CPAP Basics - 1
Hello All! Hope you doing well!
In this post my goal will be to concisely assemble everyone's collective knowledge on the basics of CPAP usage. I'll post a question along with what I believe to be the answer, and invite all comments, feedback, or replys. Next Sunday, I'll come back and post another calling it CPAP Basics - 2, etc. etc. I hope all will find this an easy way to get a good start if they are new, or for the veteran users, maybe they will find something they have not yet discovered. This blog is without a doubt a wealth of information and I'm very interested in how you'll answer these questions.
What should be the number one concern for a new CPAP patient?
My answer: By far and away, initial mask fit and cleanliness of both the cushion/pillows and skin. Your mask needs to be the right one and the right size to keep the air in your circuit and you comfortable for at least 6 hours. Our skin produces oils and acids that cause the mask to shift, in addition to discomfort. Night creams and moisturizers need to be avoided in the area where the mask contacts your skin. Also, a quick fitting tip. If you get leaks from any part other than the air entrainment port, push down on the mask with your hand and see if the leak goes away. If it doesn't you either have a fold in the cushion or the mask just isn't the right one for you. Push the mask away from your face far enough for the cushion to clear your skin and let it settle back. Try not to get emotionally attached to any certain mask, I have had people try a dozen before we found one that worked.
In this post my goal will be to concisely assemble everyone's collective knowledge on the basics of CPAP usage. I'll post a question along with what I believe to be the answer, and invite all comments, feedback, or replys. Next Sunday, I'll come back and post another calling it CPAP Basics - 2, etc. etc. I hope all will find this an easy way to get a good start if they are new, or for the veteran users, maybe they will find something they have not yet discovered. This blog is without a doubt a wealth of information and I'm very interested in how you'll answer these questions.
What should be the number one concern for a new CPAP patient?
My answer: By far and away, initial mask fit and cleanliness of both the cushion/pillows and skin. Your mask needs to be the right one and the right size to keep the air in your circuit and you comfortable for at least 6 hours. Our skin produces oils and acids that cause the mask to shift, in addition to discomfort. Night creams and moisturizers need to be avoided in the area where the mask contacts your skin. Also, a quick fitting tip. If you get leaks from any part other than the air entrainment port, push down on the mask with your hand and see if the leak goes away. If it doesn't you either have a fold in the cushion or the mask just isn't the right one for you. Push the mask away from your face far enough for the cushion to clear your skin and let it settle back. Try not to get emotionally attached to any certain mask, I have had people try a dozen before we found one that worked.
Jim Swearingen
Author of the book Sleep Well & Feel Great with CPAP, a definitive guide
For a free copy inquire with your local county librarian
CPAPtalk featured - Also available through Barnes & Noble Booksellers
Author of the book Sleep Well & Feel Great with CPAP, a definitive guide
For a free copy inquire with your local county librarian
CPAPtalk featured - Also available through Barnes & Noble Booksellers
Re: CPAP Basics - 1
To establish nightly gathering and several times a week use of the PAP flow data to modify lifestyle and deal quickly with any problems that will crop up as the therapy is used.
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: CPAP Basics - 1
I disagree. Cart before the horse. Efficacy data gathered while there is significant leakage (or other variance) is practically meaningless. That's a basic tenet of any process control.Todzo wrote:To establish nightly gathering and several times a week use of the PAP flow data to modify lifestyle and deal quickly with any problems that will crop up as the therapy is used.
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...
Unlike his passengers who died screaming as the car went over the cliff...
Re: CPAP Basics - 1
The flow data would show the leak, the problem would be quickly seen for what it is, a direction for solution would be quickly taken.LinkC wrote:I disagree. Cart before the horse. Efficacy data gathered while there is significant leakage (or other variance) is practically meaningless. That's a basic tenet of any process control.Todzo wrote:To establish nightly gathering and several times a week use of the PAP flow data to modify lifestyle and deal quickly with any problems that will crop up as the therapy is used.
I very strongly disagree with you on this.
Last time I checked only 20% of those medically commended to PAP treatment are able to use it for more than four hours per night at the one year mark. I think your attitude is the largest contributing factor in that horrible truth.
Please change your mind!
Todzo
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
-
quietmorning
- Posts: 1279
- Joined: Wed May 04, 2011 10:39 am
Re: CPAP Basics - 1
I'll answer, then I'll go above and read what everyone else has put so that my answer is true.
Number one concern? Understand Sleep Apnea. Understand YOUR Sleep Apnea - and what the numbers mean. Understand what not following through with treatment means, and understand what the benefits of treatment are. THEN - begin to negotiate your way to getting there.
Number one concern? Understand Sleep Apnea. Understand YOUR Sleep Apnea - and what the numbers mean. Understand what not following through with treatment means, and understand what the benefits of treatment are. THEN - begin to negotiate your way to getting there.
- ChicagoGranny
- Posts: 15516
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: CPAP Basics - 1
If this is about basics, not one out of one hundred will know what this means.the air entrainment port
I agree in the sense that you must understand before the first visit to the DME that you will only accept a full-data machine. Take anything less home and you are stuck.Todzo wrote:To establish nightly gathering and several times a week use of the PAP flow data to modify lifestyle and deal quickly with any problems that will crop up as the therapy is used.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: CPAP Basics - 1
And you need to be able to pick out a mask by trying them on-one after another
-under therapy pressure-(like in a shoe store-but with beds.)
If the mask don't fit the patient will quit.
-under therapy pressure-(like in a shoe store-but with beds.)
If the mask don't fit the patient will quit.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: CPAP Basics - 1
As would the awful hiss from a beginner's mal-fitted mask.Todzo wrote:The flow data would show the leak
Of course leak detection/control is always important, and data is vital when you sleep thru leaks.
But the question was "the number one concern for a new CPAP patient?" You don't need "data" to tell you that your new mask is constantly hissing because doesn't fit right. Fixing that is job one.
I believe you have misinterpreted my "attitude" along with the original question...
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...
Unlike his passengers who died screaming as the car went over the cliff...
- StuUnderPressure
- Posts: 1378
- Joined: Mon Jun 18, 2012 10:34 am
- Location: USA
Re: CPAP Basics - 1
I would start a step before that.SleepWellCPAP wrote:What should be the number one concern for a new CPAP patient?
My answer: By far and away, initial mask fit and cleanliness of both the cushion/pillows and skin.
A new CPAP patient should make sure he is getting the proper machine he needs.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap |
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
Re: CPAP Basics - 1
SleepWell, thanks for starting this thread, I think it will be helpful to newbies and veterans alike. There's always something to learn or review again.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: 14/8.4,PS=4, UMFF, 02@2L, |
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: CPAP Basics - 1
I'm with you. We can always keep trying masks. But unless somebody can afford to buy their own machine, they are usually stuck with the first one they get. It sure is hard to improve therapy without any data.StuUnderPressure wrote:I would start a step before that.SleepWellCPAP wrote:What should be the number one concern for a new CPAP patient?
My answer: By far and away, initial mask fit and cleanliness of both the cushion/pillows and skin.
A new CPAP patient should make sure he is getting the proper machine he needs.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: CPAP Basics - 1
":?: What should be the number one concern for a new CPAP patient? Great Post, SleepWellCPAP
SleepWell I hope you are doing well.
Believe this or not SleepWell, but I was thinking along the same lines as you when it comes to "What’s the Most Important?", but I was thinking more on the lines of "Data".
I certainly went after acquiring CPAP Device Data "Right Off the Bat", to the dismay of my DME, and I got it, but then only to come to realize I didn't have a clue at what I was looking at whatsoever.
“AHI”, what's that?...
“Blue Ball” with an H in it, What's That?....
“Leak Graph”, What is it and How do you Fix it?...
“Basal Level”? What's That.... etc.?????
Not having a Clue may be an understatement on my part; that’s sort of like me telling myself that "My Cat" is not the Boss.
I have enjoyed reading everyone's replies to your post and certainly agree with everything that has been posted.
I also, strongly agree with your assessment as to the Face Mask being at the top of the list. Like you I also think that when one puts their mask on their face for the first time, that's where the "Rubber Meets the Road".
To back that up a little, anyone that is going to stay in this business has to accept the fact that forevermore they are going to be a “Non-Free Sleeper” the rest of their life. In other words there will be no more just running and jumping freely and un-tethered in their beds for a night’s sleep, without first looking like they’re going to Fly a Jet.
With all the above being noted, most of us very quickly learn that the most important connection in, and to our CPAP World is Pugsy.
Jim
SleepWell I hope you are doing well.
Believe this or not SleepWell, but I was thinking along the same lines as you when it comes to "What’s the Most Important?", but I was thinking more on the lines of "Data".
I certainly went after acquiring CPAP Device Data "Right Off the Bat", to the dismay of my DME, and I got it, but then only to come to realize I didn't have a clue at what I was looking at whatsoever.
“AHI”, what's that?...
“Blue Ball” with an H in it, What's That?....
“Leak Graph”, What is it and How do you Fix it?...
“Basal Level”? What's That.... etc.?????
Not having a Clue may be an understatement on my part; that’s sort of like me telling myself that "My Cat" is not the Boss.
I have enjoyed reading everyone's replies to your post and certainly agree with everything that has been posted.
I also, strongly agree with your assessment as to the Face Mask being at the top of the list. Like you I also think that when one puts their mask on their face for the first time, that's where the "Rubber Meets the Road".
To back that up a little, anyone that is going to stay in this business has to accept the fact that forevermore they are going to be a “Non-Free Sleeper” the rest of their life. In other words there will be no more just running and jumping freely and un-tethered in their beds for a night’s sleep, without first looking like they’re going to Fly a Jet.
With all the above being noted, most of us very quickly learn that the most important connection in, and to our CPAP World is Pugsy.
Jim
- NotLazyJustTired
- Posts: 313
- Joined: Wed Mar 13, 2013 8:57 am
- Location: Gastonia, NC
- Contact:
Re: CPAP Basics - 1
The number one concern for a new CPAP patient has nothing to do with equipment. As newly diagnosed, I can tell you the number one concern is more psychological than anything. IMHO, going from sleeping "freely" to having an alien strapped to your head is more a battle of the mind than anything. I spent the entire first week of treatment reading the sticky thread, "Success stories". It helped me more than anything else I have read on this forum. Hearing all those stories and just knowing this can work and what issues I would likely face really helped me deal with it mentally. All newbies should take time to read at least part of that thread. I read all 28 pages because I found it immensely inspirational.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: SleepyHead software; Pressure 7-15 |
"So oftentimes it happens, that we live our lives in chains, and we never even know we have the key."
...from The Eagles, "Already Gone"
Sleep Well, Frank
...from The Eagles, "Already Gone"
Sleep Well, Frank
Re: CPAP Basics - 1
Hi quietmorning!quietmorning wrote:I'll answer, then I'll go above and read what everyone else has put so that my answer is true.
Number one concern? Understand Sleep Apnea. Understand YOUR Sleep Apnea - and what the numbers mean. Understand what not following through with treatment means, and understand what the benefits of treatment are. THEN - begin to negotiate your way to getting there.
I think you have the best answer so far!
Certainly I have learned through the years that my flow data is essential to guide the use and management of xPAP therapy but as NotLazyJustTired points out not knowing what to expect and how to respond to it makes for more stress and stress is a key enemy of effective xPAP therapy.
We really should have "xPAP College" as part of xPAP therapy.
Have a great week!
Todzo
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
- Drowsy Dancer
- Posts: 1271
- Joined: Mon Feb 28, 2011 9:35 am
- Location: here
Re: CPAP Basics - 1
This is a wonderful response.NotLazyJustTired wrote:The number one concern for a new CPAP patient has nothing to do with equipment. As newly diagnosed, I can tell you the number one concern is more psychological than anything. IMHO, going from sleeping "freely" to having an alien strapped to your head is more a battle of the mind than anything. I spent the entire first week of treatment reading the sticky thread, "Success stories". It helped me more than anything else I have read on this forum. Hearing all those stories and just knowing this can work and what issues I would likely face really helped me deal with it mentally. All newbies should take time to read at least part of that thread. I read all 28 pages because I found it immensely inspirational.
I'm having a hard time deciding what the very first, number one, concern of a new xPAP patient should be.
There's no doubt that you need the best mask fit possible.
There's no doubt that a data-capable machine will enable you to take charge of your treatment in a way that no brick ever can.
But there's also no doubt that wrapping your head around what's to come is critical too.
One of my favorite sayings is, "You can only have one first priority," which sounds simplistic until you try to apply it. And I'm having trouble here <weebles>.
Maybe the first concern is, having as accurate a titration as possible? I'd say have the most carefully-done PSG possible but (1) not everyone has a PSG and (2) even now I wouldn't know myself how to ensure that my PSG was well-done.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears |
| Additional Comments: Software: SleepyHead. Pressure: APAP 9.5 min/11 max, A-Flex x2 |