Can't exhale with Nasal Pillow. NORMAL????

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Beckaboos
Posts: 2
Joined: Thu Apr 11, 2013 11:31 am

Can't exhale with Nasal Pillow. NORMAL????

Post by Beckaboos » Thu Apr 11, 2013 11:41 am

Hello,
I have been using a CPAP machine for about a month. I have the Philips Respironices REMstar. I started out with the Nasal Mask and have had a hard time adjusting to it. I just switched to a Nasal Pillow. My problem is that with the nasal pillow I feel like I can't exhale. My pressure is set at 9. My question is this......should the pressure be less when using a Nasal Pillow in compared to using a Nasal Mask?

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: Can't exhale with Nasal Pillow. NORMAL????

Post by NotLazyJustTired » Thu Apr 11, 2013 11:51 am

That's not normal. A couple of possibilities: 1) you have the mask too tight. Nasal pillows should be just tight enough to sort of float on your nares without leaking. 2) You have the wrong pillow size. I recently moved from a small to a medium. It made a world of difference.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Can't exhale with Nasal Pillow. NORMAL????

Post by kaiasgram » Thu Apr 11, 2013 11:59 am

What NotLazyJustTired said about pillow size is important. And another important point: Not all nasal pillow masks are created equal. I felt suffocated with the Swift FX nasal pillow mask and almost gave up on NP masks. Then I tried the Aloha and it made all the difference, primarily I think because the air holes in the pillows are larger than on the FX. Same size pillow, but larger openings. You didn't say what NP mask you're using, but keep in mind that you can try other models before giving up on NPs.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Xney
Posts: 842
Joined: Fri Jul 20, 2012 6:53 pm

Re: Can't exhale with Nasal Pillow. NORMAL????

Post by Xney » Thu Apr 11, 2013 12:03 pm

EPR/c-flex can also help quite a bit - try setting it to 2.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP

sleepstar
Posts: 442
Joined: Wed Jan 16, 2013 4:46 am
Location: London, UK

Re: Can't exhale with Nasal Pillow. NORMAL????

Post by sleepstar » Fri Apr 12, 2013 7:24 pm

Actually this can be a very normal problem to begin with.

Nasal pillows are directly into your nose and can feel like a lot more pressure than a nasal mask. With nasal masks the air has more room to move around and it feels like less air.

Are you using a ramp? If you're not, definitely put a ramp on. Start your pressure at 5 or 6. This will feel easier to breathe. If you can put CFlex on 2 or 3 this will help too as it will decrease the pressure when you breathe out so it's easier to exhale.

Over time you'll get so used to the pressure you'll wonder if the machine is on! At the start though you definitely want to take advantage of those comfort settings (CFlex and ramp) to make it easier before you drop off to sleep.

Starting at 9cm h2o on nasal pillows would feel difficult for a lot of people. Starting the machine at a lower pressure should really help.

Best wishes