UARS, Flow Limit, and Flow Rate graphs

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: UARS, Flow Limit, and Flow Rate graphs

Post by avi123 » Tue Mar 26, 2013 2:49 pm

NotLazy, b/c you're using a Resmed machine you get in your data the degree of your Flow Limitation. The machine does it automatically by analyzing your respiration wave shape:

(It measaures the difference between a normal round wave and your flat top wave and gives it a level index between zero and one and that what you see in the Flow Limitation graph. Zero is good and one is bad)

Image

So if your Flow Limitation is as bad as in this case:


Image

Then I would look for a Sleep Clinic that could do UARS measurements with tools like these:

(as Dr Barry Krakow suggested in my above link)

http://journal.publications.chestnet.or ... sueno=2#f2

But, there is not much that you could do to cure UARS except using CPAP or AutoCPAP and hope that it will work and the side effects will away after several years.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Tue Mar 26, 2013 4:54 pm, edited 1 time in total.

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: UARS, Flow Limit, and Flow Rate graphs

Post by NotLazyJustTired » Tue Mar 26, 2013 4:53 pm

blueh2o wrote:Of course, yesterday after reading your post I immediately went home to SleepyHead and starting studying my flow shapes. Regarding "Analysis of the Flow Shapes", I find it interesting that my shapes somewhat mirror those of the women (I'm OK with my masculinity) with my shapes being primarily composed of Class 6, secondarily of Class 2, and thirdly of Class 7. The authors assume that the women's flow shape has some correlation with postmenopausal levels of progesterone. I would argue and say this subset of women are most likely to be diagnosed with UARS and thus the similarities to my flow shapes.
Having said all that, it also occurs to me that you need to have some experience studying flow shapes to be able to come to any sort of reasonable conclusions and it appears this is a relatively new part of sleep medicine science. So, you can throw everything I just said out the door because I have no idea what I'm talking about.
I hear ya and feel the same way. Problem is, you would really need to look at a very large set of flows to come to any sort of conclusion, and probably need to do that over several nights data to get anything significant. Without software, like the pattern generator this study used, it would be a very tedious task. I sure hope we see more research in this area as I feel that will drive the CPAP manufacturers to develop algorithms specifically designed to eliminate more flow limit based arousals. But like you said, what do I know?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: UARS, Flow Limit, and Flow Rate graphs

Post by Sir NoddinOff » Tue Mar 26, 2013 5:13 pm

This may be heresy on this forum, but sometimes I almost feel sorry for sleep doctors (well, not the lazy, money grubbing ones ). The reason is that all these sleep issues we've touched upon are extremely amorphous and threshold challenging. Almost once a month someone posts a sleep doctor's new theory or paper in publication. And all the acronyms!!! Yikes, brain fatigue. Also, who do you believe? I mean, sample studies with nine people... they'd laugh that out of all the statistics classes I've taken.

When I first started reading this thread, my first thought was: Oh great, more sleep stuff to study and ultimately to worry about. Not to diminish the great contributions of people like SWS et al, none-the-less I'm feeling very small in a big sleepy universe. Frankly, I don't think most of the sleep doctors (those with any integrity anyway) or faithful techs feel much better Just my thoughts after a bad night.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: UARS, Flow Limit, and Flow Rate graphs

Post by avi123 » Tue Mar 26, 2013 5:33 pm

NotLazy, b/c you're using a Resmed machine you can get in your data the degree of your Flow Limitation. The machine does it automatically by analyzing your respiration wave shape:

(It measaures the difference between a normal round wave and your flat top wave and gives it a level index between zero and one and that what you see in the Flow Limitation graph. Zero is good and one is bad)

Image

So if your Flow Limitation is as bad as in this case:


Image

Then I would look for a Sleep Clinic that could do UARS measurements with tools like these:

(as Dr Barry Krakow suggested in my above link)

http://journal.publications.chestnet.or ... sueno=2#f2

But, there is not much that you could do to cure UARS except using CPAP or AutoCPAP and hope that it will work and the side effects will go away after several years.

As a matter of fact, I doubt it if Medicare would pay the expenses if I asked a local sleep clinic to check my FL (including UARS, and RERAS):

http://www.bcbsnc.com/assets/services/p ... ndrome.pdf

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Tue Mar 26, 2013 7:23 pm, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: UARS, Flow Limit, and Flow Rate graphs

Post by NotLazyJustTired » Tue Mar 26, 2013 6:02 pm

Sir NoddinOff wrote:This may be heresy on this forum, but sometimes I almost feel sorry for sleep doctors (well, not the lazy, money grubbing ones ). The reason is that all these sleep issues we've touched upon are extremely amorphous and threshold challenging. Almost once a month someone posts a sleep doctor's new theory or paper in publication. And all the acronyms!!! Yikes, brain fatigue. Also, who do you believe? I mean, sample studies with nine people... they'd laugh that out of all the statistics classes I've taken.

When I first started reading this thread, my first thought was: Oh great, more sleep stuff to study and ultimately to worry about. Not to diminish the great contributions of people like SWS et al, none-the-less I'm feeling very small in a big sleepy universe. Frankly, I don't think most of the sleep doctors (those with any integrity anyway) or faithful techs feel much better Just my thoughts after a bad night.
I think I understand where you are coming from. You know in a perfect world, we the patients wouldn't have to know any of this stuff. We would just go to the doctor and he would say, "You have X. Here's a prescription for Y." and you would be cured. Unfortunately medical science hasn't advanced that far. But looking on the positive side, there is a whole lot more technology for us now than there was for OSA patients 20 years ago! Let's be thankful for that!

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: UARS, Flow Limit, and Flow Rate graphs

Post by -SWS » Wed Mar 27, 2013 9:31 am

49er wrote: I do have a different take on what Krakow is doing. Just because no one else in sleep medicine is doing what he is doing doesn't mean it isn't legitimate.
I agree with that, 49er. I don't think we ever had a different take on that.

However, Krakow's work with higher-than-usual PS is experimental at this point. Last I checked there was no published empiricism, peer review, or replication to give the medical community a sense of patient outcome. I'm not harping on that since good research can take years. Because his PS work is unpublished, we have no sense of how well or poorly his techniques work across a target population. I do know those huge PS values are guaranteed to introduce aerophagia and central apneas to some of the patients.

That said, Krakow's experimental PS protocol is one that I would personally consider trying if I had residual FL and standard treatment protocols could not resolve my EDS.

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: UARS, Flow Limit, and Flow Rate graphs

Post by NotLazyJustTired » Wed Mar 27, 2013 9:41 am

-SWS wrote:
49er wrote: I do have a different take on what Krakow is doing. Just because no one else in sleep medicine is doing what he is doing doesn't mean it isn't legitimate.
I agree with that, 49er. I don't think we ever had a different take on that.

However, Krakow's work with higher-than-usual PS is experimental at this point. Last I checked there was no published empiricism, peer review, or replication to give the medical community a sense of patient outcome. I'm not harping on that since good research can take years. Because his PS work is unpublished, we have no sense of how well or poorly his techniques work across a target population. I do know those huge PS values are guaranteed to introduce aerophagia and central apneas to some of the patients.

That said, Krakow's experimental PS protocol is one that I would personally consider trying if I had residual FL and standard treatment protocols could not resolve my EDS.
That is logical. One of the reasons I started this thread is that I continued to find mounting evidence that "standard treatment protocols" (as you so eloquently put it) seem as though they should work for me. I can definitely visualize a negative correlation between flow limitation and pressure. I am almost tempted to dig into the raw data and see if I can't find an empirical relationship, but I don't think I have that kind of energy.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: UARS, Flow Limit, and Flow Rate graphs

Post by NotLazyJustTired » Thu Mar 28, 2013 6:58 am

Not to belabor the point, but here is a text book Flow Limit based arousal I experienced yesterday evening while napping. The arousal was so intense it woke me up. You can clearly see the flattening of the inspiratory flow shapes leading up to the arousal. Then the huge gasp of air (the one with the large inspiratory amplitude) and then nice rounded curves (which should be expected as I was awake by then). I've also included the Flow Limit graph for illustration. The ResMed algorithm is detecting the flow limitation, but is coding it as very minor. As a result, you can see that not only is pressure not increased, it is actually decreasing just before the arousal. I have probably looked at over 1000 breaths in the last two weeks. I can see repeated patterns like this, but usually there is no large gasp of air, just nice rounded inspiratory curves flowing a period of flow limitation. I suspect, but cannot prove without EEG, that these are RERAs and continue to disturb my sleep architecture. The Responics devices actually detect these scenarios and flag them as potential RERA events.

I post this to further point out that the ResMed algorithm is not aggressively attacking flow limitation. As I've said before, not everyone will experience arousals from flow limitation, so it may just be a case of ResMed coding for the greatest common denominator. The workaround seems to be to increase the minimum pressure in APAP mode. I am getting more comfortable with the therapy and will consider trying this for myself very soon...

Image

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: UARS, Flow Limit, and Flow Rate graphs

Post by Sir NoddinOff » Thu Mar 28, 2013 5:20 pm

NotLazyJustTired wrote:Not to belabor the point, but here is a text book Flow Limit based arousal I experienced yesterday evening while napping.
I'm not seeing what you're seeing with that spike. I'm thinking machine anomaly, a hitch in breathing, a software misread, mask fart or whatever. But hey, I'm not a sleep doctor.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: UARS, Flow Limit, and Flow Rate graphs

Post by NotLazyJustTired » Fri Mar 29, 2013 9:30 am

Hey I'm no doctor either! This was definitely an arousal. It woke me up and I noted the time on the clock. I know I am having arousals per the sleep study, but they are hard to see in the data without EEG to correlate. This one was obvious. The ugly flow rate graph preceding is what has led me to believe that this is the cause. But I will humbly admit, I am speculating. If my speculation leads to better and more timely treatment, I will take it!

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

User avatar
NotLazyJustTired
Posts: 313
Joined: Wed Mar 13, 2013 8:57 am
Location: Gastonia, NC
Contact:

Re: UARS, Flow Limit, and Flow Rate graphs

Post by NotLazyJustTired » Fri Mar 29, 2013 9:39 am

I know one data point does not a trend make, but here goes. I bumped my minimum pressure up from 5 to 7 last night. Here is the graph. I have also include a graph from last week. To my eye flow limitation has been reduced! I also zoomed in and took a look at the flow curves. While I do see some inspiratory curves that are not as round as I like, they are much reduced. I'd estimate less than 50% show the patterns I was seeing at 5cm. I am encouraged. I'll keep this setting for a week or two and see how I feel. Thanks for giving me a place to blab on about this. It is helping me, and hopefully will help others as well.

Last night:

Image

Typical previous night:

Image

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; Pressure 7-15
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"

Sleep Well, Frank

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: UARS, Flow Limit, and Flow Rate graphs

Post by -SWS » Fri Mar 29, 2013 10:08 am

NotLazyJustTired wrote:Hey I'm no doctor either! This was definitely an arousal. It woke me up and I noted the time on the clock. I know I am having arousals per the sleep study, but they are hard to see in the data without EEG to correlate. This one was obvious. The ugly flow rate graph preceding is what has led me to believe that this is the cause. But I will humbly admit, I am speculating. If my speculation leads to better and more timely treatment, I will take it!
I agree that it's difficult to pinpoint stimulus/response sequelae in breaths without EEG. Yours seems like a reasonable guess in light of having noted time of awakening. Arousals often entail brief shifts in EEG frequency if not sustained shifts in sleep stage, often deteriorating sleep architecture----with or without awakenings. And I'm no doctor or medical professional either.

May your experiments in CPAP bring improved flow curves and restorative sleep... Again, good luck!

slowriter
Posts: 383
Joined: Sat Jul 20, 2019 4:37 am

Re: UARS, Flow Limit, and Flow Rate graphs

Post by slowriter » Wed Jul 24, 2019 8:00 am

NotLazyJustTired wrote:
Fri Mar 29, 2013 9:39 am
I know one data point does not a trend make, but here goes. I bumped my minimum pressure up from 5 to 7 last night. Here is the graph. I have also include a graph from last week. To my eye flow limitation has been reduced! I also zoomed in and took a look at the flow curves. While I do see some inspiratory curves that are not as round as I like, they are much reduced. I'd estimate less than 50% show the patterns I was seeing at 5cm. I am encouraged. I'll keep this setting for a week or two and see how I feel.
Guess it's been a few years, but just wondering: whatever happened with this experiment?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: UARS; VAuto Mode, 7-15, PS 5.8

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: UARS, Flow Limit, and Flow Rate graphs

Post by zonker » Thu Jul 25, 2019 3:59 pm

slowriter wrote:
Wed Jul 24, 2019 8:00 am
NotLazyJustTired wrote:
Fri Mar 29, 2013 9:39 am
I know one data point does not a trend make, but here goes. I bumped my minimum pressure up from 5 to 7 last night. Here is the graph. I have also include a graph from last week. To my eye flow limitation has been reduced! I also zoomed in and took a look at the flow curves. While I do see some inspiratory curves that are not as round as I like, they are much reduced. I'd estimate less than 50% show the patterns I was seeing at 5cm. I am encouraged. I'll keep this setting for a week or two and see how I feel.
Guess it's been a few years, but just wondering: whatever happened with this experiment?
you are unlikely to get a reply as he hasn't logged on here since 2013.

it's possible that invoking his name may make him appear, i suppose.

how are you at pentagrams?
:D
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg