Where can I download Encore Basic?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
6PtStar
Posts: 2659
Joined: Mon Jun 18, 2007 1:58 pm
Location: Texas, Is there any other place?

Re: Where can I download Encore Basic?

Post by 6PtStar » Mon Sep 23, 2013 3:27 pm

It would be helpful if you listed you equipment to make sure Encore Basic will work on your machine. Check your PM's (private message) for the link.

Jerry

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS

User avatar
tmiker
Posts: 116
Joined: Sun Sep 16, 2012 3:52 am

Re: Where can I download Encore Basic?

Post by tmiker » Tue Sep 24, 2013 3:13 pm

Hello, I am in need of Encore Basic for Windows 8. Can someone send the info? Thanks a bunch.

ablong
Posts: 123
Joined: Sun May 06, 2007 2:34 am
Location: Texas

Re: Where can I download Encore Basic?

Post by ablong » Wed Sep 25, 2013 2:47 pm

I am in the same boat as JSbarb as to the same kind of machine and I used Encore pro 1.6 till my computer failed and I replaced it with a Win 7 64 bit machine. I had copied a link to EP2 some version
which had been working up to the middle Sept . It said It could not see the card so I tried a different card which had data from the past. So no difference. Then I went and got a Back-up reader
which never had been used. and that failed also. I deleated the program and tried to reload it but It fails to load even though it says it has. Gave up
Went to my Daughters Laptop and tried to Load a copy of EP-1.8 on a vista machine and it does not read the card with certainty. It does not like the reader but will work if you switch ports and it reads
the driver again. It is just to flaky to suit me after all of these years of service with minor change. Any advice would be appreciated.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Where can I download Encore Basic?

Post by Pugsy » Wed Sep 25, 2013 10:54 pm

ablong wrote: Any advice would be appreciated.
PM sent.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

vkyk
Posts: 1
Joined: Wed Oct 16, 2013 4:14 am

Re: Where can I download Encore Basic?

Post by vkyk » Wed Oct 16, 2013 4:19 am

Hi,

I also want to try Encore Basic. Can anyone send me the link?

Foggy1
Posts: 88
Joined: Wed Oct 02, 2013 11:19 am

Re: Where can I download Encore Basic?

Post by Foggy1 » Wed Oct 16, 2013 7:43 am

I am also looking for Encore Basic.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Where can I download Encore Basic?

Post by Pugsy » Wed Oct 16, 2013 7:47 am

vkyk wrote:Can anyone send me the link?
Check your private message box
Foggy1 wrote:I am also looking
You don't have PMs activated so no one can send you anything but I see that you are showing EncoreViewer 2.1 in your profile.
If it gives you wave form graphs along with the the other reports then there's nothing in Basic that you don't have in Viewer.
The reports are identical no matter which Encore version a person uses.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Foggy1
Posts: 88
Joined: Wed Oct 02, 2013 11:19 am

Re: Where can I download Encore Basic?

Post by Foggy1 » Wed Oct 16, 2013 10:14 am

Thank you, Pugsy. If Encore viewer gives me the same data, I'll just stick with what I've got. Did turn on the PM feature though. Thanks for pointing that out.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Where can I download Encore Basic?

Post by Pugsy » Wed Oct 16, 2013 10:38 am

Foggy1 wrote: If Encore viewer gives me the same data, I'll just stick with what I've got. Did turn on the PM feature though. Thanks for pointing that out.
I have heard that the new Viewer 2.1 offers wave form graphs (the old Viewer did not) but I don't have access to it to make sure.
If it does...the reports are identical.
You can always use Sleepyhead for wave form anyway. Also wave form is not all that critical if the reports are looking good and you are feeling good. Breath by breath analysis under the microscope is really boring and unless there are major problems it really isn't even needed.
Should you wish to investigate wave form and you don't have it available in Viewer 2.1 just send me a private message.
Wave form is normally a separate report...and usually we have a little blue paperclip looking thing to click on to get it.

Viewer is the patient version of the software. Offers the exact same detailed nightly reports that Pro or Basic offers. The only difference is that with Viewer the ability to change the pressures via the software is absent. Big deal...like we don't know how to do it manually which is easier anyway
Encore Pro and Basic offer the changing of the pressure via software commands so that the DME or doctor can change a setting on the machine without having the machine in front of them. They just put it on the SD card and send the SD card back to the patient.

Viewer 2.1 is fairly new. They fixed the 32 bit vs 64 bit issue and also made it so it works with the 60 series machines (and older machines). It is still Windows only software and won't install on Windows 8 (there is a way to make it work on Win 8 though).
It is NOT compatible with the new 960 ASV model machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wb4whd
Posts: 3
Joined: Sat Oct 19, 2013 7:24 am

Re: Where can I download Encore Basic?

Post by wb4whd » Sat Oct 19, 2013 7:26 am

I have a Respironics SystemOne 50 and would like the Encore Basic please. I have tried the Philips data uploader for sleepmapper.com and it doesn't work at all, keeps telling me my log in is wrong when it's not.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Where can I download Encore Basic?

Post by Pugsy » Sat Oct 19, 2013 7:29 am

Check your PMs

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
drifterdon
Posts: 13
Joined: Tue Oct 15, 2013 1:35 pm
Location: Western Washington

Re: Where can I download Encore Basic?

Post by drifterdon » Sat Oct 19, 2013 11:37 pm

Hi Pugsy,
You sent me the link for the Rescan but it looks like I will be getting one of the Respironics machines instead.
Can I please have the link for the Encore software?
Thank you

_________________
Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead for Mac OSX
Here I go off on a new adventure

User avatar
6PtStar
Posts: 2659
Joined: Mon Jun 18, 2007 1:58 pm
Location: Texas, Is there any other place?

Re: Where can I download Encore Basic?

Post by 6PtStar » Sun Oct 20, 2013 12:00 am

Check your Private Messages.

Jerry

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS

User avatar
drifterdon
Posts: 13
Joined: Tue Oct 15, 2013 1:35 pm
Location: Western Washington

Re: Where can I download Encore Basic?

Post by drifterdon » Sun Oct 20, 2013 9:42 am

Thank you,
I got Encore to load in my Windows7 partition on my Macbook just fine.
Has anyone ever come up with any sort of tutorial for it?

Thank you,
Don

_________________
Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead for Mac OSX
Here I go off on a new adventure

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Where can I download Encore Basic?

Post by Pugsy » Sun Oct 20, 2013 9:55 am

The Help section in Basic does a good job of explaining how to manage the software.
This may help you in regards to a tutorial on what you may see. It's the only official sort of tutorial that I know of
EncoreGuide link this will show you examples of what you will see. For a different machine than yours but the basics are the same.
http://www.healthcare.philips.com/pwc_h ... _Guide.pdf

In my tutorial we talk about SleepyHead and Basic....data is data no matter what software is presented.
viewtopic/t88983/Pugsys-PointersSleepyH ... nding.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.