Old Remstar LX, how can I change the pressure?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
rackalot
Posts: 2
Joined: Sun Apr 07, 2013 10:43 pm

Old Remstar LX, how can I change the pressure?

Post by rackalot » Sun Apr 07, 2013 11:06 pm

I have an old resmed remstar lx that I got back in 2000-2001. The only 2 buttons it has is a ramp button and an on/off button. It also has a small hole on the side. Long story short, I need to change the pressure and was wondering if anyone could insruct me on how to do so. Any help would be greatly appreciated.

User avatar
Pugsy
Posts: 64979
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Old Remstar LX, how can I change the pressure?

Post by Pugsy » Mon Apr 08, 2013 8:08 am

Check your private message box. I sent you a link to the only thing I have. Look in Chapter 4.

Plug unit into wall receptacle

Then simultaneously press the Two User buttons and plug the female end of the AC power cord into the AC outlet in the back of the machine.

Hold the 2 user buttons down (doesn't say how long so be patient) to enter the dealer or clinical mode.

Then press the left user button twice until the pressure display appears.
Press the right user button until "20" appears. Each press of the right user button increments the pressure by 0.5 cm.
Chapter 4.2.4 in the manual that I sent you a link to in the private message..is where this is found.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

rackalot
Posts: 2
Joined: Sun Apr 07, 2013 10:43 pm

Re: Old Remstar LX, how can I change the pressure?

Post by rackalot » Mon Apr 08, 2013 11:18 pm

Thanks for your help! I was able to change my pressure back to where I'm more comfortable.
I recently got an S9 with a humidifier but for some reason I sleep better with my old machine......weird!