using cpap machine to dry out hose

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
springman946
Posts: 158
Joined: Mon Sep 02, 2013 12:40 pm

using cpap machine to dry out hose

Post by springman946 » Thu May 01, 2014 7:51 pm

can anyone tell me if it is a good idea to turn on my cpap machine and allow the air to blow throw the hose to dry it out after washing it someone told me that its not good because you can wear out the machine by doing that

_________________
MachineMask
Additional Comments: my machine is not listed so i going to list my machine manually I am using a Resmed Air Curve 10 ASV

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: using cpap machine to dry out hose

Post by Julie » Thu May 01, 2014 8:10 pm

Your friend is right. But in any case, what's the point of doing the hose at all? It doesn't (shouldn't) get dirty, but if you choose to rinse it once in a while then hang it over the shower rod in the a.m. Even if it doesn't dry out 100% by p.m. what difference will a few drops make compared to using humidification, plus they'll dry out quickly when you have the machine going.

User avatar
LSAT
Posts: 13411
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: using cpap machine to dry out hose

Post by LSAT » Thu May 01, 2014 8:12 pm

Once you get it dry, you will immediately go to bed and it will get wet again. I have been using my hose for over a year without drying it out. I agree...using your machine to dry the hose will wear out your machine sooner. 1 hour a day...365 hours a year.... 2000 hours in 5 years.

User avatar
ChicagoGranny
Posts: 15528
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: using cpap machine to dry out hose

Post by ChicagoGranny » Thu May 01, 2014 8:16 pm

Why are you washing the inside of the hose?
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
Pasta
Posts: 69
Joined: Sun Oct 27, 2013 4:08 am
Location: Maple Valley, WA

Re: using cpap machine to dry out hose

Post by Pasta » Fri May 02, 2014 2:24 am

My DME told me to wash my hose out every couple weeks, and I thought I was being a slacker only doing it every couple months haha. I'd stop washing it altogether if I could. It's a pain to wrestle with it in the sink. I'm starting to see a huge variation in cleaning practices in here haha.

_________________
Machine: Z2 Auto Travel CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ASV 36037, Z2 Auto for backpacking

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: using cpap machine to dry out hose

Post by Julie » Fri May 02, 2014 2:30 am

"I'd stop washing it altogether if I could." Then why don't you just stop?

Johnd80
Posts: 1
Joined: Fri May 02, 2014 12:26 am

Re: using cpap machine to dry out hose

Post by Johnd80 » Fri May 02, 2014 2:34 am

From everything I've read about hoses these are the first posts I have seen which question washing them.
I wash mine about every month in order to prevent the growth of mold which is not good in your lungs.

User avatar
Pasta
Posts: 69
Joined: Sun Oct 27, 2013 4:08 am
Location: Maple Valley, WA

Re: using cpap machine to dry out hose

Post by Pasta » Fri May 02, 2014 2:40 am

Julie wrote:"I'd stop washing it altogether if I could." Then why don't you just stop?
Johnd80 wrote:From everything I've read about hoses these are the first posts I have seen which question washing them. 
I wash mine about every month in order to prevent the growth of mold which is not good in your lungs.
^ Because of this. My DME said to so I figured it was necessary. I can see how bacteria could hang out in there too. Don't know how big a deal it is, I'd be happy if the concensus was it wasn't necessary.

_________________
Machine: Z2 Auto Travel CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ASV 36037, Z2 Auto for backpacking

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: using cpap machine to dry out hose

Post by Julie » Fri May 02, 2014 3:33 am

In 8 yrs of using Cpap I've never ever had mold or anything else unwanted in my hose(s) and haven't heard of anyone else having found any either.. I suppose if you live in a really swampy area you might have a problem, but otherwise, the only thing going through the hose (to you, not from you) is air through the machine's filter.

User avatar
Pasta
Posts: 69
Joined: Sun Oct 27, 2013 4:08 am
Location: Maple Valley, WA

Re: using cpap machine to dry out hose

Post by Pasta » Fri May 02, 2014 3:40 am

Julie wrote:In 8 yrs of using Cpap I've never ever had mold or anything else unwanted in my hose(s) and haven't heard of anyone else having found any either.. I suppose if you live in a really swampy area you might have a problem, but otherwise, the only thing going through the hose (to you, not from you) is air through the machine's filter.
Yeah, that's what I was thinking too, and it's why I don't wash my hose nearly as often as I was told to. Still wasn't sure though. The only thing I could think of was my breath blowing a little way into the hose, but even that seemed unlikely. I'm with you, I doubt it's necessary, I'm just hoping a bunch of people don't rush in to prove us wrong.

_________________
Machine: Z2 Auto Travel CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ASV 36037, Z2 Auto for backpacking

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: using cpap machine to dry out hose

Post by Pugsy » Fri May 02, 2014 6:40 am

Pasta wrote: I'm with you, I doubt it's necessary, I'm just hoping a bunch of people don't rush in to prove us wrong.
Almost always when someone comes in to try to defend rigid cleaning schedules its a newbie. Most of us cpap veterans have long ago given up the rigid cleaning schedule...and we are still here.
I started cpap therapy 5 years ago the end of this month...I have never had anything show up growing in my long hose either.
And I sure don't adhere to any sort of rigid cleaning schedule for anything.
I gave that up about 6 months into therapy when it became old real fast going through the "routine" that I thought I just HAD to do and I found out I didn't have to do it and didn't die a horrible death or catch some nasty bug.

Now there are a few cpap veterans who prefer to still maintain a rigid cleaning schedule but not near as many of those as there are newbies who have had the fear of whatever put in them by someone. I still say...whatever floats your boat and lets you sleep well...that's what a person needs to do. If someone worries about germs and bugs or whatever...then they need to do whatever it is that they need to do to alleviate those concerns.

The need to dry the hose...I have never understood that one. Why not just wash it in the evening and hook it back up to the machine and use it....it's going to get moisture in it anyway.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

houndlover
Posts: 35
Joined: Fri Apr 18, 2014 8:34 pm

Re: using cpap machine to dry out hose

Post by houndlover » Fri May 02, 2014 7:17 am

The first time I was on CPAP 6 years ago I lived in FL...super humid area...my hose grew stuff. :/ Now, that being said I had been sent home from my sleep study with a machine, non-heated hose, and a mask. Was not told anything about cleaning anything....so my humidifier tank also grew some things before I figured out on my own that I should clean it.

Now I have a heated hose and zero condensation..I clean the humidifier tank so I'm not worried about anything growing.

User avatar
lliann
Posts: 372
Joined: Wed Jun 20, 2012 6:33 pm
Location: upstate new york

Re: using cpap machine to dry out hose

Post by lliann » Fri May 02, 2014 7:40 am

Eeeesh.

So, and lets just say I am not a perfectionist nor an alarmist, yet I will try to be informed.


I have been washing my hose. Not often but once in a while. Its a royal pain in the butt washing the hose. Is there a solid reason why I should wash it? Ever..? For any reason? Can it grow "stuff"

Inquiring minds want to know.<g>

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead for Mac s9 auto settings 11-16

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: using cpap machine to dry out hose

Post by Julie » Fri May 02, 2014 7:45 am

IF at some point, whether it's next week or next year you SEE something (or smell something - tho' that would be more likely a machine generated odor) that doesn't belong, then go ahead and deal with it, but if all is well... if it ain't broke... etc. I think I washed a hose... somewhere around 2007... or was it 2009?

User avatar
lliann
Posts: 372
Joined: Wed Jun 20, 2012 6:33 pm
Location: upstate new york

Re: using cpap machine to dry out hose

Post by lliann » Fri May 02, 2014 8:04 am

Julie wrote:IF at some point, whether it's next week or next year you SEE something (or smell something - tho' that would be more likely a machine generated odor) that doesn't belong, then go ahead and deal with it, but if all is well... if it ain't broke... etc. I think I washed a hose... somewhere around 2007... or was it 2009?

Sounds good to me. Less is more.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead for Mac s9 auto settings 11-16