trouble with sleepyhead

Post by theoldman22 » Thu Oct 17, 2013 8:25 pm

Picture below is (I hope) of what I get trying to show downloaded information using Sleepyhead v0.9.3, which was downloaded about a month ago. What am I doing wrong? Computer is using Windows XP Pro.

theoldman22

[img]http://s721.photobucket.com/user/2Paul2 ... 1208257112[/img]

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed S9 VPAP Auto

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: trouble with sleepyhead

Post by robysue » Thu Oct 17, 2013 9:18 pm

theoldman22,

I could not see the image in the post. Copy, pasting, and editing the provided url gives me the following image. Is this the correct image? If this is the right image, what's the question? Is it the question about why the fonts are so messed up???

Image

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: trouble with sleepyhead

Post by jedimark » Thu Oct 17, 2013 9:23 pm

Your system fonts are way too big...

What screen resolution are you running on?

I'm glad you brought this up, because I need to address this issue for people with not so good eyesight and tiny screens.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
Tech Guy
Posts: 64
Joined: Fri Sep 13, 2013 5:32 pm

Re: trouble with sleepyhead

Post by Tech Guy » Thu Oct 17, 2013 11:19 pm

Open Sleepy head software. Click File / Preferences then Appearance. You can adjust the font size to 10 which is the default. That should fix it .
PR System One & Humidifier

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: trouble with sleepyhead

Post by avi123 » Thu Oct 17, 2013 11:56 pm

Oldman from NM, since you have a Resmed machine why don't you use ResScan? ResScan helped me achieve an envious treatment.
See my last year's Stats:

Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
theoldman22
Posts: 38
Joined: Wed Oct 10, 2012 8:39 pm
Location: New Mexico

Re: trouble with sleepyhead

Post by theoldman22 » Fri Oct 18, 2013 3:13 pm

thanks to all who replied. Appreciate.

Yes, the URL in my post was of the image I was trying to send. The instructions I used for the attempt were from one of Pugsy's post:
Once the screen shot is created save it in jpg format.
Upload the image to a host site. I use Photobucket it is free, there are others.
Once the image is uploaded then copy the ENTIRE IMG address. Be sure to include the opening and closing IMG in brackets. Paste that copied address into a post here.
If you want to use Photobucket they provide a box with the various url addresses and you just choose the last one in the box.
Copy/paste that last line url address into the body of your post.
Use the preview button. If you can't see the image try again because if you can't see it we can't.
I tried many times using these instructions but never got the image--just the URL.

Later today I will reduce thw font size as suggested and see if I get a usable versuin of SleepyHead.

Also, a ine person sent me the link to download ResCan--will try that also.

theoldman22

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed S9 VPAP Auto

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: trouble with sleepyhead

Post by DoriC » Fri Oct 18, 2013 5:31 pm

robysue wrote:theoldman22,

I could not see the image in the post. Copy, pasting, and editing the provided url gives me the following image. Is this the correct image? If this is the right image, what's the question? Is it the question about why the fonts are so messed up???

Image
Preferences, Appearance, Uncheck Use Pixmap Caching, that should also help.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
theoldman22
Posts: 38
Joined: Wed Oct 10, 2012 8:39 pm
Location: New Mexico

Re: trouble with sleepyhead

Post by theoldman22 » Fri Oct 18, 2013 11:58 pm

Sreen resolution is 1024 x 768 pixels. Font size in SleepyHead is 10. The image (hope it comes thru) is just one page of what was available. There is a preview pane above where I am typing that shows the image. I am hoping to get orderly pages with graphs of several variables for the night.

Image

the old man22

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed S9 VPAP Auto

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: trouble with sleepyhead

Post by jedimark » Sat Oct 19, 2013 12:30 am

theoldman22 wrote:Sreen resolution is 1024 x 768 pixels. Font size in SleepyHead is 10. The image (hope it comes thru) is just one page of what was available. There is a preview pane above where I am typing that shows the image. I am hoping to get orderly pages with graphs of several variables for the night.
That looks really weird. It could be caused by a windows Display-Per-Inch accessibility setting that's out of whack in your control panel..

Unfortunately I don't have access to a WinXP system anymore, I think I gave my last notebook with WinXP away, and my XP VM on the mac got upgraded to Win8. :/

I'll have a play with the DPI settings on my win7 notebook and see if I can replicate it.

If your seeing this on first install, there may be a bug in the Qt cross platform libraries with setting font point size.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: trouble with sleepyhead

Post by jedimark » Sat Oct 19, 2013 12:37 am

Here is some information on the Windows DPI settings I was talking about:

http://www.microsoft.com/resources/docu ... x?mfr=true

Also, there is a "cleartype" font option that may make things clearer afterwards.

http://support.microsoft.com/kb/306527

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: trouble with sleepyhead

Post by DoriC » Sat Oct 19, 2013 8:53 am

DoriC wrote:
robysue wrote:theoldman22,

Image
Preferences, Appearance, Uncheck Use Pixmap Caching, that should also help.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
theoldman22
Posts: 38
Joined: Wed Oct 10, 2012 8:39 pm
Location: New Mexico

Re: trouble with sleepyhead

Post by theoldman22 » Sat Oct 19, 2013 4:21 pm

Thank you. Did the following:

1. Changed computer font from Large to Normal.
2. Turned on Clear Type.
3. Unchecked PixMap Cacbing

Enclosed is an image showing the results--thank you very much.

Image

Still have download showing Mode as BiPap--in Profile changed to APAP. The ResMed VPAP Auto machine is set to Auto with
IP max 16, EP min 5, PS 3.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed S9 VPAP Auto

User avatar
Pugsy
Posts: 65270
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: trouble with sleepyhead

Post by Pugsy » Sat Oct 19, 2013 7:43 pm

theoldman22 wrote:Still have download showing Mode as BiPap--in Profile changed to APAP. The ResMed VPAP Auto machine is set to Auto with
IP max 16, EP min 5, PS 3.
There is no APAP mode on the S9 VPAP
There's cpap mode, fixed bilevel mode which is called "S" and auto bilevel mode.
There is a bug in SleepyHead with the bilevel pressure machines (both the S9 VPAP and the PR S1 BiPap) not showing the correct PS. It's being worked on now.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: trouble with sleepyhead

Post by jedimark » Sat Oct 19, 2013 9:14 pm

Pugsy wrote:
theoldman22 wrote:Still have download showing Mode as BiPap--in Profile changed to APAP. The ResMed VPAP Auto machine is set to Auto with
IP max 16, EP min 5, PS 3.
There is no APAP mode on the S9 VPAP
There's cpap mode, fixed bilevel mode which is called "S" and auto bilevel mode.
There is a bug in SleepyHead with the bilevel pressure machines (both the S9 VPAP and the PR S1 BiPap) not showing the correct PS. It's being worked on now.
Yup, I only just fixed the VPAP auto stuff.. It will be in the next release.
I've still got more BiPAP/ASV stuff to do... just finding heaps of little easy things to clean up first.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: trouble with sleepyhead

Post by archangle » Sat Oct 19, 2013 11:37 pm

BTW, VirtualBox is great for running multiple versions of Windoze, linux, etc. on one machine.

It's not exactly the same as a real PC, but you can probably see a lot of problems with it. Display problems may be a little iffy, but it's fun to play with anyway.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.