Got my new S9 AutoSet!!

Post by islandboy5150 » Sun Oct 06, 2013 8:54 am

After 8 years of running on a brick, a FP 604 sleep style institutional looking over rated leaf blower, I finally got an S9 AutoSet! Love how quiet the motor is. And the menu screens. Awesome! The look overall is great! The hose feels a bit cheesy, though. The FP hose is way better. You can tell resmed absolutely took the cheapest way out with the hose. It couldn't be any cheaper even if it was made in China. Also, noticed right off that it sounds a bit like dearth vader breathing. It's actually pretty loud in that regard. But, that's the only aspect of the machine itself that I'm not happy with. Wish it didn't sound like air blowing through a hollow plastic box, even without the humidifier.
I immediately went into the clinicians menu to snoop around after my first night using it to discover they set it up wrong. And they set it up right in front of me, asking me all the questions, then dialing it in. For instance, they asked if I used or wanted ramp. I said heck no, don't want any ramp, it's just a pain in the butt. Ramp was set to on, 45 minutes, starting at 4! But, they did give me an auto, when I really didn't need one, and because I requested one, and with no push back at all. But the low pressure they had it start at was starving me of air, just like during the sleep study. If I couldn't tolerate a low start pressure of 4 during the study and didn't get to sleep until 2am after I begged the tech during the study to crank up the pressure, then why set my new AutoSet to start at 4 with a 45 minute ramp to boot? And after they asked me while setting it up? It takes a minimum of 14 to keep my airway open.
But, I'm glad I checked the setup. Over all very happy with the machine. Now that it's set correctly. Now just have to figure a way to reduce the dearth vader berating noise.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PAP-CAP for Headgear, Hose Cozy and Hose Boss

SleepyToo2
Posts: 1005
Joined: Sun Sep 11, 2011 7:55 am
Location: North of Philadelphia, PA

Re: Got my new S9 AutoSet!!

Post by SleepyToo2 » Sun Oct 06, 2013 9:09 am

Two things:

Firstly, have you changed your mask along with the machine? I have found that some masks course different noises than others, so this may be one of the reasons for the.Darth Vader noise?

Secondly, I find that using my machine without the humidifier makes it seem very noisy. I have a different machine to you, but my thinking is that the humidifier may actually be like a silencer on the exhaust pipe on a car.

Good luck with quieting the beast!

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Use SleepyHead software.
Not a medical professional - just a patient who has done a lot of reading

User avatar
islandboy5150
Posts: 301
Joined: Fri Oct 15, 2010 9:14 pm
Location: Temecula, California
Contact:

Re: Got my new S9 AutoSet!!

Post by islandboy5150 » Sun Oct 06, 2013 9:20 am

SleepyToo2 wrote:Two things:

Firstly, have you changed your mask along with the machine? I have found that some masks course different noises than others, so this may be one of the reasons for the.Darth Vader noise?

Secondly, I find that using my machine without the humidifier makes it seem very noisy. I have a different machine to you, but my thinking is that the humidifier may actually be like a silencer on the exhaust pipe on a car.

Good luck with quieting the beast!
Thanks SleepyToo2. Yea, got a new mask at the same time. Hollow breathing noise is coming from the machine. Both with and without the humidifier. I leaned over the machine and took a few breaths. It sounds like air blowing through a hollow plastic box. Pretty loud, at least to me.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PAP-CAP for Headgear, Hose Cozy and Hose Boss

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Got my new S9 AutoSet!!

Post by Sheriff Buford » Sun Oct 06, 2013 1:19 pm

Islandboy: where can I get a look at a FP hose? I've used the both the Resmed and generic hose.

Sheriff

User avatar
islandboy5150
Posts: 301
Joined: Fri Oct 15, 2010 9:14 pm
Location: Temecula, California
Contact:

Re: Got my new S9 AutoSet!!

Post by islandboy5150 » Sun Oct 06, 2013 1:49 pm

Sheriff Buford wrote:Islandboy: where can I get a look at a FP hose? I've used the both the Resmed and generic hose.

Sheriff
Not sure, Sheriff. Unless take a picture of one and post it maybe. i've never seen one pictured online, but thats been a while since I looked too.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PAP-CAP for Headgear, Hose Cozy and Hose Boss

User avatar
Pugsy
Posts: 65266
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Got my new S9 AutoSet!!

Post by Pugsy » Sun Oct 06, 2013 2:32 pm

Islandboy....are you talking about the F & P heated hose when compared to the S9 ClimateLine heated hose?
Or are you using the slimline non heated hose with the S9?
Did you use a regular non heated hose with your old machine?
I have not ever seen a non heated F & P hose to compare it with non heated hose for the other machines.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Got my new S9 AutoSet!!

Post by avi123 » Sun Oct 06, 2013 2:55 pm

@
@
See here how I set the pressures on my S9 APAP:

If my prescribed therapeutic pressure on CPAP was 13 cm then I would start the APAP pressures on 13 cm max and 10.0 cm min. Next, I would spend a night with this setting and get a pressure graph on ResScan. The graph would show the pressure curve plus the two limit lines. Next, I would start adjusting the set-up on the APAP so that the pressure curve only "kisses" those limit lines. The reasons: I don’t want the machine to suddenly boost the pressure needlessly and give me emergent central apneas and cause me to have CompSA. Also, cause more leaks b/c of the higher pressure. As to the low pressure limit I set it so to have the machine respond more quickly to events.

Example

Image

Wishing you luck on your new machine and within a year you should be showing better STATS Data than my:

Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
islandboy5150
Posts: 301
Joined: Fri Oct 15, 2010 9:14 pm
Location: Temecula, California
Contact:

Re: Got my new S9 AutoSet!!

Post by islandboy5150 » Sun Oct 06, 2013 3:05 pm

Heated S9 hose, compared to the FP heated hose. The FP is a little bigger, heavier, but a lot more flexible, pliable, and robust. FP Just seems like a lot more durable hose. Plus, the S9 hose has soft rubber ends, and I find it's harder to connect and much harder to disconnect from my mask. I feel like I'm gonna tear the hose apart at the point where the connector and hose come together. I have to really be careful to grab the rubber connector and only pull on that, while making sure to not put any tension on the hose part. It's a bit of a pita compared to the FP hose ends. The FP ends are hard plastic, that I could grab and twist off the elbow on the mask without fear of damaging the hose. The FP hose connector is easy on, easy off friction fit. The S9 heated hose has soft pliable rubber ends, and you have to be careful where you grab it and you have to kind of wiggle and pull, which if you are grabbing on the hose, will surely rip it.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PAP-CAP for Headgear, Hose Cozy and Hose Boss

User avatar
okitoki
Posts: 88
Joined: Mon May 24, 2010 1:10 am
Location: Perth, Australia

Re: Got my new S9 AutoSet!!

Post by okitoki » Mon Oct 07, 2013 1:15 am

hmmm... strange with the "breathing" noise, as I have also just started using the S9 Autoset... Im still using my old F&P 431 Full face mask and I find that the machine is seriously dead silent when Im using it.. My start off pressure is at 6, and there are times that I had to pull my mask off my face just to feel the big gush of air and noise to confirm that it is actually on... no joking.. it is that quiet for me... even the breathing feels natural for me without the sensation of air being forced in.....

Coming from the S8, this machine is awesome...

_________________
MaskHumidifier

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Got my new S9 AutoSet!!

Post by Sheriff Buford » Mon Oct 07, 2013 5:00 am

okitoki wrote:hmmm... strange with the "breathing" noise, as I have also just started using the S9 Autoset... Im still using my old F&P 431 Full face mask and I find that the machine is seriously dead silent when Im using it.. My start off pressure is at 6, and there are times that I had to pull my mask off my face just to feel the big gush of air and noise to confirm that it is actually on... no joking.. it is that quiet for me... even the breathing feels natural for me without the sensation of air being forced in.....

Coming from the S8, this machine is awesome...
I put my hand over the vent to feel the air (vent)...... to see if the machine is working. Also, in cpap mode, the S9 EPR defaults to medium. That gush of air (on EPR mode) gives me that "Darth Vader" sound. I tried to turn the EPR off (in cpap mode), but the machine wouldn't let me.

Sheriff

knightlite
Posts: 152
Joined: Sun Dec 30, 2012 4:21 pm
Contact:

Re: Got my new S9 AutoSet!!

Post by knightlite » Mon Oct 07, 2013 5:14 am

Try a pad a cheek liner to see if that helps with the vader sound , They are great with this mask . I find that the humidifier helps as well as the hose cozy .

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: apap ds560 sleepyhead software/with patch, resmed hose cozy, pressure set 10 toa 14.5 , aflex of 3 ,ramp off, not using heated hose
now using mirage quattro ffm with pad a cheek liner , hose hanger on head board with 18 inch short hose /swivel to 6 foot hose-- DS560 apap set 10-14.5

User avatar
islandboy5150
Posts: 301
Joined: Fri Oct 15, 2010 9:14 pm
Location: Temecula, California
Contact:

Re: Got my new S9 AutoSet!!

Post by islandboy5150 » Mon Oct 07, 2013 11:27 am

knightlite wrote:Try a pad a cheek liner to see if that helps with the vader sound , They are great with this mask . I find that the humidifier helps as well as the hose cozy .
I do, and have for years. Karen even special made for me a liner out of fleece instead of the lint free type of material. I use a hose cozy, hangers, and a pap-cap. Thats all been dialed in for years now.
The machine works good though, AHI is under 2 most of the time. On the FP 604 I couldnt hear the air going through the machine at all. On the S9, I can. And the tone of the sound changes with my breathing. Maybe its just me and the machine is working as designed.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PAP-CAP for Headgear, Hose Cozy and Hose Boss