where is my Flow rate Graph

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
meggie
Posts: 6
Joined: Fri Aug 09, 2013 3:37 pm

where is my Flow rate Graph

Post by meggie » Sat Aug 24, 2013 1:59 pm

I have a Respironics machine. My daily graphs in SH are Event Flags, AHI, Pressure, Leak & Snore but not Flow Rate. Do I need it? What will it tell me? How do I get it to show up in my reports? Thank you everyone for helping me figure all this out.

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Additional Comments: I have the earlier model System One DS550HS (A-flex)

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: where is my Flow rate Graph

Post by JDS74 » Sat Aug 24, 2013 2:39 pm

Hi Meggie,

It would help if you showed your equipment list, then those who can help will have the information to help.
I understand that not all Respironics machines record the flow data.
Welcome to the list.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

meggie
Posts: 6
Joined: Fri Aug 09, 2013 3:37 pm

Re: where is my Flow rate Graph

Post by meggie » Sat Aug 24, 2013 2:54 pm

How do I post the equipment list?
thanks

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Additional Comments: I have the earlier model System One DS550HS (A-flex)

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: where is my Flow rate Graph

Post by Pugsy » Sat Aug 24, 2013 3:04 pm

The Flow Rate graph in SleepyHead will go absent if the SD card was not in the PR S1 machine properly all night.
You can go into the SD card and look in the folder that is the P serial number folder for a P0 folder...look for the most recent file that is the file type xxx.005 which is the flow rate graph. If there isn't one...it didn't get written to the SD card. If it's there then maybe there's a problem within the file and SH can't read it..minor corrupted file.
Also make sure that the Flow rate graph didn't accidentally get turned off...go to Preferences/Graphs and make sure there is a check mark by the Flow rate graph.

How to post your equipment?
Explained here
wiki/index.php/Registering_Equipment_in_User_Profile

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

meggie
Posts: 6
Joined: Fri Aug 09, 2013 3:37 pm

Re: where is my Flow rate Graph

Post by meggie » Sat Aug 24, 2013 3:23 pm

I have just downloaded my data for the first time. The instructions that came with the SD card were to insert the card and then remove it as soon as the data was written, which I did. I then immediately loaded the data into SH and did not reinsert the card. Are you saying that I should leave the card in the machine all the time? thanks, you are so helpful. . .

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Additional Comments: I have the earlier model System One DS550HS (A-flex)

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: where is my Flow rate Graph

Post by Pugsy » Sat Aug 24, 2013 3:28 pm

Yes, leave the SD card in the machine all the time...especially at night if you want flow rate graphs.
I think those instructions are likely tied into the Respironics web based data gathering and flow rate graphs are not part of that data reporting.

Flow rate files are extremely large and the PR S1 machine doesn't store them on its internal memory. They are written directly to the SD card while the machine is in use.
Now the other data that is collected by the machine is stored in its internal memory for about a month before it gets written over but the flow rate graph files aren't stored at all.

Leave the SD card in the machine at all times except for when you are importing to your software and you won't lose the flow rate graphs unless there is a minor bug in the file (and that has been known to happen on occasion but not often).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.