College3girls
Posts: 95
Joined: Wed Feb 05, 2014 6:46 pm
Location: Upstate NY

Re: AHI still erratic, time to wingding?

Post by College3girls » Sat Jun 07, 2014 10:17 pm

I think Mom's mess may help someone else, so I'm continuing to post here.
Robysue, hopefully I have the right graphs, with them zoomed at the right places. If not, be patient with me. I'll keep trying until I get them right. (Hubby would say I am very trying)
Image
Image
Image
Image

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Assisting Mom, on Resmed BIPAP with climate control; w/Amahara View FF
Ready for a good night's sleep.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: AHI still erratic, time to wingding?

Post by robysue » Sun Jun 08, 2014 1:03 am

College3girls,

I'd like the Flow Wave data (the trace of the actual breaths) as well as the graphs you've shown. To save some room and get it all on the one graph: Drop the Event table and squash the other graphs to make room for the Flow Wave graph. You can squash the vertical height of the Pressure graph quite a bit if you need too.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: AHI still erratic, time to wingding?

Post by robysue » Sun Jun 08, 2014 1:09 am

palerider wrote:
College3girls wrote:Thank you Robysue. I am calling Mom tonight to have her change her settings. I will plan a trip to her in a week to download her SD card to my computer again. She is not tech savy enough to download to photobucket or be able to post anything, which is why I've been figuring out how to do it to the best of my ability.
do you think you could set up a batch file for her to zip up the contents of the sd car, and email them to you?

then you can unzip the files and import them into your sleepyhead...

and/or email the file to robysue
College3girls,

palerider has a good idea if you want to do it. You can continue to post here for now with the data that you have in SH on your computer. But if we're still actively working on ideas AND if the data for this week looks really bad once you go back to your mom's next week to download the data, then shipping me the SD card contents is not a bad idea. If you decide to do this, then PM me for my email address and I'll also give you instructions if you don't know how to save the SD card's contents to your hard drive or create the zip file.

I've done this for a couple of other people in the past. I typically ask permission to post my thoughts about the data to the forum rather than in a PM. And I also ask permission to post whatever screen shots I think are needed in my reply.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: AHI still erratic, time to wingding?

Post by Sludge » Sun Jun 08, 2014 4:15 am

robysue wrote:...shipping me the SD card contents is not a bad idea. If you decide to do this, then PM me for my email address and I'll also give you instructions if you don't know how to save the SD card's contents to your hard drive or create the zip file.
You can also upload it to Dropbox or GoogleDocs.
You Kids Have Fun!!

User avatar
College3girls
Posts: 95
Joined: Wed Feb 05, 2014 6:46 pm
Location: Upstate NY

Re: AHI still erratic, time to wingding?

Post by College3girls » Sun Jun 08, 2014 9:50 am

I like the idea of the zip files and email. Mom does know how to download her SD card to sleepyhead, but I have never created or used a zipfile. I'm willing to learn if someone would like to pm with instructions.

It would be great if mom could get her data to me daily, and then I would have it to pass on to forum members trying to untangle all the knots of her treatment.

Please feel free to post anything on the forum.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Assisting Mom, on Resmed BIPAP with climate control; w/Amahara View FF
Ready for a good night's sleep.

User avatar
College3girls
Posts: 95
Joined: Wed Feb 05, 2014 6:46 pm
Location: Upstate NY

Re: AHI still erratic, time to wingding?

Post by College3girls » Sun Jun 08, 2014 7:11 pm

Here on the more zoomed in graphs for Jun 1-4, showing the wave flow. I tried to get areas where the events clustered and the pressure changed rapidly.
Image
Image
Image
Image

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Assisting Mom, on Resmed BIPAP with climate control; w/Amahara View FF
Ready for a good night's sleep.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: AHI still erratic, time to wingding?

Post by robysue » Sun Jun 08, 2014 10:25 pm

College3girls,

I still think there's something odd going on every time your mom gets above 21/17 (or so it seems) in this data. Even when there are NOT apneas scored and the pressure is that high, the breathing looks funny: The last figure you posted is a closer look at what seems to be going on in all the nasty clusters. Here's a marked up version of that figure:

Image

The pressure increases about 8--8.5 cm in 3.5--4 minutes in response to the OAs and flow limitations in the first red box. That's a huge pressure increase in a very short period of time. And instead of the breathing stabilizing as a result of this pressure increase, it seems to get MORE unstable and MORE strange looking.

By 23:12, the pressure is near its peak level and right after the OA scored around then, a really strange looking pattern has clearly developed and the pattern persists for the rest of this zoomed in shot. The disturbed breathing has a pattern that is really regular in both time and shape---there's a huge increase in flow rate (possible hyperventilation) about every 2 minutes followed by a sharp decrease the flow rate. Even after the apneas "stop" the breathing remains very disturbed. Throughout the green box, the nadir of the 2-minute cycle doesn't (quite) get small enough to be scored as either a hypopnea or an apnea, but it's still there. And then the cycle starts all over again. And the pattern continues from right after the OA scored at 23:12ish all the way until 23:36. By the end of this period, the flow rate is sufficiently low at the nadir of the cycle that it does get scored as a pair of some kind of events and an OA at the very end. The whole period is a full 24 minutes of really weird looking breathing.

But there's also significant flow restriction being scored here as well.

So I still think that the extra high pressure may be somehow contributing to the clearly unstable breathing here. But I'm also only a well read patient. Maybe the next time Sludge stops by he can say something about whether what I'm noticing indicates a possible problem with the high pressures triggering or contributing to the unstable breathing.

Let me know how your mom does with the max IPAP set at 21.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5