SleepyHead Software - How do I interpret average total leaks

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mandypandy2246
Posts: 6
Joined: Wed Aug 21, 2013 9:09 pm

SleepyHead Software - How do I interpret average total leaks

Post by mandypandy2246 » Sun Aug 25, 2013 12:06 am

In the SleepyHead software - how do I interpret average total leaks? Is this supposed to be zero?

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: SleepyHead Software - How do I interpret average total leaks

Post by kaiasgram » Sun Aug 25, 2013 12:54 am

mandypandy2246 wrote:In the SleepyHead software - how do I interpret average total leaks? Is this supposed to be zero?
ResMed or PR machine?

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Software - How do I interpret average total leaks

Post by Pugsy » Sun Aug 25, 2013 4:55 am

Probably a Respironics machine because I don't think that with Resmed machines there is a "Total Leak" data point.

Total Leak with Respironics PR S1 machines include the mask's expected vent rate plus any excess leak.
Vent rates vary with masks and pressures used. Won't ever see 0.0 numbers for Total Leak.

I have a section on leak evaluation here in this thread.
viewtopic/t88983/Pugsys-PointersSleepyH ... nding.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

nicmart

Re: SleepyHead Software - How do I interpret average total leaks

Post by nicmart » Thu Nov 28, 2013 9:33 am

After some false starts with conventional masks, whose odors and discomfort can't tolerate, I have used a cloth mask, which is much more tolerable. Sleepyhead reports that my first night leak rate average was just 1. Seems too good to be true.

I'm using a ResMed 29 Autoset with H5i humidifier and ClimateLine tube.

Heavy spacey feeling when awakening as some others report.

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Software - How do I interpret average total leaks

Post by Pugsy » Thu Nov 28, 2013 9:53 am

nicmart wrote:After some false starts with conventional masks, whose odors and discomfort can't tolerate, I have used a cloth mask, which is much more tolerable. Sleepyhead reports that my first night leak rate average was just 1. Seems too good to be true.

I'm using a ResMed S9 Autoset with H5i humidifier and ClimateLine tube.
The S9 machines reports only excess leak which is a different leak number that Respironics machine report.
An average leak of 1.0 L/min with a S9 machine is indeed possible. That's a great leak number. Your number to avoid is 24 L/min.

Regarding the heavy spacey feeling....unknown cause. Might just be you need more sleep or the sleep you got wasn't so good for some reason...might be related to meds....might be related to apnea events.
More information and detective work is needed before we can offer any ideas about it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
nicmart
Posts: 1
Joined: Thu Nov 28, 2013 9:40 am

Re: SleepyHead Software - How do I interpret average total leaks

Post by nicmart » Thu Nov 28, 2013 10:40 am

Pugsy wrote:
nicmart wrote:After some false starts with conventional masks, whose odors and discomfort can't tolerate, I have used a cloth mask, which is much more tolerable. Sleepyhead reports that my first night leak rate average was just 1. Seems too good to be true.

I'm using a ResMed S9 Autoset with H5i humidifier and ClimateLine tube.
The S9 machines reports only excess leak which is a different leak number that Respironics machine report.
An average leak of 1.0 L/min with a S9 machine is indeed possible. That's a great leak number. Your number to avoid is 24 L/min.

Regarding the heavy spacey feeling....unknown cause. Might just be you need more sleep or the sleep you got wasn't so good for some reason...might be related to meds....might be related to apnea events.
More information and detective work is needed before we can offer any ideas about it.
Thank you.

I'm new, and don't know the ropes. Is there a thread where people upload screen shots of their Sleepyhead date for others to comment upon, or is this considered gauche?

Don't know why I can't figure out how to add profile info. I can't even find anything about doing so in the FAQ.

User avatar
StuUnderPressure
Posts: 1378
Joined: Mon Jun 18, 2012 10:34 am
Location: USA

Re: SleepyHead Software - How do I interpret average total leaks

Post by StuUnderPressure » Thu Nov 28, 2013 10:47 am

If you want to get a 2nd opinion (2nd to SleepyHead software that is), install the ResMed ResScan software. It has a pretty leak graph.

An "average" leak of 1.0 would be acceptable to me IF the highest leak was not much more than that.

I do NOT have any problems with leaks, but if I were to have a leak that went up even to just 7 or so, that would wake me up.

My leak number MOST nights is 0.00 & I sleep best when that is the case.

But, I am sure the SleepyHead software is accurate.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02

SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Software - How do I interpret average total leaks

Post by Pugsy » Fri Nov 29, 2013 8:50 am

nicmart wrote:Is there a thread where people upload screen shots of their Sleepyhead date for others to comment upon, or is this considered gauche?
No special thread and in fact it is preferred that you just start a thread of your own for this purpose. That way your stuff doesn't get mixed up with other people's stuff and you get the attention you deserve.

How to upload the image of the reports explained here if you haven't seen it
viewtopic.php?f=1&t=81072&p=737779#p737779

General basic explanation of what you see in SH here
viewtopic/t88983/Pugsys-PointersSleepyH ... nding.html
A lot of the basic questions are covered here.

So start a new thread...that's the way we prefer to look at reports.

_________________