S9 Auto Off doesn't seem to be working

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
JeffL
Posts: 507
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

S9 Auto Off doesn't seem to be working

Post by JeffL » Thu Sep 19, 2013 12:50 pm

I usually don't use the auto off, but this morning, I removed my mask, and waited for my S9 to turn off. It just kept blowing air.
The auto on works fine.
This is a new machine that I was given to replace one that became noisy. The auto off on the original machine worked.
I played with it this morning, and found that instead of shutting off, I get a message about a High Leak.
There isn't a setting for having it start automatically, but not stop automatically. There's just the SmartStart on/off option.
Any ideas?

_________________
MachineMask

User avatar
Pugsy
Posts: 65390
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: S9 Auto Off doesn't seem to be working

Post by Pugsy » Thu Sep 19, 2013 1:00 pm

Try setting Smart Start to "ON" and Leak Alert to "Off" and see if it works the way you want it to.
I don't have my S9 to verify but I am pretty sure that's what I used to get the S9 to come on when I took a couple of breaths and turn off in about a minute if I removed the mask.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
JeffL
Posts: 507
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: S9 Auto Off doesn't seem to be working

Post by JeffL » Thu Sep 19, 2013 2:34 pm

I've gone through the clinical manual, and can't find any reference to Leak Alert or anything like that.

_________________
MachineMask

User avatar
bosstow
Posts: 67
Joined: Fri May 04, 2012 6:17 pm

Re: S9 Auto Off doesn't seem to be working

Post by bosstow » Thu Sep 19, 2013 2:41 pm

JeffL wrote:I've gone through the clinical manual, and can't find any reference to Leak Alert or anything like that.
No Leak Alert setting unless you have an S9 VPAP model.

On my VPAP Auto, if you turn on Leak Alert, it disables SmartStart. In other words you can't have both SmartStart (which includes the automatic stop) and Leak Alert turned on at the same time.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Max IPAP 25, Min EPAP 10, PS 5, ramp off. Climateline MAX tubing, CMS-50F Pulse Oximeter, ResScan & SleepyHead
Also have a Resmed S9 Autoset as a backup/travel machine

User avatar
Pugsy
Posts: 65390
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: S9 Auto Off doesn't seem to be working

Post by Pugsy » Thu Sep 19, 2013 2:47 pm

Sorry, I didn't click on the link to see if you had the VPAP. I just assumed that you had the leak alert feature since you mentioned something about a high leak message. My bad.

With the S9 machine SmartStart is the only way to have Auto On and Auto Off that I know of.
It's either on or off. When I let it do the Auto Off thing it would turn off probably in about a minute. I no longer have the S9 VPAP so I can't test it.

If SmartStart set to On...allows for the machine to turn on with a few breaths then that part is working properly
When set to on if it won't turn off when you remove the mask after a reasonable length of time then I don't know of anything except that there seems to be a problem with the sensor.
Perhaps someone who does have a S9 can do a little experiment and actually time the amount of time it normally takes for the machine to turn off when the mask is removed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
bosstow
Posts: 67
Joined: Fri May 04, 2012 6:17 pm

Re: S9 Auto Off doesn't seem to be working

Post by bosstow » Thu Sep 19, 2013 2:55 pm

Pugsy wrote:Perhaps someone who does have a S9 can do a little experiment and actually time the amount of time it normally takes for the machine to turn off when the mask is removed.
Experiment? Sure!

OK, just tested on both the S9 Autoset and the S9 VPAP Auto. It took 7 seconds on both machines to stop when the mask was fully removed and the nasal pillows were unobstructed (i.e. massive/full leak)

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Max IPAP 25, Min EPAP 10, PS 5, ramp off. Climateline MAX tubing, CMS-50F Pulse Oximeter, ResScan & SleepyHead
Also have a Resmed S9 Autoset as a backup/travel machine

User avatar
bosstow
Posts: 67
Joined: Fri May 04, 2012 6:17 pm

Re: S9 Auto Off doesn't seem to be working

Post by bosstow » Thu Sep 19, 2013 3:12 pm

Also, this is in the Troubleshooting section of the S9 Information Guide:

Image

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Max IPAP 25, Min EPAP 10, PS 5, ramp off. Climateline MAX tubing, CMS-50F Pulse Oximeter, ResScan & SleepyHead
Also have a Resmed S9 Autoset as a backup/travel machine

User avatar
StuUnderPressure
Posts: 1378
Joined: Mon Jun 18, 2012 10:34 am
Location: USA

Re: S9 Auto Off doesn't seem to be working

Post by StuUnderPressure » Thu Sep 19, 2013 3:16 pm

bosstow wrote: Experiment? Sure!

OK, just tested on both the S9 Autoset and the S9 VPAP Auto. It took 7 seconds on both machines to stop when the mask was fully removed and the nasal pillows were unobstructed (i.e. massive/full leak)
My S9 AutoSet also takes only seconds to automatically turn itself off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02

SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html

User avatar
JeffL
Posts: 507
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: S9 Auto Off doesn't seem to be working

Post by JeffL » Thu Sep 19, 2013 5:20 pm

I don't see anything that applies.
I have the S9 Autoset, set for Auto with a pressure of 5 to 12.
The SmartStart is On.

I just happened to get a call from my supplier about an hour ago, and I asked about it. For what it's worth I was told that the newer versions don't turn off automatically anymore. I guess I could verify that with ResMed.

_________________
MachineMask

User avatar
bosstow
Posts: 67
Joined: Fri May 04, 2012 6:17 pm

Re: S9 Auto Off doesn't seem to be working

Post by bosstow » Thu Sep 19, 2013 6:24 pm

JeffL wrote:I don't see anything that applies.
I have the S9 Autoset, set for Auto with a pressure of 5 to 12.
The SmartStart is On.

I just happened to get a call from my supplier about an hour ago, and I asked about it. For what it's worth I was told that the newer versions don't turn off automatically anymore. I guess I could verify that with ResMed.
I'm fairly certain that is NOT true what they told you regarding newer versions not turning off automatically when SmartStart is enabled.

As a test, you could increase your minimum pressure to 7 or 8 (temporarily only for this test), then start it up and take your mask off once it comes up to pressure and see if it then is able to auto shutoff. Then put your min pressure back to 5 after the test.

I'm wondering if when you happen take off the mask with a low "current pressure" (especially since your profile says you are using pillows and the chart above states that pressure < 7 can cause auto off not to function) that the auto off won't work. Your minimum pressure is below 7.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Max IPAP 25, Min EPAP 10, PS 5, ramp off. Climateline MAX tubing, CMS-50F Pulse Oximeter, ResScan & SleepyHead
Also have a Resmed S9 Autoset as a backup/travel machine

User avatar
Lazer1234
Posts: 301
Joined: Sat Aug 17, 2013 5:08 am
Location: Sweden

Re: S9 Auto Off doesn't seem to be working

Post by Lazer1234 » Fri Sep 20, 2013 4:02 am

With 7 cm and epr 1 does not shut down
With 7.2 cm and epr 1 does not shut down
With 7.2 cm and epr off it will shut down
I have pillows and slim line

Edit: The pressure may vary depending on the model. The S9 Autoset cs-a is 6 cm in force.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: APAP 7,8 - 20 EPR 3 No Ramp ResScan 5.1 SleepyHead 0.9.8.1
Everything I write I translate through Google Translate.
Hope you have patience with that, sometimes it can get a little crazy.
/Lazer1234

Gparr
Posts: 125
Joined: Thu May 23, 2013 4:38 pm

Re: S9 Auto Off doesn't seem to be working

Post by Gparr » Fri Sep 20, 2013 12:18 pm

I noticed with my S9 since I had to bump up my pressure to 8 the auto on works fine...however the auto off does not work.

_________________
MachineMask
Murphy's New Law

If it doesn't fit, force it...if it breaks, it was meant too!

User avatar
Lazer1234
Posts: 301
Joined: Sat Aug 17, 2013 5:08 am
Location: Sweden

Re: S9 Auto Off doesn't seem to be working

Post by Lazer1234 » Fri Sep 20, 2013 1:25 pm

If you set epr to 1 it should work, not 2 or 3.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: APAP 7,8 - 20 EPR 3 No Ramp ResScan 5.1 SleepyHead 0.9.8.1
Everything I write I translate through Google Translate.
Hope you have patience with that, sometimes it can get a little crazy.
/Lazer1234

User avatar
SleepWellCPAP
Posts: 370
Joined: Fri Mar 22, 2013 11:44 am
Location: Naples Florida
Contact:

Re: S9 Auto Off doesn't seem to be working

Post by SleepWellCPAP » Fri Sep 20, 2013 5:31 pm

Not sure if this has been mentioned yet, but if you are using a nasal pillow mask with small size pillows the S9 may or may not turn off automatically.

It's a known, not sure if problem is the right word. The auto off feature, activated ironically by turning smartstart to on, has problems with nasal pillows. Three seconds is the longest it should take to turn off.

Hope that helps, good luck.
Jim Swearingen
Author of the book Sleep Well & Feel Great with CPAP, a definitive guide
For a free copy inquire with your local county librarian
CPAPtalk featured - Also available through Barnes & Noble Booksellers

User avatar
JeffL
Posts: 507
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: S9 Auto Off doesn't seem to be working

Post by JeffL » Fri Sep 20, 2013 8:32 pm

Quit looking! I got to the bottom of this.

First of all. I don't know who the bigger idiots are, my provider, who told me that the auto off doesn't work in the newer models, or the person (who will go unnamed) at ResMed who replied to my inquiry.

I stated in my message to ResMed the SmartStart was enabled, and that my provider told me that auto off doesn't work in the newer units.
This is what I got back from RedMed:

Thank you for contacting ResMed. We recommend that you contact your local homecare provider to confirm that the Smart start/stop feature is enabled on your new unit since that feature is available on all of our S9 AutoSet units (new and older versions).

Based on Bosstow's info and suggestion, I experimented by altering the min. pressure. What I found was it will turn off at a pressure setting of 8 (takes a count of 6 or 7). From 7.2 to 7.8, I can hear the airflow drop, and it becomes less steady, but sometimes it will turn off, and sometimes it won't.

Looking at the pressure graphs in my ResScan charts, it looks as if I float around on either side of 8.

_________________
MachineMask