got my resmed s9 ? about leak

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
twink
Posts: 6
Joined: Mon Mar 03, 2014 1:39 pm

got my resmed s9 ? about leak

Post by twink » Fri Mar 07, 2014 10:35 am

Hello,
I picked up the s9 elite yesterday and used it last night. On my resmed s8 my leak, with the same mask I am using was only 0.2 Now with the s9 it's reading 4l/s. Would have to double check to see if it's L/s, or L/m, but I think seconds. Could my leak be worse with this machine. Mask fit says excellent. Do have the new slim line tubing.
Thanks for any replies.
Twink

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: got my resmed s9 ? about leak

Post by avi123 » Fri Mar 07, 2014 10:39 am

It's 4 Liter per Min and it's good b/c it is below 24 L/Min, the allowed maximum.

When I started CPAPing at the beginning of 2011, I had lousy results using my S9 Elite and a full face Fisher & Paykel mask. Looking back at it I can't believe it how bad they were:

Image

Image

Image

Compare those results to my present data with my present machine, new mask, and higher pressure:

{nice results}:

Image

Image


Image



Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Fri Mar 07, 2014 11:16 am, edited 5 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65119
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: got my resmed s9 ? about leak

Post by Pugsy » Fri Mar 07, 2014 10:39 am

The S9 reports in liters per minute. Your S8 reported in liters per second.
With the S9 your leak number to avoid is 24 L/min...so your 4 L/min means minimal leaks .

24 L/min is the same as your old number to avoid with the S8 of 0.40...just multiply liters per second by 60 seconds.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Sat Mar 08, 2014 3:15 am, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

twink
Posts: 6
Joined: Mon Mar 03, 2014 1:39 pm

Re: got my resmed s9 ? about leak

Post by twink » Fri Mar 07, 2014 3:03 pm

Wow, thank you for the replies. Love the data! That's great. So I guess I'm good.!!!
Twink