Need Advice Please

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Need Advice Please

Post by cpierc01 » Sun Dec 15, 2013 4:47 pm

I have been on BIPAP for a little over 6 months now and everything so far has been going good. Until this week my AHI is usually around 1 or2. Starting Wednesday it has went up. I am sorry I could not get the graphs to upload. Nothing has changed no new meds. Just wanted to see if this happens sometimes or should I be concerned. My average for the 6 months is 1.5 I think. Thank you for the help

Tuesday was
AHIApnea / Hypopnea Index 0.96
HypopneaHypopnea 0.75
ObstructiveObstructive Apnea 0.21
Clear AirwayClear Airway Apnea 0.00
RERAResp. Effort Related Arousal 0.43
Flow LimitFlow Limitation 0.00
VSnoreVibratory Snore 0.00
PB/CSRPeriodic Breathing 0.00%


Wed AHI
Apnea / Hypopnea Index 3.23
HypopneaHypopnea 2.59
ObstructiveObstructive Apnea 0.54
Clear AirwayClear Airway Apnea 0.11
RERAResp. Effort Related Arousal 0.32
Flow LimitFlow Limitation 0.00
VSnoreVibratory Snore 0.00
PB/CSRPeriodic Breathing 1.06%

Thur AHI
Apnea / Hypopnea Index 3.15
HypopneaHypopnea 1.99
ObstructiveObstructive Apnea 0.94
Clear AirwayClear Airway Apnea 0.21
RERAResp. Effort Related Arousal 0.52
Flow LimitFlow Limitation 0.00
VSnoreVibratory Snore 0.00
PB/CSRPeriodic Breathing 0.32%

Fri
AHIApnea / Hypopnea Index 4.03
HypopneaHypopnea 2.18
ObstructiveObstructive Apnea 1.52
Clear AirwayClear Airway Apnea 0.33
RERAResp. Effort Related Arousal 0.54
Flow LimitFlow Limitation 0.00
VSnoreVibratory Snore 0.00
PB/CSRPeriodic Breathing 0.15%

SatAHI
Apnea / Hypopnea Index 4.35
HypopneaHypopnea 3.85
ObstructiveObstructive Apnea 0.49
Clear AirwayClear Airway Apnea 0.00
RERAResp. Effort Related Arousal 0.99
Flow LimitFlow Limitation 0.00
VSnoreVibratory Snore 0.00
PB/CSRPeriodic Breathing 0.51%

User avatar
SleepWellCPAP
Posts: 370
Joined: Fri Mar 22, 2013 11:44 am
Location: Naples Florida
Contact:

Re: Need Advice Please

Post by SleepWellCPAP » Sun Dec 15, 2013 4:51 pm

According to the data you posted, it looks like your elevated AHI is mostly hypopneas. That would indicate either the reading is inaccurate or you're needing a little bit more pressure. I wonder what your leak rate is. Are you trying a new mask?
Jim Swearingen
Author of the book Sleep Well & Feel Great with CPAP, a definitive guide
For a free copy inquire with your local county librarian
CPAPtalk featured - Also available through Barnes & Noble Booksellers

cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Re: Need Advice Please

Post by cpierc01 » Sun Dec 15, 2013 4:54 pm

Leak rate is low using nasal pillows but I have noticed lately that i do feel some air quite a bit and have to reposition a lot before I fall a sleep. I was going to put on new nasal pillow tonight since the one I have been using is 6 months old. Maybe that is it since it all of a sudden went up.

cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Re: Need Advice Please

Post by cpierc01 » Sun Dec 15, 2013 4:55 pm

other info not sure how to read the leak rate

Channel Min Med 95% Max
EPAPExpiratory Pressure (cmH20)
W-Avg: 5.00 5.00 5.00 5.00 5.00
IPAPInspiratory Pressure (cmH20)
W-Avg: 11.00 11.00 11.00 11.00 11.00
PSPressure Support (cmH20)
W-Avg: 6.00 6.00 6.00 6.00 6.00
Minute Vent.Minute Ventilation (L/min)
W-Avg: 8.08 0.12 7.75 12.12 22.00
Resp. RateRespiratory Rate (breaths/min)
W-Avg: 16.43 11.37 16.20 18.80 21.60
LeaksLeak Rate (L/min)
W-Avg: 0.48 0.00 0.00 2.00 3.00
Total LeaksTotal Leak Rate (L/min)
W-Avg: 20.84 20.00 21.00 22.00 24.00
SnoreSnore (unknown)
W-Avg: 1.44 0.00 0.00 6.00 25.00
Insp TimeInspiratory Time (seconds)
W-Avg: 1.43 0.26 1.40 1.66 2.80
Exp TimeExpiratory Time (seconds)
W-Avg: 2.24 0.52 2.26 2.72 3.26
Tidal VolumeTidal Volume (ml)
W-Avg: 483.42 13.33 480.00 700.00 1180.00

User avatar
SleepWellCPAP
Posts: 370
Joined: Fri Mar 22, 2013 11:44 am
Location: Naples Florida
Contact:

Re: Need Advice Please

Post by SleepWellCPAP » Sun Dec 15, 2013 5:03 pm

Six months on a set of nasal pillows is a significant amount of time for sure and the leakage does look little high. (At IPAP 11 cm H2O your 48 L/min leak should be closer to 35 L/min) The idea of trying a new set of pillows and adding some lanolin (Lansinoh) if you still have trouble getting a good seal would be a good first start.

If that didn't work, if you feel comfortable adjusting your pressure, you might try raising your EPAP from 5 to 6.

Hope that helps.
Jim Swearingen
Author of the book Sleep Well & Feel Great with CPAP, a definitive guide
For a free copy inquire with your local county librarian
CPAPtalk featured - Also available through Barnes & Noble Booksellers

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need Advice Please

Post by Pugsy » Sun Dec 15, 2013 5:10 pm

SleepWellCPAP wrote:leakage does look little high. (At IPAP 11 cm H2O your 48 L/min leak should be closer to 35 L/min)
Where are you seeing Leak being 48 L/min? I don't see it.
I see leak 0.48 but that is the minimum excess leak number (or SH's close approximation of excess leak)...the worst he ever got was 24.00 max under the Total leak column.
I am betting the leak line is so straight and flat I would be pea green with envy if I could see the actual graph.
cpierc01 wrote:LeaksLeak Rate (L/min)
W-Avg: 0.48 0.00 0.00 2.00 3.00
Total LeaksTotal Leak Rate (L/min)
W-Avg: 20.84 20.00 21.00 22.00 24.00

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
SleepWellCPAP
Posts: 370
Joined: Fri Mar 22, 2013 11:44 am
Location: Naples Florida
Contact:

Re: Need Advice Please

Post by SleepWellCPAP » Sun Dec 15, 2013 5:21 pm

You're right Pugsy. Puzzling indeed, total leaks don't appear to have changed.

I'd still recommend trying a new set of pillows though and see how that goes before I would recommend a pressure increase.
Jim Swearingen
Author of the book Sleep Well & Feel Great with CPAP, a definitive guide
For a free copy inquire with your local county librarian
CPAPtalk featured - Also available through Barnes & Noble Booksellers

cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Re: Need Advice Please

Post by cpierc01 » Sun Dec 15, 2013 6:17 pm

Thank You for the responses and yes normally the leak line is very low. But I have been having air leak around the seal that I have to keep adjusting. I will change the pillows out tonight and see ow it goes. I am puzzled by this if I needed more pressure strange that it went changed in one day.

cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Re: Need Advice Please

Post by cpierc01 » Mon Dec 16, 2013 8:05 pm

Changed out the nasal pillows last night and AHI went back down to 1.0 from 4.3 the night before. Thanks everyone and I will keep a check on it.

Guest

Re: Need Advice Please

Post by Guest » Tue Dec 17, 2013 3:04 pm

i would make sure that your mask has a proper fit as well, lot of times patients do not realize the fit is off.

changing your settings you should consult with your sleep provider before doing so.

cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Re: Need Advice Please

Post by cpierc01 » Tue Dec 17, 2013 6:50 pm

Thanks just an update second night AHI was 0.5 so hopefully it was that I need to change out the Nasal Pillows. I am new to this so I am learning.

User avatar
SleepWellCPAP
Posts: 370
Joined: Fri Mar 22, 2013 11:44 am
Location: Naples Florida
Contact:

Re: Need Advice Please

Post by SleepWellCPAP » Tue Dec 17, 2013 6:56 pm

Glad to hear it cpierc01, good for you!