Sleepyhead

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bradleyjoe
Posts: 55
Joined: Sun Mar 29, 2009 8:13 am
Location: Nebraska

Sleepyhead

Post by bradleyjoe » Thu Aug 08, 2013 8:25 pm

When I look at time the sleep, wake time does not add up to the hours. It does not do it every day but sum are it said sleep 21:15 wake 06:45 hours 17:49. Any ideas would help Brad

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: pressure 6-20

User avatar
DiverCTHunter
Posts: 484
Joined: Thu Jan 05, 2012 11:48 am
Location: Cleveland, TN

Re: Sleepyhead

Post by DiverCTHunter » Thu Aug 08, 2013 8:34 pm

Aliens!

Seriously though, I'd probably start by double-checking the clock and maybe formatting your card. IIRC, the S9 firmware's only flaw is that it tends to go bonkers if the SD card's even slightly corrupted.
When in doubt, open the case. Remember: If you can't open it, you don't own it!

Prescribed APAP range - 6-10 cm/H2O, titrated at 8.
Current range - 9.0-11.5 cm/H2O - still searching for the magic "zero night" but averaging 2.2 AHI

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead

Post by Pugsy » Thu Aug 08, 2013 8:46 pm

It's a minor bug in the SleepyHead software. We call it the double session bug.
You are probably using an older version of SleepyHead. The most recent version 0.9.3 fixed that issue.

Uninstall your current version of SleepyHead. Don't worry about all your old data as it is stored your MyDocuments folder...called SleepApp and the uninstall won't bother it.
Go here
http://sourceforge.net/projects/sleepyhead/files/
make sure you get the version 0.9.3 files dated May 19, 2013 and install it and your sleep hours should straighten out just fine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

bradleyjoe
Posts: 55
Joined: Sun Mar 29, 2009 8:13 am
Location: Nebraska

Re: Sleepyhead

Post by bradleyjoe » Thu Aug 08, 2013 9:15 pm

Thanks for your replays I will re format and get the newer sleepyhead. Thanks again Brad

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: pressure 6-20

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead

Post by Pugsy » Thu Aug 08, 2013 9:23 pm

What are you going to reformat? I don't think anything needs to be reformatted.
Just uninstall SH...install the newer version and that's pretty much it.
Nothing gets changed on the SD card...the old data will still be available but it will be fixed minus the double session issues.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

bradleyjoe
Posts: 55
Joined: Sun Mar 29, 2009 8:13 am
Location: Nebraska

Re: Sleepyhead

Post by bradleyjoe » Thu Aug 08, 2013 9:27 pm

Thanks Pugsy

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: pressure 6-20

bradleyjoe
Posts: 55
Joined: Sun Mar 29, 2009 8:13 am
Location: Nebraska

Re: Sleepyhead

Post by bradleyjoe » Thu Aug 08, 2013 9:33 pm

That did the trick Pugsy all hour readings look to be right. Thank you so much Brad

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: pressure 6-20