I had the worst time last night.
Any insight or suggestions? I am frustrated. I woke up feeling leaks quite often.

-
Pugsy
- Posts: 65400
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Fri Feb 14, 2014 9:20 am
Same discussion as before...your minimum pressure is allowing too many obstructive events to happen...
leaks that woke you up were probably related to the machine increasing the pressure to try to better prevent the airway collapse.
Do you think you could use 7.5 or 8.0 minimum comfortably?
For sure try 7.0 tonight.
Pattern of clusters is very suspicious for REM Stage sleep...might have some supine sleeping thrown in.
I may have to RISE but I refuse to SHINE.
-
Julie
- Posts: 20051
- Joined: Tue Feb 28, 2006 12:58 pm
Post
by Julie » Fri Feb 14, 2014 9:22 am
Does your mask normally fit well without many leaks?
Are you using anything like a 'Pap pillow'?
Something might have just shifted the mask sideways though you didn't wake up, and it stayed like that for half the night.
-
amybeth
- Posts: 30
- Joined: Thu Jan 09, 2014 7:47 am
Post
by amybeth » Fri Feb 14, 2014 9:23 am
Thanks Pugsy! I had increased my min from 5.0 to 6.0, but will jack that up tonight for sure! At 6.0 I don't even feel anything, feels totally normal. So I think putting it up will be no problem at all.