0.0

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: 0.0

Post by -SWS » Thu Jul 04, 2013 8:25 am

Papit wrote:
-SWS wrote:
JohnBFisher wrote: I think the "flow limitation" is an artifact of the algorithm used to identify possible limitation of flow and a precursor to snoring.
John's guess might be spot on. Papit, that hypothesis is fairly easy to test. In Resscan's top window panes set up the FL and detailed flow graphs at 8 hours resolution, one graph on top of each other. Use the mouse to drag the individual graphs either higher or lower. Then, in the bottom window panes, set up the FL and detailed flow graphs in 1 minute resolution---also one graph on top of the other. Now click the top graphs right where severe FL is scored. Then use the left and right arrows at the bottom of the 1-minute graphs to scroll back and forth for detailed viewing.

Are the 1-minute flow curves rounded at the tops, or severely distorted? If the 1-minute flow graphs are nice and rounded, then you are probably looking at reporting or interpretation artifact in the algorithm as John suspects. If, on the other hand, the 1-minute flow graphs are moderately or severely distorted at the tops, then you are probably looking at obstructive flow limitation. Generally, central phenomena will have rounded flow tops while obstructive phenomena will have distorted wave tops (deviating from sinusoid). Mixed phenomena can have both over the course of several breaths. Additionally, mixed and central phenomena will often present with flow-amplitude waxing and waning known as periodic breathing. Cheyne-Stokes is one of many possible periodic breathing types. I also think your periodic breathing does not fit the classic Cheyne-Stokes pattern.

Back to FL analysis. If you observe FL by examining wave tops at 1-minute resolution, AND you now see more severe residual FL than you have on past xPAP treatment platforms, then your residual FL might be iatrogenic vocal cord closures thanks to much larger PS values on ASV. Even "normal" test subjects tend to present vocal cord closures/adduction in response to high PS values . . .

I wouldn't worry about that phenomenon if you are sleeping better and feeling better by day. And if you are experiencing iatrogenic vocal cord adduction in response to high PS values, you may very well see those go away with time and physiologic adaptation. Alternately, you might want to eventually experiment by limiting PS max.
Thank you for the guidance, -SWS. I chose to examine two very elevated graphed 'flow limitations' that occurred at 8:39:25am and 8:40:00am in yesterday's (7/3/13) sleep. As you can see below, as usual, lots of FL spikes were graphed during the sleep. Also pictured are a series of 1-minute duration detailed views of the 1-minute resolution Flow and FL graphs that I scrolled through to examine my breathing flow wave tops before the two strong FLs' occurred, during the FL's, and after the FL's. Please check me on this. (1) It appears that my Flow wave generally is fairly sinusoidal in shape with relatively minimal distortion at the tops in the absence of FL's. (2) Also, the two FL's examined (typical, after looking at many others during this sleep) had minor/no significant affect in causing distortion to the Flow waves when they occurred. Again, thanks for the clear explanation and the time you took on this.

Image

Image

Image

Image

Image

Image
Hi, Papit. At this point I'm favoring John's hypothesis (central phenomena falsely scored as FL).

Of those six 1-minute graphs, Rescan scored FL-positive on the fourth and fifth graphs but not the others. Would you mind changing y-axis scale on that pair of graphs and perhaps add a couple more FL-positive 1-minute graphs? You can change scale in Resscan by selecting Tools=> Options=> Preferences (tab)=> Detailed Graph Ranges (button)=> Flow (graph selection). Then change the "lower" and "upper" values to -45 and +45 respectively.

While you're in there, what Minute Ventilation values does Resscan return during moments scored as FL-positive? Just place the mouse cursor over the Minute Ventilation graph to get a popup box with recent-averaged and current minute ventilation values.

User avatar
Dreamingofsleep
Posts: 229
Joined: Sat Feb 23, 2013 11:16 am

Re: 0.0

Post by Dreamingofsleep » Thu Jul 04, 2013 12:48 pm

Papit!!
How amazing and awesome Congrats and congrats you are obviously doing something or a lot of things right!

Dreaming

User avatar
Papit
Posts: 843
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: 0.0

Post by Papit » Thu Jul 04, 2013 8:54 pm

-SWS wrote:
Papit wrote:
-SWS wrote:
JohnBFisher wrote: I think the "flow limitation" is an artifact of the algorithm used to identify possible limitation of flow and a precursor to snoring.
John's guess might be spot on. Papit, that hypothesis is fairly easy to test. . . .
Are the 1-minute flow curves rounded at the tops, or severely distorted?
Thank you for the guidance, -SWS. I chose to examine two very elevated graphed 'flow limitations' that occurred at 8:39:25am and 8:40:00am in yesterday's (7/3/13) sleep.

-SWS wrote: Hi, Papit. At this point I'm favoring John's hypothesis (central phenomena falsely scored as FL).

Of those six 1-minute graphs, Rescan scored FL-positive on the fourth and fifth graphs but not the others. Would you mind changing y-axis scale on that pair of graphs and perhaps add a couple more FL-positive 1-minute graphs? You can change scale in Resscan by selecting Tools=> Options=> Preferences (tab)=> Detailed Graph Ranges (button)=> Flow (graph selection). Then change the "lower" and "upper" values to -45 and +45 respectively.

While you're in there, what Minute Ventilation values does Resscan return during moments scored as FL-positive? Just place the mouse cursor over the Minute Ventilation graph to get a popup box with recent-averaged and current minute ventilation values.
See the images below. I annotated Minute Ventilation readings on the images of the FL events in the graphs and, for relative perspective, at other times during flow as well. For additional FL examples, I picked a cluster of four (marked A, B, C, D), two very strong, one medium and one low strength. They occurred in my most recent sleep (7/4/13). I also added images of the Flow between the FL events. The cluster occurred between 10:12am and 10:16, well after sleep onset and well before awakening. In addition, for whatever correlation value might exist, Pressure and "High Rate" Pressure graphs are included (although I must admit I have not found a definition of "Pressure (High Rate)". The two pressure graphs differ. Eight images follow. The first two have been updated as you indicated.

Image

Image


==================================================================================================


>> FOUR ADDITIONAL FL EVENTS taken from today's 7/4/13 sleep <<


Image

Image

Image

Image

Image

Image

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: 0.0

Post by -SWS » Thu Jul 04, 2013 11:11 pm

Hi Papit- The good news is that your minute ventilation values don't suggest hypoventilation. Regarding the events scored as FL by Resscan. Upon closer examination I think those are mostly obstructive FL. I would invite others to share their opinion(s) based on the most recent images and Figure 1 of this article as reference:
http://journal.publications.chestnet.or ... 1079416#f1

Figure 1's class-1 wave shape in the above article is a normal inspiratory curve. Class 2 through class 7 are common flow-limitation wave shapes--thus obstructive. Additionally, the FL classes or morphologies can be combined or superimposed in a single breath.

Here's my take for each of your above graphs, Papit:
graph1- flow limited (tail end of expiration before FL also reflects classic obstructive narrowing)
graph2- flow limited
graph3- sinusoidal (although 3rd full cycle might have class 7 flattening)
graph4- perhaps FL or disorganized breathing associated with wake/shift/arousal---but probably not central
graph5- flow limited
graph6- flow limited
graph7- perhaps central phenomena based on waxing/waning (although minor waxing and waning is fairly common in ordinary OSA)
graph8- perhaps FL or disorganized breathing associated with wake/shift/arousal

Again, other opinions are most welcome...

User avatar
Papit
Posts: 843
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: 0.0

Post by Papit » Fri Jul 05, 2013 2:58 am

-SWS wrote:Hi Papit- The good news is that your minute ventilation values don't suggest hypoventilation. Regarding the events scored as FL by Resscan. Upon closer examination I think those are mostly obstructive FL. I would invite others to share their opinion(s) based on the most recent images and Figure 1 of this article as reference:
http://journal.publications.chestnet.or ... 1079416#f1

Figure 1's class-1 wave shape in the above article is a normal inspiratory curve. Class 2 through class 7 are common flow-limitation wave shapes--thus obstructive. Additionally, the FL classes or morphologies can be combined or superimposed in a single breath.

Here's my take for each of your above graphs, Papit:
graph1- flow limited (tail end of expiration before FL also reflects classic obstructive narrowing)
graph2- flow limited
graph3- sinusoidal (although 3rd full cycle might have class 7 flattening)
graph4- perhaps FL or disorganized breathing associated with wake/shift/arousal---but probably not central
graph5- flow limited
graph6- flow limited
graph7- perhaps central phenomena based on waxing/waning (although minor waxing and waning is fairly common in ordinary OSA)
graph8- perhaps FL or disorganized breathing associated with wake/shift/arousal

Again, other opinions are most welcome...
Hi, -Sws. That's a very informative article and study, particularly in giving those of us trying to discern some interpretation or meaning from our zoomed-in 1-minute or less wave forms.

I notice that my graphed FL's are very short in time duration in every example we looked at, about 2 seconds or a half of one breath (an inhalation). Slightly over two complete cycles are taking 10 seconds on the Flow graph. My respiratory rate is 15 breaths/minute or 4 seconds per complete inhale+exhale. Then, (1) Do I assume correctly that each complete wave (one cycle) is the equivalent of one-half of a breath, an inhalation cycle -- that is then graphically followed by a complete sinusoidal exhalation cycle on the Flow graph? And (2) Are 2-second FL events, that the S9 VPAP Adapt is designed to quickly act upon, a cause for concern?

Please note. I'm beginning to wonder about the FL graph's applicability to the S9 VPAP Adapt's various mode settings. I see it as relevant to the machine's CPAP mode. However, I'm thinking now that mathematically the FL graph may or may not be useful in both of the ASV modes; i.e., the standard ASV selection and the ASV Auto mode selection. (3) Can you evaluate that? I don't see any clarifying specifics in ResScan's literature so far. I'm still looking. Any thoughts?

Here is something I found in the company's literature. On Page 73 (see link below) it seems to be telling us that detected flow limitations are used by at least some S9 machines to adjust pressure and treat (inhalation) flow. Then, (4) Might the FL's simply be indicators of precisely when the machine did what it is supposed to do and treat central apneas before they can fully occur; i.e., might the FL graph in effect be a machine performance monitor that flags when and how many centrals would have occurred were it not for the VPAP Adapt? --p.73, http://www.apneaboard.com/ResScan_Inter ... -Guide.pdf

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: 0.0

Post by -SWS » Fri Jul 05, 2013 8:40 am

Papit wrote: I notice that my graphed FL's are very short in time duration in every example we looked at, about 2 seconds or a half of one breath (an inhalation). Slightly over two complete cycles are taking 10 seconds on the Flow graph. My respiratory rate is 15 breaths/minute or 4 seconds per complete inhale+exhale. Then, (1) Do I assume correctly that each complete wave (one cycle) is the equivalent of one-half of a breath, an inhalation cycle -- that is then graphically followed by a complete sinusoidal exhalation cycle on the Flow graph? And (2) Are 2-second FL events, that the S9 VPAP Adapt is designed to quickly act upon, a cause for concern?
Flow limitations sometimes result in sleep arousals (RERAs). Hypothetically, if one's sleep is significantly deteriorated by RERA's, then flow limitations are a cause for concern. My view is that residual flow limitations are not a cause for concern if they do not result in significant numbers of sleep-deteriorating RERA's. Please see my comments below about the S9 ASV's pressure response to FL as an obstructive precursor signal.
Papit wrote: Please note. I'm beginning to wonder about the FL graph's applicability to the S9 VPAP Adapt's various mode settings. I see it as relevant to the machine's CPAP mode. However, I'm thinking now that mathematically the FL graph may or may not be useful in both of the ASV modes; i.e., the standard ASV selection and the ASV Auto mode selection. (3) Can you evaluate that? I don't see any clarifying specifics in ResScan's literature so far. I'm still looking. Any thoughts?
Traditionally, sleep medicine has been concerned with flow limitations based on: a) the flow limitation's potential for RERA type disturbances (not all FLs result in RERA's), and b) the flow limitation's association as precursor for more severe obstructive events such as apnea or hypopnea.

Papit wrote: Here is something I found in the company's literature. On Page 73 (see link below) it seems to be telling us that detected flow limitations are used by at least some S9 machines to adjust pressure and treat (inhalation) flow. Then, (4) Might the FL's simply be indicators of precisely when the machine did what it is supposed to do and treat central apneas before they can fully occur; i.e., might the FL graph in effect be a machine performance monitor that flags when and how many centrals would have occurred were it not for the VPAP Adapt? --p.73, http://www.apneaboard.com/ResScan_Inter ... -Guide.pdf
I'm fairly certain the S9 ASV algorithm does not intentionally interpret FL as a precursor to central events. Theoretically, severe FL can reduce minute and tidal volumes, which in turn can affect the flow-targeting aspect of an ASV algorithm. When placed in auto ASV mode, the most recent S9 ASV model will interpret FL wave shape (not volume) as a precursor signal to more severe obstructive events. The obstructive-addressing aspect of the auto-ASV algorithm will then adjust pressure in an attempt to prevent subsequent obstructive apnea or hypopnea.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: 0.0

Post by avi123 » Fri Jul 05, 2013 2:25 pm

Image
Image

Dr Barry Krakow on Flow Limitations (and UARS+RERA):

https://www.youtube.com/watch?v=_FrxWQBm3iQ

Papit, get rid of it!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: 0.0

Post by -SWS » Fri Jul 05, 2013 11:38 pm

Papit wrote: Here is something I found in the company's literature. On Page 73 (see link below) it seems to be telling us that detected flow limitations are used by at least some S9 machines to adjust pressure and treat (inhalation) flow. Then, (4) Might the FL's simply be indicators of precisely when the machine did what it is supposed to do and treat central apneas before they can fully occur; i.e., might the FL graph in effect be a machine performance monitor that flags when and how many centrals would have occurred were it not for the VPAP Adapt? --p.73, http://www.apneaboard.com/ResScan_Inter ... -Guide.pdf
I don't think residual obstruction (FL) is a benchmark of how well or poorly the central component was addressed.

As a side note the initial S9 ASV model does not offer ASV-auto mode, whereas the most recent model does. ASV-auto mode refers to this obstructive-addressing aspect of the latest S9 ASV algorithm:
Resmed wrote:In ASVAuto mode, the EPAP also responds to flow limitation, snore and obstructive apneas on the next breath, in proportion to the severity of the event.
http://www.resmed.com/us/products/s9_vp ... s&sec=true

That means the initial S9 ASV model will not adjust EPAP in response to FL, snore, or OA. Rather, those must be manually addressed/titrated with a fixed EPAP. Similarly, the latest S9 ASV model will not raise EPAP in response to FL, snore, or OA *if* legacy "ASV mode" is enabled rather than "ASV-auto mode".

Also, I think it's hard to find other doctors in sleep medicine who share Dr. Krakow's views about needing to eliminate all FL in absence of RERA's. Dr. Krakow's FL views are certainly not shared by consensus sleep medicine. Kudos to him for exploring that hypothesis. Many of us are waiting for empiricism and/or peer replication supporting Dr. Krakow's FL views.

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: 0.0

Post by JohnBFisher » Fri Jul 05, 2013 11:53 pm

I tend to agree with -SWS. In general, if you feel fine to brilliant, then don't sweat the FL or hypopnea values. If you still feel tired, explore it. But remember, being tired may be due to other issues than just sleep.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
Papit
Posts: 843
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: 0.0

Post by Papit » Sat Jul 06, 2013 1:03 am

JohnBFisher wrote:I tend to agree with -SWS. In general, if you feel fine to brilliant, then don't sweat the FL or hypopnea values. If you still feel tired, explore it. But remember, being tired may be due to other issues than just sleep.
John, -SWS, Avi and all, thanks so much for your continued input. Do I still feel tired and sometimes sleepy during the day despite my new ASV having almost fully reduced my sustained high centrals, and the few obstructives I had, to near zero? Yes, I do. I notice no significant change in that after nearly a month. And so I will continue to explore this with my doc. Although I'm disappointed that I'm still tired, as you say, that may well be due to other issues. And I am very much comforted to know that I am now probably far less likely to expire in my sleep due to lengthy apneas, especially central apnea in my case. That's huge.

So if anyone's sleep doc floats the notion of upgrading from a cpap machine to ASV after less than stellar treatment effectiveness on cpap for several months, I would say go for it. And don't linger.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Last edited by Papit on Sat Jul 06, 2013 11:39 pm, edited 1 time in total.
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: 0.0

Post by NateS » Sat Jul 06, 2013 9:41 am

Papit wrote: So if anyone's sleep doc floats the notion of upgrading from a cpap machine to ASV after less than stellar treatment effectiveness on cpap for several months, I would say go for it. And don't linger.
I was very lucky to be able to persuade my doctor to let me go straight to ASV. It only happened because I first came to cpaptalk, read and learned to insist upon a copy of my full sleep report(s) before accepting the machine first prescribed for me, learned what to look for when I got my report, followed up on recommended reading and then wrote my sleep doctor a letter quoting from my first and second studies re centrals and politely making my argument to go straight to ASV, which resulted in a very amicable phone conversation with him, a 3rd study to support my/his request for authorization to Medicare.

I was very fortunate, thanks to this group, not to have to mess around with a straight CPAP machine first. I know how unsatisfactory that would have been because, by accident, I spent one night on CPAP mode by accident a few months ago, and what a negative experience and terrible events graph!

Best wishes, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: 0.0

Post by avi123 » Sat Jul 06, 2013 10:09 am

The damaging effects from hi Flow Limitations (FL):


In 2011 while on an S9 Elite (plain CPAP at a pressure of 7 cm) my FL was above average:

Image

As a result, you can see in the following Minute Ventilation (MV) top graph, that the FL caused a drop of more than half in the MV.
The FL just robbed me of breathing air:


Image

Raising the pressure on the S9 Elite to around 13 cm (with EPR = 3) did eliminate the FL.

The above has direct connection to the following by -SWS:

As a side note the initial S9 ASV model does not offer ASV-auto mode, whereas the most recent model does. ASV-auto mode refers to this obstructive-addressing aspect of the latest S9 ASV algorithm:


Resmed wrote:In ASVAuto mode, the EPAP also responds to flow limitation, snore and obstructive apneas on the next breath, in proportion to the severity of the event.
http://www.resmed.com/us/products/s9_vp ... s&sec=true

That means the initial S9 ASV model will not adjust EPAP in response to FL, snore, or OA. Rather, those must be manually addressed/titrated with a fixed EPAP. Similarly, the latest S9 ASV model will not raise EPAP in response to FL, snore, or OA *if* legacy "ASV mode" is enabled rather than "ASV-auto mode".


But since an auto machine is contraindicated for CSAS patients, it means that the manual version of the ASV mode is the one to go by but the EPAP pressure needs to be set and fine tuned to treat (eliminate) most FLs.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Sun Jul 07, 2013 2:59 am, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: 0.0

Post by Sir NoddinOff » Sat Jul 06, 2013 1:05 pm

NateS wrote: I was very lucky to be able to persuade my doctor to let me go straight to ASV. It only happened because I first came to cpaptalk, read and learned to insist upon a copy of my full sleep report(s) before accepting the machine first prescribed for me, learned what to look for when I got my report, followed up on recommended reading and then wrote my sleep doctor a letter quoting from my first and second studies re centrals and politely making my argument to go straight to ASV, which resulted in a very amicable phone conversation with him, a 3rd study to support my/his request for authorization to Medicare.
Best wishes, Nate
My emphasis above: Glad to hear your ASV plan worked out so well for you, Nate. However I'm a little shocked that Medicare went along with this. It's my general understanding that one must have months of documented failure while using conventional APAP, then the same for BiLevel, before even remotely qualifying for an ASV machine. Maybe I missed some trick? And certainly my sleep doctor hasn't enlightened me on this topic.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: 0.0

Post by chunkyfrog » Sat Jul 06, 2013 1:27 pm

My thought: A good doctor should be able to adequately convey the medical necessity of proceeding with the proper treatment,
as opposed to subjecting a patient to an extended period of inadequate treatment, with all the risks associated with that.
Kudos to Nate for persevering until he got the results he needed.
I would certainly hope that Medicare could be found liable if their "rules" result in harm to the patient.
A competent doctor's input is invaluable.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: 0.0

Post by JohnBFisher » Sat Jul 06, 2013 2:18 pm

Papit wrote:... Do I still feel tired and sometimes sleepy during the day despite my new ASV having almost fully reduced my sustained high centrals, and the few obstructives I had, to near zero? Yes, I do. I notice no significant change in that after nearly a month. ...
While I agree you should pursue this further with your doctor, don't forget that getting out of shape (as we often do when we have uncontrolled sleep apnea) also drains our energy. It can take quite some time (months) to reverse that impact. But hang in there. You should be able to feel more rested and in better shape if you continue to work toward it.

P.S. Don't forget that "great" nights don't always happen. "Horrible" nights can also occur.

_________________