need help interpreting data!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
toryg51
Posts: 15
Joined: Mon Sep 09, 2013 8:35 am
Location: Upstate NY

need help interpreting data!

Post by toryg51 » Mon Sep 23, 2013 4:05 pm

Hi, everybody!
Inordinately proud that I was able to actually figure out how to get into my machine stats!!! Unfortunately, that was about the end of it for being able to post it here. I was able to copy this:
ResMed
S9 AutoSet
Mode: CPAP 8cmH2O
Date Sleep Wake Hours
9/22/2013 18:58 04:16 09:18:04
AHIApnea / Hypopnea Index 3.44
HypopneaHypopnea 0.11
ApneaUnspecified Apnea 0.11
ObstructiveObstructive Apnea 1.18
Clear AirwayClear Airway Apnea 2.04

Event Breakdown


Statistics
Channel Min Med 95% Max
PressurePressure (cmH20)
W-Avg: 7.98 5.60 8.00 8.00 8.00
EPAPExpiratory Pressure (cmH20)
W-Avg: 5.98 4.00 6.00 6.00 6.00
Minute Vent.Minute Ventilation (L/min)
W-Avg: 5.81 0.00 5.50 8.62 16.84
Resp. RateRespiratory Rate (breaths/min)
W-Avg: 15.21 0.00 15.20 20.00 30.00
Flow LimitFlow Limit Graph (0-1)
W-Avg: 0.00 0.00 0.00 0.01 0.27
LeaksLeak Rate (L/min)
W-Avg: 8.68 0.00 4.80 30.00 48.00
SnoreSnore (unknown)
W-Avg: 0.04 0.00 0.04 0.08 0.18
Insp TimeInspiratory Time (seconds)
W-Avg: 1.97 0.20 1.80 3.16 6.24
Exp TimeExpiratory Time (seconds)
W-Avg: 2.07 0.04 2.06 3.16 4.70
Tidal VolumeTidal Volume (ml)
W-Avg: 393.80 0.00 360.00 680.00 1400.00

As you may remember from a previous post, it's the leaking I'm worrying about. Can someone interpret this for me? These are last night's stats. My AHI's the previous three nights (my first three nights on CPAP) were 5.09, 3.80, 2.36. which I don't think look too bad. And the leak rate has been about the same, sometimes a little less, sometimes a little more. So should I be sweating the leaks?

THANK YOU EVERYBODY - can't imagine doing this without your support!!!

_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also use a dental night guard. Pressure set at 8.

User avatar
Pugsy
Posts: 65307
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: need help interpreting data!

Post by Pugsy » Mon Sep 23, 2013 4:16 pm

The S9 machine reports excess leak only. ResMed says that they can compensate for leaks up to 24 L/min.
Look at the leak graph line...how much time is spent above 24 L/min?

In the SleepyHead tutorial I talked about how sometimes the numbers look worse than things really are so when numbers look iffy then I suggest that people evaluate the leak line itself.
viewtopic/t88983/Pugsys-PointersSleepyH ... nding.html
You might give it a glance to see if it helps you understand things a bit more easily.
Pay special attention to the leak explanation and examples.

Now any leak that wakes a person up needs to be dealt with because it disturbs sleep.
Small short lived leaks into large leak territory may not be nearly as scary as the numbers first indicate they might be.
5 to 15 minutes in large leak territory out of 7 or 8 hours of sleep and a person sleeps right through it, is not the end of the world and doesn't negate the otherwise less than large leak territory time.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

toryg51
Posts: 15
Joined: Mon Sep 09, 2013 8:35 am
Location: Upstate NY

Re: need help interpreting data!

Post by toryg51 » Mon Sep 23, 2013 4:33 pm

Thanks, Pugsy,
Jeepers, I wish I could figure out how to copy and paste that graph so you could see it. It seems like I'm in large leak territory a fair amount of time, but again, if my apneas are relatively low (which they seem to be - am I right?) maybe I should just not worry about it and keep an eye on things.

_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also use a dental night guard. Pressure set at 8.

User avatar
Pugsy
Posts: 65307
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: need help interpreting data!

Post by Pugsy » Mon Sep 23, 2013 5:05 pm

Screen shots of the detailed reports is the easiest way for us to see them
How to post them here on the forum is discussed in this thread.
viewtopic.php?f=1&t=81072&p=737779#p737779

If you are spending prolonged periods of time well above the 24 L/min line then the low AHI could be from the machine not being able to sense and record properly. We need to be able to trust the data and if large leaks are going on for prolonged periods of time then we can't trust the data.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.