Heater light on Respironics one machine?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Islandwoman

Heater light on Respironics one machine?

Post by Islandwoman » Sun Dec 22, 2013 2:59 pm

Hello, I am just a about 30 days into CPap. I have the Respironics one which came with a regular hose. I just switched to the heated hose and a new top on the humidifier. I was told by the provider that the wavy blue lines would turn to yellow and also read that online. My lines are still blue and wonder if that means something is not working.

tetragon
Posts: 539
Joined: Wed Oct 19, 2011 4:06 am
Location: Toronto, ON
Contact:

Re: Heater light on Respironics one machine?

Post by tetragon » Sun Dec 22, 2013 6:04 pm

Did you also switch to the 80W power supply? You can tell which one you have by reading the label on the underside of the power brick.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Software: SleepyHead, loosely following HEAD

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Heater light on Respironics one machine?

Post by Pugsy » Sun Dec 22, 2013 6:34 pm

Did you also get the new power supply? The old one won't work..not enough power wattage.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Islandwoman
Posts: 319
Joined: Sun Dec 22, 2013 8:15 pm
Location: Pacific Northwest

Re: Heater light on Respironics one machine?

Post by Islandwoman » Sun Dec 22, 2013 8:27 pm

Yes, They sent me a much larger power supply. It says 80Watts on the bottom but it is connected to an extension cord. Could that be that the problem?

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Heater light on Respironics one machine?

Post by Pugsy » Sun Dec 22, 2013 8:32 pm

I don't know about the extension cord but I have the heated hose and it should turn yellow if everything is working properly.
Remove the extension cord just to check to see if that will work...bring the machine into the kitchen or somewhere to test it.
Check all the connections..especially at the power brick. Something isn't working right for some reason.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Islandwoman
Posts: 319
Joined: Sun Dec 22, 2013 8:15 pm
Location: Pacific Northwest

Re: Heater light on Respironics one machine?

Post by Islandwoman » Sun Dec 22, 2013 9:12 pm

Found the problem, took it apart and found the new internal connector wire had loosened a bit. Redid that and them manually changed hose size an heated hose to on. Yellow light now. Thank you so much at 70 this is a big life change but I now have more energy and feel very positive(and I wont have a cold nose!).