Smart Start for a Series One REMstar?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Wyowanderer
Posts: 15
Joined: Sun Apr 25, 2010 6:54 pm
Location: Rocky Mountain region, planet Earth

Smart Start for a Series One REMstar?

Post by Wyowanderer » Fri Feb 14, 2014 3:03 pm

My ResMed S8 was set up for smart start- it would automatically start when you breathed into it, and would shut off after 6 seconds of flow with the mask off.
Can the Remstar be set up the same way?

_________________
Mask
Strength and muscle and jungle work........

User avatar
Pugsy
Posts: 65269
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Smart Start for a Series One REMstar?

Post by Pugsy » Fri Feb 14, 2014 3:06 pm

Yes, those options are available just called something different.

In the clinical set up menu...
Auto On.....set to On
Auto Off....set to On

Will let the machine start up after it senses a couple of breaths and and turn off when it senses no more breathing but note that the auto off thing takes about a minute to turn off. Much longer than Resmed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Smart Start for a Series One REMstar?

Post by Janknitz » Fri Feb 14, 2014 4:54 pm

I find I have to block the anti-asphyxiation valve with my finger to get my PRS1 to turn on.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Pugsy
Posts: 65269
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Smart Start for a Series One REMstar?

Post by Pugsy » Fri Feb 14, 2014 8:38 pm

I guess we need to amend remarks. I keep forgetting about that little quirk with Auto On.

If someone is using a full face mask and the pressure used along with the force of your own breathing can't be accurately "sensed" by the machine then it may not turn itself on even with Auto On set to On.

BTW ResMed has a similar issue with the S9 and of all masks..nasal pillow masks...at pressures below 7 cm...sometimes it won't kick in. You would think with that issue the nasal pillow masks would be the last of the masks to have that happen but it can and is mentioned in the manual.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Wyowanderer
Posts: 15
Joined: Sun Apr 25, 2010 6:54 pm
Location: Rocky Mountain region, planet Earth

Re: Smart Start for a Series One REMstar?

Post by Wyowanderer » Wed Feb 19, 2014 12:18 pm

Excellent. Couldn't find it in the clinician's manual. Thanks for the help, all.

_________________
Mask
Strength and muscle and jungle work........