Mask cushions: Control III or boiling with baking soda?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Mask cushions: Control III or boiling with baking soda?

Post by Sir NoddinOff » Fri Feb 07, 2014 2:33 pm

I'm currently using a Quattro FX Full Face Mask. I tried the ever popular boiling technique on my detachable cushion, which includes water and baking soda to prolong their life, however it's a big PITA. I can say that it seems to work okay tho. Has anybody tried soaking their cushions in Control III disinfectant to restore their viability? Will it ruin the cushions? As a side note: the boiling method doesn't seem to hurt them. Also, I'm not fond of the smell of Control III, but I can later run them thru a bath of Dawn dish detergent solution with the rest of my gear. Thanks all.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
LSAT
Posts: 13395
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Mask cushions: Control III or boiling with baking soda?

Post by LSAT » Fri Feb 07, 2014 2:55 pm

I use baby wipes every night before I put on my mask. I rarely clean the cushion, but, every 2-3 months I boil it.

Wulfman...

Re: Mask cushions: Control III or boiling with baking soda?

Post by Wulfman... » Fri Feb 07, 2014 6:56 pm

I'm a minimalist, but when I do wash my UMFF (which is a lot like the Quattro, which I also have), I use Dawn dishwashing liquid (the original blue). I soak/wash the entire mask once (assembled without headgear) in HOT water in a small plastic bucket (about a gallon size), then rinse twice in HOT water, then let it soak in COLD water for awhile (restores some "spring" to the cushion). Then, shake out most of the water and hand dry with a flour sack dish towel.
I wash the headgear and fleece strap covers in the washing machine with a load of clothes on a cool water setting. We use ALL laundry soap. Then, I hang that stuff to dry on a clothes hanger.

Been doing that for nearly nine years. As I've said before, I get YEARS worth of use out of both the headgear and cushions.


Den

.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Mask cushions: Control III or boiling with baking soda?

Post by Sir NoddinOff » Fri Feb 07, 2014 8:26 pm

Thanks guys. I guess cleaning one's mask cushions with Control III is not being done much on this forum.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Mask cushions: Control III or boiling with baking soda?

Post by chunkyfrog » Fri Feb 07, 2014 10:23 pm

Silicone is highly absorbent. Soaking in anything with an odor is a very bad idea.
The smell may never come out. Soaking can also make the silicone cloudy and affect the texture.
It happened to me.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Mask cushions: Control III or boiling with baking soda?

Post by Sir NoddinOff » Tue Feb 11, 2014 1:48 pm

chunkyfrog wrote:Silicone is highly absorbent. Soaking in anything with an odor is a very bad idea.
The smell may never come out. Soaking can also make the silicone cloudy and affect the texture.
It happened to me.
CF: How do you clean your cushions

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Mask cushions: Control III or boiling with baking soda?

Post by squid13 » Tue Feb 11, 2014 1:54 pm

I'm like wulfman in that I wash my stuff in Dawn dish soap, rinse well and let dry.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Mask cushions: Control III or boiling with baking soda?

Post by chunkyfrog » Tue Feb 11, 2014 4:19 pm

Right now, I use unscented baby wipes (Up &Up from Target)
I use a new cushion every month. I have been keeping the old ones, and may do a boil test.
Wisp is part silicone and part hard plastic (Lexan?) so I'm not sure what will happen.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her
Last edited by chunkyfrog on Thu May 15, 2014 11:24 am, edited 1 time in total.

User avatar
MrStein
Posts: 154
Joined: Mon Oct 15, 2012 2:38 pm
Location: Near Nashville Tn

Re: Mask cushions: Control III or boiling with baking soda?

Post by MrStein » Wed Feb 12, 2014 11:48 am

Do this in this sequence:

Boil 3-4 mins in distilled water with 2 teaspoons of baking soda. Use enough water to cover the cushions.
When finished, rinse in distilled water.
Follow mixing instructions on Control III bottle, soak cushion for ten mins. Rinse and air dry.
Cushion will be like brand new, unless they are yellowed. Nothing seems to get the yellow out.
Even if they are yellow they will still be supple and soft like new. This is the method I used and never fails.
I rotated about 5 cushions and just kept stashing or giving away the ones the DME sent.
I now use the Sleepweaver Anew so I no longer have to do all of this.
The method I describe above is totally safe and really works well.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: 12cm-16cm Auto, Encore Basic 2.0 and SleepyHead
"The future of yesterday is tomorrow's today"

User avatar
knothead
Posts: 538
Joined: Fri Jan 24, 2014 8:58 am
Location: Corpus Christi, Tx

Re: Mask cushions: Control III or boiling with baking soda?

Post by knothead » Wed Feb 12, 2014 12:39 pm

I clean my mask, cushion & hose with Dawn. I simply dunk the mask, cushion in the solution, swish it around, rinse with tap water. I then mix a solution of white vinegar and water and allow all items mentioned to soak for about 10-15 minutes. I then rinse thoroughly, especially my hose, let all air dry and walla...clean smelling equipment. Been using this method for 4 years now with no ill effects. Easy to do also...

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: BP down & in control, Resmed S9 Autoset for backup & travel.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Mask cushions: Control III or boiling with baking soda?

Post by Sir NoddinOff » Wed Feb 12, 2014 5:48 pm

chunkyfrog wrote: Wisp is part silicone and part hard plastic (Lexan?) so I'm not sure what will happen <when boiling>.
CF. It's interesting you should mention that: I use the Wisp mask from time to time and wondered if a person could use some kitchen tongs (while wearing a protective rubber glove) and simply dip the silicone flanges part into boiling water containing baking soda - obviously you'd take care not to immerse the hard plastic part in the hot water. In other words, maybe it's not be necessary to submerge the whole soft/hard cushion to affect a restoration of the soft flanges. Sadly, I don't have a Wisp cushion to experiment with at this time. Wonder if anyone's tried this method yet
knothead wrote: I then mix a solution of white vinegar and water and allow all items mentioned to soak for about 10-15 minutes. I then rinse thoroughly, especially my hose, let all air dry and walla...clean smelling equipment. Been using this method for 4 years now with no ill effects. Easy to do also...
Yeah, Knothead. I've read a lot of positive things about white vinegar. I often use Palmolive anti-bacterial dish soap to clean masks because it has 2% lactic acid as the disinfecting agent - it seems to keep the cushions 'springy' if I don't soak them too long, which seems to cause cloudiness in the flanges.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Mask cushions: Control III or boiling with baking soda?

Post by archangle » Thu Feb 20, 2014 1:57 am

I've had good luck with washing silicone parts in the dishwasher weekly, plus swapping with a spare mask part and letting the cleaned part dry for a whole week.

ResMed does say not to dishwash, but it works well for me.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Mask cushions: Control III or boiling with baking soda?

Post by Sir NoddinOff » Fri Feb 28, 2014 1:52 pm

Here's the best thread I've seen on boiling with baking soda... before and after pics and all
This CPAPtalk link should answer your question to the max
viewtopic/t74686/Boiling-Stuff-Part-2.html

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

Sandra_ON
Posts: 18
Joined: Sun Feb 16, 2014 2:53 pm

Re: Mask cushions: Control III or boiling with baking soda?

Post by Sandra_ON » Fri Feb 28, 2014 3:57 pm

The respiration therapist that I see just said to pour a couple of drops of dishwashing liquid into the mask and rub around with your fingers, rinse and let dry. No mention made of boiling or buying any special products for washing and that's what I've been doing.

_________________
MaskHumidifier
Additional Comments: Pressure at 5/11

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: Mask cushions: Control III or boiling with baking soda?

Post by Bill44133 » Fri Feb 28, 2014 4:05 pm

Sandra_ON wrote:The respiration therapist that I see just said to pour a couple of drops of dishwashing liquid into the mask and rub around with your fingers, rinse and let dry. No mention made of boiling or buying any special products for washing and that's what I've been doing.
The longer you wear the mask the cushions start wear and lose there ability to seal and will leak like crazy. Dipping the cushion in boiling water brings back the silcone to life. Works wonders

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19