DevilBliss DV54 AutoAdjust CPAP -Help
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Feeling a bit worse since increasing the lower pressure. But i'll give it another 3 or 4 days and see if that make a difference.
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Hi alexchetty,alexchetty wrote:Feeling a bit worse since increasing the lower pressure. But i'll give it another 3 or 4 days and see if that make a difference.
Are you under a doctor's care or are you doing this on your own?
Also, is there any way you could buy the smart link module so you could access sleepyhead software?
49er
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Use SleepyHead |
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
on my own.
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
07October2013
95th Percentile Pressure 13.0
90th Percentile Pressure 12.0
AHI 4.5
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 2.75 (Is this a concern?)
EPI 31 (Is this a concern? Should I be using a full face mask?)
--Pressure at low 6.5 - high 20
--Note: AHI has decreased slightly.
--going to give it another couple of days.. maybe my body is adjusting to the change.
--Still too early (6:48am Cape Town, South Africa) to tell whether i'm feeling better, worse or the same.
95th Percentile Pressure 13.0
90th Percentile Pressure 12.0
AHI 4.5
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 2.75 (Is this a concern?)
EPI 31 (Is this a concern? Should I be using a full face mask?)
--Pressure at low 6.5 - high 20
--Note: AHI has decreased slightly.
--going to give it another couple of days.. maybe my body is adjusting to the change.
--Still too early (6:48am Cape Town, South Africa) to tell whether i'm feeling better, worse or the same.
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Great news alexchetty that your AHI is down. Definitely give it a few more days for your body to adjust.alexchetty wrote:07October2013
95th Percentile Pressure 13.0
90th Percentile Pressure 12.0
AHI 4.5
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 2.75 (Is this a concern?)
EPI 31 (Is this a concern? Should I be using a full face mask?)
--Pressure at low 6.5 - high 20
--Note: AHI has decreased slightly.
--going to give it another couple of days.. maybe my body is adjusting to the change.
--Still too early (6:48am Cape Town, South Africa) to tell whether i'm feeling better, worse or the same.
Not concerned NRI has increased for now. However, the fact that it has from .5 which consists mostly of central leads me to believe that you do want to be conservative increasing your pressure particularly since your AHI is trending downwards.
Regarding the EPI 31, yes, you do either need to switch to a full face mask or if you're happy with your nasal mask, you could try wearing a chin strap to see if that prevents mouth breathing. Another option may be using tape to keep your mouth shut. Not everyone is comfortable with that but I wanted to throw it out there.
49er
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Use SleepyHead |
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Thanks for the reply 49er.
I will monitor my progress and post results daily to keep this forum up to date.
I will monitor my progress and post results daily to keep this forum up to date.
Re: DevilBliss DV54 AutoAdjust CPAP -Help
You're very welcome. Kudos you for your diligence in monitoring your therapy.alexchetty wrote:Thanks for the reply 49er.
I will monitor my progress and post results daily to keep this forum up to date.
49er
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Use SleepyHead |
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
08October2013
95th Percentile Pressure 10.0
90th Percentile Pressure 9.5
AHI 5.0
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 0.5
EPI 31
--Pressure at low 6.5 - high 20
--7:06am Cape Town, South Africa - Feeling better than yesterday this morning.
95th Percentile Pressure 10.0
90th Percentile Pressure 9.5
AHI 5.0
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 0.5
EPI 31
--Pressure at low 6.5 - high 20
--7:06am Cape Town, South Africa - Feeling better than yesterday this morning.
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Great you're feeling better alexchetty. Maybe a few more days at this pressure to see what the AHI does. But more importantly, to see if you continue to feel better.alexchetty wrote:08October2013
95th Percentile Pressure 10.0
90th Percentile Pressure 9.5
AHI 5.0
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 0.5
EPI 31
--Pressure at low 6.5 - high 20
--7:06am Cape Town, South Africa - Feeling better than yesterday this morning.
Madbear had a good point that if you were using an S9, you might be seeing a lower AHI. So that is another reason it might be just as important to go by how you feel vs. your AHI.
Have you made a decision on how you're going to deal with the high EPI?
49er
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Use SleepyHead |
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
I've got a both a full face mask and a nasal mask. will wait a couple more days to see how i'm feeling and if the AHI drops further or at lease stabilises before switching to the full face mask.
-
alexchetty
- Posts: 94
- Joined: Wed Oct 02, 2013 11:10 pm
- Location: Cape Town
Re: DevilBliss DV54 AutoAdjust CPAP -Help
09October2013
95th Percentile Pressure 11.5
90th Percentile Pressure 10.5
AHI 4.0
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 1.25
EPI 24
--Pressure at low 6.5 - high 20
--Note: AHI has decreased
--Going to stay on this pressure for another day, then increase. (Do you think adjusting the lower pressure from 6.5 to 8 is too a big jump?)
95th Percentile Pressure 11.5
90th Percentile Pressure 10.5
AHI 4.0
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 1.25
EPI 24
--Pressure at low 6.5 - high 20
--Note: AHI has decreased
--Going to stay on this pressure for another day, then increase. (Do you think adjusting the lower pressure from 6.5 to 8 is too a big jump?)
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Hi alexchetty,alexchetty wrote:09October2013
95th Percentile Pressure 11.5
90th Percentile Pressure 10.5
AHI 4.0
Pressure Plateau Time 0%
High Leak Flow Time 0%
NRI 1.25
EPI 24
--Pressure at low 6.5 - high 20
--Note: AHI has decreased
--Going to stay on this pressure for another day, then increase. (Do you think adjusting the lower pressure from 6.5 to 8 is too a big jump?)
How do you feel? Unless you are feeling worse, since the AHI had decreased, I would wait a few more days to see if this is going to continue.
49er
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Use SleepyHead |
-
Mr Concerned
- Posts: 43
- Joined: Wed Oct 09, 2013 11:43 pm
Re: DevilBliss DV54 AutoAdjust CPAP -Help
49er can you write me a prescription for Cetirizine? I have this really bad rash that won't go away.
Re: DevilBliss DV54 AutoAdjust CPAP -Help
Mr. Concerned,Mr Concerned wrote:49er can you write me a prescription for Cetirizine? I have this really bad rash that won't go away.
Instead of acting like a troll, why don't you tell us who you are so we understand the context of why you are so concerned about forums like this one?
49er
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Use SleepyHead |
-
Mr Concerned
- Posts: 43
- Joined: Wed Oct 09, 2013 11:43 pm
Re: DevilBliss DV54 AutoAdjust CPAP -Help
I am an internet forum user. Just like you. I don't give out my real name. Just like you. I could really be anyone.... much like you. I really don't see why you are so concerned about my identity. When I read through this forum I was not immediately alarmed by all of the anonymous forum users. I assumed they were people from all over the world that had devised clever pseudonyms to use for identification purposes.49er wrote:Mr. Concerned,Mr Concerned wrote:49er can you write me a prescription for Cetirizine? I have this really bad rash that won't go away.
Instead of acting like a troll, why don't you tell us who you are so we understand the context of why you are so concerned about forums like this one?
49er