Post by JeffH » Wed Jul 03, 2013 5:40 pm

Patty (my GF) got her new Resmed VPAP Adapt SV today. When I was setting it up I noticed it had an extra setting that my machine (same one) doesn't have. Her's has in mode setting, CPAP, ASV/AUTO, and ASV. Mine only has CPAP and ASV. What is the ASV/AUTO setting?

Thanks

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
JeffH
Posts: 2153
Joined: Mon Dec 04, 2006 3:44 pm

Re: S9 VPAP Adapt question

Post by JeffH » Wed Jul 03, 2013 5:53 pm

Never mind, I found it.

choice of auto-adjusting EPAP in ASVAuto mode or fixed EPAP in ASV mode.

From Resmed website. New feature, at least to me. My machine is a couple of years old (6400 hrs) and only has CPAP and ASV modes.


JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: S9 VPAP Adapt question

Post by DreamDiver » Wed Jul 03, 2013 9:07 pm

JeffH wrote:choice of auto-adjusting EPAP in ASVAuto mode or fixed EPAP in ASV mode.
What mode is she going to use?

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF

User avatar
JeffH
Posts: 2153
Joined: Mon Dec 04, 2006 3:44 pm

Re: S9 VPAP Adapt question

Post by JeffH » Thu Jul 04, 2013 7:25 am

DreamDiver wrote:
JeffH wrote:choice of auto-adjusting EPAP in ASVAuto mode or fixed EPAP in ASV mode.
What mode is she going to use?

ASV at first. Probably going to call her Doc next week and see what he says about ASV/AUTO.


JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
MagsterMile
Posts: 393
Joined: Mon Dec 24, 2012 1:02 pm
Location: Northern Illinois

Re: S9 VPAP Adapt question

Post by MagsterMile » Thu Jul 04, 2013 7:35 am

That is interesting about the auto-adjusting epap. My epap is fixed at 6. I sometimes think I'd like to have it set lower. I wonder what goes into the thinking that causes the epap to be set at the number it is.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 series ResMed VPAP Adapt-pressure: Epap 6.0 ps5 - 13.0

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: S9 VPAP Adapt question

Post by JohnBFisher » Thu Jul 04, 2013 9:40 am

MagsterMile wrote:That is interesting about the auto-adjusting epap. My epap is fixed at 6. I sometimes think I'd like to have it set lower. I wonder what goes into the thinking that causes the epap to be set at the number it is.
Two answers to that one ..

First, it was the original design. Auto Adjusting BiPAP or ASV requires a lot more technology behind the scenes. So, it's more expensive.

Second, not all patients can tolerate an auto adjusting BiPAP or ASV. In fact, some patients actually score worse when the EPAP adjusts. But those are the minority and the auto adjusting on the high end says that most patients using an ASV unit should adjust to an auto adjusting setup.

I suspect this was added to keep the Resmed ASV unit competitive with the Respironics ASV unit.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
JeffH
Posts: 2153
Joined: Mon Dec 04, 2006 3:44 pm

Re: S9 VPAP Adapt question

Post by JeffH » Fri Jul 05, 2013 8:01 am

For John:

At this point in time do you think most of the differences in machines are mostly software rather than mechanical? I'm thinking that they use the same machine but just load different software on it.

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: S9 VPAP Adapt question

Post by avi123 » Fri Jul 05, 2013 7:01 pm

JeffH wrote:For John:

At this point in time do you think most of the differences in machines are mostly software rather than mechanical? I'm thinking that they use the same machine but just load different software on it.

JeffH
By the suggested set-up of the two machines I can tell that they differ in technology:

(view it in unzoomed Window)

Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: S9 VPAP Adapt question

Post by jdm2857 » Fri Jul 05, 2013 7:36 pm

I think JeffH was asking about differences between machines within a brand, not between brands.
jeff

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: S9 VPAP Adapt question

Post by JohnBFisher » Fri Jul 05, 2013 11:47 pm

JeffH, to be honest, I duo not know for certain. I do know that in previous versions the mechanical components were quite different. I suspect that may still be the case. For example, the motor used in an ASV unit must be more robust than the average CPAP motor. Why put that type of motor in EVERY unit when only a fraction of the units sold need that much motor. So, while I do not *know*, I am pretty certain the answer is that it is more than just software.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: S9 VPAP Adapt question

Post by Papit » Sat Jul 06, 2013 12:18 am

JeffH wrote:Never mind, I found it.
choice of auto-adjusting EPAP in ASVAuto mode or fixed EPAP in ASV mode.
From Resmed website. New feature, at least to me. My machine is a couple of years old (6400 hrs) and only has CPAP and ASV modes.
JeffH
Hi Jeff. I wonder if you (or others here) know what would prompt one's doc to prescribe use of the ASV Auto mode rather than the regular ASV mode for a patient on the S9 VPAP Adapt?

Also, I'm curious what your Flow Limitation graph looks like as a user of either or both modes. FL's were always very very low when I used cpap for over a year. So I'm not sure what to make of so many strong FL spikes now appearing on my FL graph. The FL graph typically isn't preselected to show up on ResScan by S9 Adapt machine users. So if you (any Adapt users) don't see it and you'd like to check it out and comment for me, you can call it up as follows:

To view your Flow Limitation graph, open your patient account in ResScan as usual and then select the REPORT menu. Then click the CUSTOMIZE tab. Expand the Detailed Graphs by clicking the “ (+) ” next to it on the right side of the page (under Report Layout). Then do the same on the left side of the page (under Available Items). Now select Flow Limitation and click “Add”. And then click OK.
---- OR ----
To view your Flow Limitation graph in SleepyHead, open SH, click the File menu and select Preferences. Then select the Graphs tab, check the box next to “Flow Limit” in both the upper “Daily Graphs” list and further below in the “Overview Graphs” list. Then click OK and you will have added the Flow Limitation graph to the other graphs you can see.


Let me know how you do by PM or post here.
All the best.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Last edited by Papit on Mon Jul 08, 2013 12:19 am, edited 3 times in total.
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

User avatar
JeffH
Posts: 2153
Joined: Mon Dec 04, 2006 3:44 pm

Re: S9 VPAP Adapt question

Post by JeffH » Sat Jul 06, 2013 8:33 am

jdm2857 wrote:I think JeffH was asking about differences between machines within a brand, not between brands.
correct

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
JeffH
Posts: 2153
Joined: Mon Dec 04, 2006 3:44 pm

Re: S9 VPAP Adapt question

Post by JeffH » Sat Jul 06, 2013 8:36 am

Papit wrote:
JeffH wrote:Never mind, I found it.
choice of auto-adjusting EPAP in ASVAuto mode or fixed EPAP in ASV mode.
From Resmed website. New feature, at least to me. My machine is a couple of years old (6400 hrs) and only has CPAP and ASV modes.
JeffH
Hi Jeff. I wonder if you (or others here) know what would prompt one's doc to prescribe use of the ASV Auto mode rather than the regular ASV mode for a patient on the S9 VPAP Adapt?

Also, I'm curious what your Flow Limitation graph looks like as a user of either or both modes. FL's were always very very low when I used cpap for over a year. So I'm not sure what to make of so many strong FL spikes now appearing on my FL graph. The FL graph typically isn't preselected to show up on ResScan by S9 Adapt machine users. So if you (any Adapt users) don't see it and you'd like to check it out and comment for me, you can call it up as follows:

To view your Flow Limitation graph, Open ResScan and select any date. Select Tools>Options>Report tab>OK. Click OK in the Select Data window for the selected detailed data session(s). Click the Customize menu tab. Expand (+) Detailed Graphs to see a list of your graphs on the right (Report Layout) and on the left (Available Items). Select Flow Limitation under the Available Items list and click “Add” it. Click OK.

Hope things are getting better and better for you and your GF. All the best.

Papit, you are correct, my flow limitations don't show up. I tried following your directions to get them to but never found the custom menu tab you speak of. Could you maybe clarify your instructions>


Thanks

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: S9 VPAP Adapt question

Post by Papit » Sun Jul 07, 2013 1:19 am

Jeff and any other ASV users (Respironics BiPAP Auto or ResMed S9 VPAP Adapt),

Here are instructions (edited) for adding the Flow Limitation graph to your Detailed Graphs and Reports. I'm looking for users of the ResMed S9 VPAP Adapt and Respironics BiPAP Auto SV to see how common (because of a software artifact) or how patient-specific (therefore varying) are the occurrence of frequent and strong graphed FL spikes. My FL graph during the 18 months I used cpap and apap showed relatively few, very weak flow limitation 'events'. On my new ASV, S9 Vpap Adapt, the graph shows tons of strong and medium ones. Are you seeing the same on your S9 Adapt or BiPAP Auto SV machine? How about when you were using a cpap or apap?

To view your Flow Limitation graph, open your patient account in ResScan as usual and then select the REPORT menu. Then click the CUSTOMIZE tab. Expand the Detailed Graphs by clicking the “ (+) ” next to it on the right side of the page (under Report Layout). Then do the same on the left side of the page (under Available Items). Now select Flow Limitation and click “Add”. And then click OK.
---- OR ----
To view your Flow Limitation graph in SleepyHead, open SH, click the File menu and select Preferences. Then select the Graphs tab, check the box next to “Flow Limit” in both the upper “Daily Graphs” list and further below in the “Overview Graphs” list. Then click OK and you will have added the Flow Limitation graph to the other graphs you can see.

Let me know how you do by PM or post here.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Last edited by Papit on Mon Jul 08, 2013 12:13 am, edited 1 time in total.
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

User avatar
Stormynights
Posts: 2273
Joined: Wed Mar 10, 2010 7:01 pm
Location: Oklahoma

Re: S9 VPAP Adapt question

Post by Stormynights » Sun Jul 07, 2013 9:32 am

I followed your instructions exactly and even went back to double check. I have added the flow limitations to both sleepy head and rescan but neither of the programs are actually showing the graphs.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure EPAP 5.8 IPAP 9.4-21.8 PS 3.6/16 S9 Vpap Adapt ASV