Re: leakage

Post by MaryGordon » Mon Apr 21, 2014 11:15 am

Thank you. I have them set as private on Flickr: the problem is that I don't really want them public because of some of the people who view my Flickr site. I don't want to open a photobucket a/c as I get in too much muddle when I have too many accounts. but I will if this isn't working, as I so much appreciate the help on this forum.

User avatar
Pugsy
Posts: 65384
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: leakage

Post by Pugsy » Mon Apr 21, 2014 11:25 am

Your links send me to Yahoo sign in page and I don't have a yahoo account.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Mon Apr 21, 2014 11:28 am

oh that's rubbish.... I'll try again. Thanks for trying.

User avatar
Bama Rambler
Posts: 189
Joined: Wed Feb 19, 2014 10:25 am

Re: leakage

Post by Bama Rambler » Mon Apr 21, 2014 11:32 am

The links you posted give me a "Page Not Found" error.
Since the earlier one you posted worked in the post I made, they should work.

Did you follow these instructions?
Go to the pic that you want to embed and select the share box.
Then the html/bbcode box.
Under the actual code are 3 buttons, select the BBCode button.
Then select the entire code, copy it and then paste it in the post you want it to show in.

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: CPAP 15cmH2O

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Mon Apr 21, 2014 11:40 am

[img]<a%20href="https://www.flickr.com/photos/mtps/1395 ... l2014"></a>[/img]
this (ie my incompetence!) is driving me mad!! too much to get my head round after a busy day's work!! And I am meant to be good at this sort of stuff believe it or not!

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Mon Apr 21, 2014 11:42 am

Image
or even this...!

User avatar
Bama Rambler
Posts: 189
Joined: Wed Feb 19, 2014 10:25 am

Re: leakage

Post by Bama Rambler » Mon Apr 21, 2014 11:53 am

Looks like they changed the way they do the share dialog a little.

go to the picture you want to share and then click the share button.
In the middle of the dialog that pops up there 3 choices, Link -- HTML -- BBCode, Select the BBCode text.
Select the size you want (i.e. Medium 640x480)
Then select the entire text in the link box and paste it in the post.

I did it for one of my pictures and here's what it looks like.

ImageIMG_20130719_065517 by Bama Rambler, on Flickr

And here's what the code looks like

Code: Select all

[url=https://flic.kr/p/fcsmua][img]https://farm4.staticflickr.com/3746/9318618033_7118531e58.jpg[/img][/url][url=https://flic.kr/p/fcsmua]IMG_20130719_065517[/url] by [url=https://www.flickr.com/people/88683008@N08/]Bama Rambler[/url], on Flickr

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: CPAP 15cmH2O

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Mon Apr 21, 2014 12:28 pm

Your patience is really touching. thank you. Try again!
Imagepressure-leak-resp-21april2014 by Mary Stevens, on Flickr
Imageleak-day20April2014 by Mary Stevens, on Flickr
Imageleaks21April2014 by Mary Stevens, on Flickr

User avatar
Pugsy
Posts: 65384
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: leakage

Post by Pugsy » Mon Apr 21, 2014 12:44 pm

Looks like you start the night with a good leak line but then it goes to heck when the pressure starts to increase likely in response to some sort of apnea event trying to happen and also some of the leaks look like they have a plateau appearance which usually means some sort of mouth openings.

I don't have a working link to ResMed's ResScan guide that shows that example. ResMed moved their stuff all around on their website.

See if this will work for you
https://dl.dropboxusercontent.com/u/978 ... nGuide.pdf

Looks like your minimum pressure is quite low. Perhaps fitting the mask to higher pressure would help or starting with a higher minimum might help reduce any leaking related to pressures. If what is shown is all related to mouth opening though...you have a problem when it exceeds 24 L/min for any length of time..it could be impacting sleep quality and allowing some apnea events to materialize because of sub optimal pressure. Especially when you see over 30 L/min leak and this also affects the machine's ability to sense and flag any events that might occur. Meaning they could happen and the machine not be able to flag them because it can't sense them.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Bama Rambler
Posts: 189
Joined: Wed Feb 19, 2014 10:25 am

Re: leakage

Post by Bama Rambler » Mon Apr 21, 2014 1:41 pm

MaryGordon wrote:Your patience is really touching. thank you. Try again!
And there you go! It's just a matter of figuring out the right sequence of events in order to get them to show up.
Glad you finally got it figured out.

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: CPAP 15cmH2O

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Mon Apr 21, 2014 2:11 pm

You are just so very kind! The drop box link has worked fine and I will look at the material tomorrow.
As for Flickr, should you be interested to see my home town and other places I've lived my photos are herehttps://www.flickr.com/photos/mtps/sets/

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Wed Apr 23, 2014 1:18 am

The leakage problems, and indeed the sleep quality problems, that I was experiencing seem to be well on the way to being controlled.
I have done three things:
  • the straps are looser;
  • I am taping the arms of the mask to my cheeks so that the looser fitting doesn't permit it to slide all over the shop;
  • and as for the chin strap, by chance I found that by wearing the cup of the chin strap actually over my mouth where it is comfortable and not too tight, my mouth doesn't open.
The only thing I have to really remember to do is to take any meds before getting this all strapped and taped into place!
Leakage is down to almost zero.
I also suspect that I was getting a poor oxygen flow before. Once again I am now feeling refreshed in the morning.

User avatar
Pugsy
Posts: 65384
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: leakage

Post by Pugsy » Wed Apr 23, 2014 7:33 am

Glad you seem to have sorted out your issues.

Thank you for the photo tour. I did go to your album and stroll through your pictures. Really nice work. I especially like the castles. Castles have always fascinated me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: leakage

Post by chunkyfrog » Wed Apr 23, 2014 7:45 am

Your pictures are very nice.
Castles are so fascinating.
To think of the things people will do in order to feel safe!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

MaryGordon
Posts: 38
Joined: Fri Mar 28, 2014 5:00 am
Location: Scotland

Re: leakage

Post by MaryGordon » Thu Apr 24, 2014 1:22 am

Thanks guys, very much!