Does anyone know how to change the pressure settings for a Devilbiss DV54D.

Wulfman...

Re: Change pressure in a Devilbiss DV54D

Post by Wulfman... » Sat Feb 01, 2014 1:01 pm

Up5Times wrote:Does anyone know how to change the pressure settings for a Devilbiss DV54D.
Of course. Don't you? I downloaded the documents from DeVilbiss' website.

But, here's a snippet to get into the machine.

Den

.

First - With the device pow­ered and turned off, press and hold the DOWN and DELAY buttons.

Second - While holding the two buttons, press the ON/OFF button.
The display will show “Clinical Menu”. You can now access the Clinical Setup Menu.

Use the LEFT but­ton and RIGHT button to navigate through the menu

Use the UP button and DOWN button to adjust the settings


.

Up5times

Re: Change pressure in a Devilbiss DV54D

Post by Up5times » Sun Feb 02, 2014 3:33 am

Thanx Wulfman, Greatly appreciated.. Sleep Well!

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Change pressure in a Devilbiss DV54D

Post by chunkyfrog » Sun Feb 02, 2014 6:16 pm

Props to Den!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Guest

Re: Change pressure in a Devilbiss DV54D

Post by Guest » Mon Feb 03, 2014 3:55 pm

Wulfman,

I got into the DeVilbiss web site and for the life of me I can't find the info you referred to.

Could you please explain how you got it.

Thanx

Wulfman...

Re: Change pressure in a Devilbiss DV54D

Post by Wulfman... » Mon Feb 03, 2014 4:05 pm

Guest wrote:Wulfman,

I got into the DeVilbiss web site and for the life of me I can't find the info you referred to.

Could you please explain how you got it.

Thanx
Well, they changed/deleted the link to the location where they had a WHOLE BUNCH of downloadable files.
Fortunately, I happened to find them and downloaded them before they did. (August of 2013)

Here is one link that I found that will still work to get the quick setup guide. Copy the link and remove the spaces.

http : // www . readbag . com/internetage-ws-cpaphelp-manuals-devilbiss-devilbiss-intellipap-autoadjust-quicksetupguide


Den

.

Up5times

Re: Change pressure in a Devilbiss DV54D

Post by Up5times » Mon Feb 03, 2014 5:40 pm

Wulfie,

Got it. You da Man....................

Thanx 1,000,000

Wulfman...

Re: Change pressure in a Devilbiss DV54D

Post by Wulfman... » Mon Feb 03, 2014 5:50 pm

Glad I could be of assistance.


Den

.

guestuser

Re: Change pressure in a Devilbiss DV54D

Post by guestuser » Tue Jul 07, 2015 8:52 am

just would like to add that as of Today July 7 2015, the link you mentioned is still good.

Thanks for the link.

raul

Re: Change pressure in a Devilbiss DV54D

Post by raul » Sun Nov 20, 2016 12:05 am

Thank you for this. I just tried the link and it still works. I was finally able to adjust the upper limit of my Intellipap Autoadjust CPAP machine to the level prescribed by my doctor.