Very scared because I don't know what is going on

Post by fmj13 » Tue Jan 14, 2014 9:08 pm

Earlier this weekend, I woke up in the morning and my sound machine was off. Well on Monday morning, it was the case again, but the moisturizing gloves I had worn (and are really comfortable) were also off. That told me I must have turned off the sound machine and not remembered. I was so tired on Monday that I had to take a nap at work during lunch. I went to a show on Monday night (early enough to go to bed at a normal time) and I then downloaded the information from my card. It shows a flow gap that I have no memory of (the second big gap):

http://www.flickr.com/photos/109690148@N08/11957616963/

Well tonight, I looked through and according to both sleepyhead and rescan, my usage has been really low the past few days:
http://www.flickr.com/photos/109690148@N08/11957768224/
While it explains why I have been so particularly tired lately, it is terrifying to me. I don't have any memory of turning off the machine, I wake up each morning with my Swift FX on and the VPAP on. I have no idea why this is happening or what to do about it. My Swift FX fits perfectly and I find it comfortable. I have not been overly congested lately either.

Please help!

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Evora Nasal CPAP Mask with Headgear
Additional Comments: on CPAP since 2007, Bipap since July 2013

aj43

Re: Very scared because I don't know what is going on

Post by aj43 » Tue Jan 14, 2014 10:49 pm

Are you taking Ambien or other sleeping pills?

User avatar
star444
Posts: 83
Joined: Sat Dec 07, 2013 7:53 pm

Re: Very scared because I don't know what is going on

Post by star444 » Wed Jan 15, 2014 1:13 am

I have no real experience with the things you are experiencing, but hang in there because I'm sure others will be able to share their experiences with you.

_________________
Machine: Somnetics Transcend Micro Travel Machine
Mask: SleepWeaver Advance Small Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: Pressure = 10.5 - 13.5 (Wisp > Quattro FX for Her > SleepWeaver Anew > SleepWeaver Advanced)
Sleep Study AHI = 70.5

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Very scared because I don't know what is going on

Post by Julie » Wed Jan 15, 2014 4:09 am

Don't be 'terrified' - you slept without a machine for years before diagnosis (as are so many thousands still doing it), so a few bad nights or even weeks are unlikely to finish you off just yet, especially as you're getting some therapy, if not enough.

fmj13
Posts: 202
Joined: Sun Jul 21, 2013 3:42 pm

Re: Very scared because I don't know what is going on

Post by fmj13 » Wed Jan 15, 2014 5:34 am

aj43 wrote:Are you taking Ambien or other sleeping pills?
I've never taken a sleeping pill. I checked sleepyhead this morning and it continues. I go to sleep at my regular time 10:30-11 PM and wake up between 6-6:40 AM on week days, later on the weekends. Is it possible there is something wrong with my machine?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Evora Nasal CPAP Mask with Headgear
Additional Comments: on CPAP since 2007, Bipap since July 2013

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Very scared because I don't know what is going on

Post by Sheriff Buford » Wed Jan 15, 2014 6:21 am

I agree with Julie... don't be scared or terrified... it is not all that unusual. Normally, new folks will unknowenly take the mask off at night. We recogmend using tape to awaken you at night when you try to remove the mask. If you are truly turning the machine off at night... I say this because maybe the machine is doing itself.... I suspect you are reaching over and turning the machine off as you would an alarm clock. I would "rig-up" a three-sided box of some sort, and place the machine in the open side- with the back of the machine receiving un-restricted air. That way, to turn the machine off, you would have to take the machine out of the box and to successfully perform this task, you's have to be awake.

I can betcha' one thing... you'll probably will get some forum members thinking-up a gismo (lookin' at you Goof-proof ) that will work for you and your machine. I can see some of them drawing up a spec sheet... of sorts... with a material list and everything also. The homemade things some folks here can come up with constantly amazes me...

Sheriff

Sleep well
Posts: 1
Joined: Wed Jan 15, 2014 6:12 am

Re: Very scared because I don't know what is going on

Post by Sleep well » Wed Jan 15, 2014 6:31 am

I did a similar thing. Lasts time I had my data read by my supplier I had two periods when I had periods of 1 and two hours with no flow. I still had my mask on and my machine Running was running in the morning. I felt terrible I must say and could not work it out.
I use the Resmed S 9 Elite and the only thing I can think of as I get up at night that I pressed the wrong button when I got back into bed . Being the settings or control button then got up again , as I do,and pressed the on and off button as I should.

It has made me a little anxious and try to check I have exhaust air when I am up and down at night. I am fairly new to cpap so I am still in that new anxious state to some extent but getting better all the time. When I feel .it is all to much I think about how bad I was before treatment and how much more I can do now even on a less than perfect night. This keeps me at it. If I had no other conditions I would be even better.

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: Very scared because I don't know what is going on

Post by RandyJ » Wed Jan 15, 2014 6:35 am

Do you have "Auto Off" enabled? If you do, I would disable it. That way the machine cannot possibly turn off unless it loses power or you turn it off with the button. At least if you check this you can eliminate it as a cause.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
Pugsy
Posts: 65328
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Very scared because I don't know what is going on

Post by Pugsy » Wed Jan 15, 2014 7:51 am

It is also possible that there is something wrong with the machine.

With SmartStart disabled..that would eliminate large leaks maybe causing the machine to turn off all by itself.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

fmj13
Posts: 202
Joined: Sun Jul 21, 2013 3:42 pm

Re: Very scared because I don't know what is going on

Post by fmj13 » Wed Jan 15, 2014 7:09 pm

Pugsy wrote:It is also possible that there is something wrong with the machine.

With SmartStart disabled..that would eliminate large leaks maybe causing the machine to turn off all by itself.
How do you do that? I checked my instructions and it mentions to do it, but not how.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Evora Nasal CPAP Mask with Headgear
Additional Comments: on CPAP since 2007, Bipap since July 2013

User avatar
Pugsy
Posts: 65328
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Very scared because I don't know what is going on

Post by Pugsy » Wed Jan 15, 2014 7:20 pm

http://www.apneaboard.com/resmed-s9-cpap-setup

In the clinical setup menu there is a feature called SmartStart which is essentially auto on feature...turns the machine on automatically when breathing is sensed. When it is turned on then that also means the auto off feature is turned on and in the event that a large leak happens the machine will turn itself off.
I have no idea what your leak situation might have been but in this situation it's best to simply remove any auto on or off features from the equation right now.

SmartStart is right under Sleep Quality in the Options menu

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

fmj13
Posts: 202
Joined: Sun Jul 21, 2013 3:42 pm

Re: Very scared because I don't know what is going on

Post by fmj13 » Wed Jan 15, 2014 8:12 pm

Pugsy wrote:http://www.apneaboard.com/resmed-s9-cpap-setup

In the clinical setup menu there is a feature called SmartStart which is essentially auto on feature...turns the machine on automatically when breathing is sensed. When it is turned on then that also means the auto off feature is turned on and in the event that a large leak happens the machine will turn itself off.
I have no idea what your leak situation might have been but in this situation it's best to simply remove any auto on or off features from the equation right now.

SmartStart is right under Sleep Quality in the Options menu
Thanks for the instructions, however it is already set to off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Evora Nasal CPAP Mask with Headgear
Additional Comments: on CPAP since 2007, Bipap since July 2013

fmj13
Posts: 202
Joined: Sun Jul 21, 2013 3:42 pm

Re: Very scared because I don't know what is going on

Post by fmj13 » Wed Jan 15, 2014 8:13 pm

Sheriff Buford wrote:I agree with Julie... don't be scared or terrified... it is not all that unusual. Normally, new folks will unknowenly take the mask off at night. We recogmend using tape to awaken you at night when you try to remove the mask. If you are truly turning the machine off at night... I say this because maybe the machine is doing itself.... I suspect you are reaching over and turning the machine off as you would an alarm clock. I would "rig-up" a three-sided box of some sort, and place the machine in the open side- with the back of the machine receiving un-restricted air. That way, to turn the machine off, you would have to take the machine out of the box and to successfully perform this task, you's have to be awake.

I can betcha' one thing... you'll probably will get some forum members thinking-up a gismo (lookin' at you Goof-proof ) that will work for you and your machine. I can see some of them drawing up a spec sheet... of sorts... with a material list and everything also. The homemade things some folks here can come up with constantly amazes me...

Sheriff
Well, I'm not new to using CPAP and not that new to using this type of device specifically, and I haven't done this before, so I really don't understand why it is happening now.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Evora Nasal CPAP Mask with Headgear
Additional Comments: on CPAP since 2007, Bipap since July 2013

User avatar
Pugsy
Posts: 65328
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Very scared because I don't know what is going on

Post by Pugsy » Wed Jan 15, 2014 8:18 pm

fmj13 wrote:
Thanks for the instructions, however it is already set to off.
I was so hoping you wouldn't be telling me that it was already off.

Get in touch with your machine supplier tomorrow.
How long have you had this machine?
There's a problem...either with the writing of the data to the SD card or with the machine turning itself off inappropriately...both would be issues where the warranty should come into play.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
mrsbongo1
Posts: 35
Joined: Mon Mar 18, 2013 12:56 pm
Location: Southern California

Re: Very scared because I don't know what is going on

Post by mrsbongo1 » Thu Jan 16, 2014 2:54 am

My machine was doing the same thing, had it replaced in July, worked fine. Last week the new machine started turning off again on me, I called Resmed asked if there was any know issues, said no, wanted to know if I wanted to talk to tech support, said sure. Fellow there started out treating me like a moron, I had the correct answer to all his questions, then he asked about the Climate line hose, which I had just switch out. Bingo, the hose looked fine only a few months old but was the reason for my machines start and stop. I think it was the reason for the replacement of my first machine too. Well I put that hose in the trash, put on a new one and have not had any trouble since. I'm evenly thinking of turning the smart start/stop back on! Give it a try, I really don't like sleeping in my FF mask for an hours with the machine off.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead 9.2. Pressure 13. CMS 50-F Oximeter