beeping sound service requied (respronic m series auto)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mckooi
Posts: 206
Joined: Tue Jul 24, 2007 7:56 pm

beeping sound service requied (respronic m series auto)

Post by mckooi » Mon Dec 30, 2013 5:19 pm

Hi,

Problem: beeping sound and blue light on buttons blinking (keep going on).
Panel message: service required.
cpap model: respronics m series auto aflex (since 2008).

Last night the cpap was working but did not auto detect and start itself like i usually do after wearing the mask. So i pressed the start button and it works. Now, this morning i turn it off then on it did not start the blower but keep beeping as descripted above. Please help.

Went to a camp site drove on bumpy road. Cound this been the cause?
I am on pressure setting : min 16.5cm to max 18.5cm. (Without cpap to sleep i dont wake up feel well)
ahi: low 3 to 4 with cpap (without cpap ahi 65+ severe osa).

Please help and Thank you and God bless.

Mckooi

User avatar
squid13
Posts: 2300
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: beeping sound service requied (respronic m series auto)

Post by squid13 » Mon Dec 30, 2013 5:31 pm

Does it give you an error code?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

mckooi
Posts: 206
Joined: Tue Jul 24, 2007 7:56 pm

Re: beeping sound service requied (respronic m series auto)

Post by mckooi » Mon Dec 30, 2013 5:33 pm

Problem solved.
Caused: smart card wrongly inserted.
Solution: fllip over the smart card and reinsert and beeping sound stop and blinking light stop.
Reason that cause this to happen most likely foggy head and not careful or mind not sharp.
Thank you and God bless.

mckooi
Posts: 206
Joined: Tue Jul 24, 2007 7:56 pm

Re: beeping sound service requied (respronic m series auto)

Post by mckooi » Sat Dec 30, 2017 4:45 am

Hi All,

The beeping sound again with blue light blinking. I had tried the old trick by inserting the smart card. Again not working. Anyone please help.

Thank you.

Mckooi

User avatar
Julie
Posts: 20026
Joined: Tue Feb 28, 2006 12:58 pm

Re: beeping sound service requied (respronic m series auto)

Post by Julie » Sat Dec 30, 2017 5:02 am

Well it's not a new machine and they do wear out eventually. How long have you used it regularly? Most insurance co's replace after 5 yrs - is that relevant to you?

mckooi
Posts: 206
Joined: Tue Jul 24, 2007 7:56 pm

Re: beeping sound service requied (respronic m series auto)

Post by mckooi » Sat Dec 30, 2017 5:22 am

Hi Julie,

Since 2008. I need to make it work. Because financial is very limited at the moment.

Been sleeping without cpap over months. Don't feel really good anymore. Cant think well....everything is deteriorating....Please help.

Mckooi

User avatar
Julie
Posts: 20026
Joined: Tue Feb 28, 2006 12:58 pm

Re: beeping sound service requied (respronic m series auto)

Post by Julie » Sat Dec 30, 2017 5:35 am

There are people here who occasionally have machines to donate and should be along soon. I'm sorry I can't do more!

Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Re: beeping sound service requied (respronic m series auto)

Post by Woody » Sat Dec 30, 2017 6:44 am

You may want to simply disconect the power wait a minute and reconnect it. If that
doesn't work I have a S8 Vontage, an auto cpap with only about 400 hrs use on it I would
take $120 for.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software

craigclive
Posts: 8
Joined: Sun Nov 12, 2017 4:03 pm
Location: Chestnut Hill Cove, MD

Re: beeping sound service requied (respronic m series auto)

Post by craigclive » Sat Dec 30, 2017 9:13 am

I experienced the same thing. Everyone indicated that the machine was trash. Someone suggested taking the machine apart but I could not figure out how to do that. I ended up trashing mine. looking for a travel cpap.

_________________
Machine: DreamStation CPAP Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: the dreamstation machine is a DX500xxxx

User avatar
Pugsy
Posts: 64936
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: beeping sound service requied (respronic m series auto)

Post by Pugsy » Sat Dec 30, 2017 9:45 am

Do you have any money at all that you could spare towards a newer machine?

I don't have a donated machine that I could offer at this time but maybe we could find a good used machine for minimal cash outlay.

There are organizations that can help but they might also require some sort of funds plus a doctor has to be involved. Do you still have a doctor who could write a RX for the machine?

There's the Reggie White Foundation...and this one https://www.sleepapnea.org/community/cp ... p-machine/

Finally you might call the people at acbio.com who repair these machines and run your problem by them to see if they have any ideas for you to try.
It's going to cost more to fix it than the machine is worth most likely but maybe they have some ideas.
http://acbio.com/

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: beeping sound service requied (respronic m series auto)

Post by JimW159 » Sat Dec 30, 2017 11:55 am

Where do you live?
If in the US, see the last two posts in this thread: viewtopic.php?f=1&t=157767&p=1212467&hi ... r#p1212467

If Pugsy is looking, do you still agree the solution will be beneficial to the OP in regard to its functional relationship with the OP's stated machine? BTW - the machine has about 4500 hours on it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
Pugsy
Posts: 64936
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: beeping sound service requied (respronic m series auto)

Post by Pugsy » Sat Dec 30, 2017 12:39 pm

JimW159 wrote:Where do you live?
If in the US, see the last two posts in this thread: viewtopic.php?f=1&t=157767&p=1212467&hilit=+remstar#p1212467

If Pugsy is looking, do you still agree the solution will be beneficial to the OP in regard to its functional relationship with the OP's stated machine? BTW - the machine has about 4500 hours on it.
The model 550 that you have is the next newer model from the OP's original M series Auto...Pretty much the same machine with just a couple of new features and it would be an excellent option for the OP in this situation.

I used to use the M Series Auto (my very first machine back in 2009) and when the 550 came out I jumped on (when I found a cheap one) because it offered more data and the SD card thing and it was the exact same machine in terms of functionality. Win-win.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: beeping sound service requied (respronic m series auto)

Post by JimW159 » Sat Dec 30, 2017 12:45 pm

Pugsy wrote: The model 550 that you have is the next newer model from the OP's original M series Auto...Pretty much the same machine with just a couple of new features and it would be an excellent option for the OP in this situation.
I used to use the M Series Auto (my very first machine back in 2009) and when the 550 came out I jumped on (when I found a cheap one) because it offered more data and the SD card thing and it was the exact same machine in terms of functionality. Win-win.
Now we must wait to see if he responds - my track record of giving away old machines is not good, thus far.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
Pugsy
Posts: 64936
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: beeping sound service requied (respronic m series auto)

Post by Pugsy » Sat Dec 30, 2017 12:53 pm

JimW159 wrote: Now we must wait to see if he responds - my track record of giving away old machines is not good, thus far.
I hear you...you wouldn't think that it wouldn't be so difficult to give away a machine.
Mckooi is a long time forum member so maybe he will be back sooner than later.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: beeping sound service requied (respronic m series auto)

Post by TedVPAP » Sat Dec 30, 2017 1:07 pm

JimW159 wrote:
Pugsy wrote: The model 550 that you have is the next newer model from the OP's original M series Auto...Pretty much the same machine with just a couple of new features and it would be an excellent option for the OP in this situation.
I used to use the M Series Auto (my very first machine back in 2009) and when the 550 came out I jumped on (when I found a cheap one) because it offered more data and the SD card thing and it was the exact same machine in terms of functionality. Win-win.
Now we must wait to see if he responds - my track record of giving away old machines is not good, thus far.
That's funny. Maybe it's the creepy face you make when you say "freeeeeeeeee" that scares people away.