Change pressure on PR 260 SE brick?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
alpha752
Posts: 108
Joined: Sat Mar 08, 2014 8:19 pm

Change pressure on PR 260 SE brick?

Post by alpha752 » Sat Apr 12, 2014 2:10 pm

Is there a way to change the pressure on my old PR 260 SE brick? I was thinking about keeping it as a back up, but I am finding I need a little more pressure then it's set at and would like to be able to adjust it. I know how to on the rest of the series 60 machines, but as it doesn't have a regular LCD or scroll wheel I'm not sure on this POS. Is it possible?

User avatar
Pugsy
Posts: 65282
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Change pressure on PR 260 SE brick?

Post by Pugsy » Sat Apr 12, 2014 3:03 pm

http://www.apneaboard.com/forums/Thread ... structions
Check this link above for instructions on how to enter the setup menu.

You might also request the provider manual from here
Be sure that you follow the instructions and request this wording exactly
Respironics PR System One REMstar SE
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

Your brick would make a good back up machine and you can change the settings by just going into the clinical setup menu.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.