Therapy hours question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
walker3117
Posts: 4
Joined: Sat Mar 29, 2014 8:56 am

Therapy hours question

Post by walker3117 » Sat Mar 29, 2014 9:08 am

I have a remstar auto aflex system one. I'm trying to understand why the therapy hours vary so much (as much as 5 hours) when I use the machine all night every night. Thanks.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Therapy hours question

Post by robysue » Sat Mar 29, 2014 9:32 am

walker3117,
Welcome to the fourm!
walker3117 wrote:I have a remstar auto aflex system one. I'm trying to understand why the therapy hours vary so much (as much as 5 hours) when I use the machine all night every night. Thanks.
First question: Where are you getting the "therapy hours" numbers from? The machine's LCD or SleepyHead or Encore? If you are getting them from the machine's LCD, what time of day are you checking the machine? The PR System One REMStar's update their LCD data at noon Greenwich Meridian Time, and if your normal sleep period crosses the noon GMT line, the LCD will only show the data for before noon GMT. The data recorded after that goes into the next day's numbers.

And then there's this: Therapy hours are "scored" only when the machine can detect actual breathing. In the presence of some large leaks, the machine may be on, but it cannot detect breathing. And that time will not count towards "therapy hours."

So the next question is: What do your leaks look like?

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Therapy hours question

Post by chunkyfrog » Sat Mar 29, 2014 10:30 am

Walker: welcome to the forum.
To more effectively help you, we need you to fill out the equipment portion of your profile.
Sadly, some machines do not have leak data--we hope you have a "good one"

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Wulfman...

Re: Therapy hours question

Post by Wulfman... » Sat Mar 29, 2014 10:44 am

walker3117 wrote:I have a remstar auto aflex system one. I'm trying to understand why the therapy hours vary so much (as much as 5 hours) when I use the machine all night every night. Thanks.
If you're not using software to evaluate your data, I would strongly recommend it. It'll take the guesswork out of the data.
I believe your "problem" is due to WHEN you're checking your data. (if it's in the LCD screen)
As I understand it, the System One doesn't update the information until after Noon. (and that may be GMT, too.)


Den

.

walker3117
Posts: 4
Joined: Sat Mar 29, 2014 8:56 am

Re: Therapy hours question

Post by walker3117 » Sat Mar 29, 2014 3:12 pm

I'm getting the number from the LCD screen. It lists 0 for large leaks and as far as I can tell I'm not having much if any leakage (I feel around the nasal mask for leaks). I check the numbers after 12 noon GMT (5 am PST). Could there be leaks that I can't feel? Maybe I'll try my full face mask and see what happens. I'm really confused about this.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Therapy hours question

Post by Julie » Sat Mar 29, 2014 3:19 pm

Check the #'s before noon for the previous night.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Therapy hours question

Post by robysue » Sat Mar 29, 2014 3:25 pm

walker3117 wrote:I'm getting the number from the LCD screen. It lists 0 for large leaks and as far as I can tell I'm not having much if any leakage (I feel around the nasal mask for leaks). I check the numbers after 12 noon GMT (5 am PST). Could there be leaks that I can't feel? Maybe I'll try my full face mask and see what happens. I'm really confused about this.
The LCD data will only contain the data that's recorded before 5 am PST (noon GMT). If you sleep later than 5 am (PST) the data from 5 AM until wake won't show up until the NEXT day on the LCD. It could be that's what's causing the discrepancy between what the machine says the therapy hours are and what you think they should be.

And yes, there can be leaks you don't feel; much of the time they're not relevant, but sometimes they can be. And very rarely the machine can lose track of your breathing even when the leaks are not in the official Large Leak range.

Rather than switching to a FFM, why don't you try downloading SleepyHead and seeing what the data actually looks like for the full sleep session? SleepyHead will use the time zone setting on your computer to adjust the GMT timestamps on the data so that it shows up in a way that is meaningful relative to your actual sleep schedule.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

walker3117
Posts: 4
Joined: Sat Mar 29, 2014 8:56 am

Re: Therapy hours question

Post by walker3117 » Sat Mar 29, 2014 7:24 pm

Thank you for your help. I can see I need to get a card reader. Does it matter what kind I get? Does it reset the data card every time you take it out?

User avatar
cpapernewbie
Posts: 685
Joined: Fri Oct 13, 2006 5:53 pm

Re: Therapy hours question

Post by cpapernewbie » Sat Mar 29, 2014 9:28 pm

you might not need to buy a card reader. System One is using an SD card - the reader for this SD card is usually already built-in in desktop and laptops since more than 10 years ago.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: CPAP history: dumb tank, auto, PR M, PR System 1, PR BIPAP, PR System 1 model 60, Resmed S9, Resmed S10, Dreamstation
Resmed airsense 10

walker3117
Posts: 4
Joined: Sat Mar 29, 2014 8:56 am

Re: Therapy hours question

Post by walker3117 » Sat Mar 29, 2014 10:36 pm

Thanks, I'll check my computer.

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Therapy hours question

Post by Pugsy » Sun Mar 30, 2014 12:03 am

walker3117 wrote:Does it matter what kind I get? Does it reset the data card every time you take it out?
If your computer doesn't have a slot for the SD card you can get any generic reader/adapter to use a USB port.

Removing the SD card to use the software doesn't reset anything or do anything to the data on the SD card or the machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.