Machine not registering full sleep time?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
GayleM
Posts: 58
Joined: Sat Feb 02, 2013 11:47 pm
Location: Texas

Machine not registering full sleep time?

Post by GayleM » Fri Oct 25, 2013 5:12 pm

My husband got his machine (Resmed S9 Elite) about 6 weeks ago. He's working nights right now. He went to bed about 8 this morning, up at 11:15 for about 15 minutes then back to bed til about 3:15 this afternoon. I checked the time on his machine as soon as he got up and it said 1.0 hours. I just loaded his card to both ResScan & SleepyHead and they both show only 1 hour. Why would it only show 1 hour, I don't understand!!

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Machine not registering full sleep time?

Post by JDS74 » Fri Oct 25, 2013 5:35 pm

'cause the machine resets its counters at noon.
It should show additional hours on the previous day's data.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
GayleM
Posts: 58
Joined: Sat Feb 02, 2013 11:47 pm
Location: Texas

Re: Machine not registering full sleep time?

Post by GayleM » Fri Oct 25, 2013 5:40 pm

That makes sense that it would reset at noon, but wouldn't it have 3.25 hours then, since he woke up at 3:15?

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Machine not registering full sleep time?

Post by JDS74 » Fri Oct 25, 2013 6:37 pm

Noon for the machine may not be noon for you.
On PR System One machines, you have to tell the software what time zone you are in with respect to the time zone the machine thinks it is in.
I presume something similar applies to ResMed machines. Just a guess.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
GayleM
Posts: 58
Joined: Sat Feb 02, 2013 11:47 pm
Location: Texas

Re: Machine not registering full sleep time?

Post by GayleM » Fri Oct 25, 2013 7:01 pm

Ok, thanks for the info!! Will have to figure out how to set the time zone.

User avatar
Pugsy
Posts: 65262
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Machine not registering full sleep time?

Post by Pugsy » Fri Oct 25, 2013 7:39 pm

SleepyHead has a time zone setting. ResScan doesn't that I know of so if the time is off in ResScan then that's the machine clock.
The S9 machines have a clock that has to be set to local time. It's possible that the machine's clock is off...so check it too.
The clock is available only in the clinical set up menu though. Have you seen it to verify the time that the machine is set to is correct?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
GayleM
Posts: 58
Joined: Sat Feb 02, 2013 11:47 pm
Location: Texas

Re: Machine not registering full sleep time?

Post by GayleM » Fri Oct 25, 2013 10:27 pm

The clock is the culprit! I just checked it and it said 21:05 and it is 23:15. The 2 hours difference plus the 1 hour it showed adds up to the 3 hours after noon that I would have expected. I also adjusted the MIN Pressure a bit (it was set to 5) to 6. His MAX is 8. He has complained about "not getting enough air" so he keeps pushing the Mask Fit, which I've read is at 10 or Max Pressure, whichever is greater. I'm thinking after we take his SD card on Monday for his last compliance check, that I will change his Max to 10 and maybe his MIN to 8 so he will be more comfortable. Although, I can't imagine being comfortable at that pressure after me being at 21-16 for almost a year now!!!

Thank you so much for helping me out!! This group is awesome!!