ResScan reports now won't print

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

ResScan reports now won't print

Post by wardmiller » Fri Mar 22, 2013 9:51 am

I've been using ResScan for six months without a serious problem. I down load last night's data to my lap top and to my desktop (for backup) each morning. This week I wanted to print a ResScan Report for my doctor visit today. While viewing Detailed Graphs on my desktop, I clicked on Report, then on the last session, then OK. The message "Creating report" came up as usual, but then the area where I should see the report was blank. I waited several minutes, still blank.

I then went to the laptop and brought up the report as usual. Then I printed it to my network's laser printer without a problem.

What should I do to the desktop so it will print reports as it used to do?

Also, I've noticed recently when displaying ResScan graphs on the laptop the time strip at the top of the graphs is off to the right about 45 minutes. Putting the cursor on data points shows the correct time of the events, and looking at the same data on the desktop, the time strip is correct. It is not earth shattering, but makes it difficult to correlate real time to the graphs and to correlate PulseOx data. It is not related to DST, because the error existed before then. Suggests on how to correct the time strip?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Pugsy
Posts: 65262
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan reports now won't print

Post by Pugsy » Fri Mar 22, 2013 10:14 am

Do both the laptop and the desktop have the same ResScan version on them? Which version?
I don't know what is going on with ResScan time clock on the desktop. Assuming that it's clock is correct 45 minutes is a weird number to be off.
I think I would be probably reinstalling the ResScan on the desktop.
Version 4.2 is the latest (and greatest, cough, cough) you can read about it in UncleBob's signature line.
memberlist.php?mode=viewprofile&u=38643

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: ResScan reports now won't print

Post by wardmiller » Fri Mar 22, 2013 12:07 pm

I'm using ResScan Version 03.16.018 on both computers. I'll get the later one and load it on the desktop and see what happens.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: ResScan reports now won't print

Post by wardmiller » Sat Mar 23, 2013 2:40 pm

Pugsy wrote: I think I would be probably reinstalling the ResScan on the desktop.
Version 4.2 is the latest (and greatest, cough, cough) you can read about it in UncleBob's signature line.
memberlist.php?mode=viewprofile&u=38643
I downloaded and installed the 4.2 version. Installation worked okay. But it still shows a blank screen when trying to print a report. I thought maybe it did not recognize my printer so I looked for a place where ResScan would ask about the printer, but could not find anything like that. Not being able to print a report is a real shortcoming. <sigh>

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Pugsy
Posts: 65262
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan reports now won't print

Post by Pugsy » Sat Mar 23, 2013 3:22 pm

I don't even own a printer so I am no help with ResScan printing of reports.
Can you not generate a report and save it as a pdf document and then print from there?
Maybe someone with more ResScan experience could offer some ideas. I don't really use ResScan all that much.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ResScan reports now won't print

Post by NateS » Sat Mar 23, 2013 6:56 pm

wardmiller wrote:
I downloaded and installed the 4.2 version. Installation worked okay. But it still shows a blank screen when trying to print a report. I thought maybe it did not recognize my printer so I looked for a place where ResScan would ask about the printer, but could not find anything like that. Not being able to print a report is a real shortcoming. <sigh>
Have you opened the Windows Control Panel and checked your Printer setup?

Are you able to print from other programs?

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: ResScan reports now won't print

Post by wardmiller » Sun Mar 24, 2013 8:13 am

NateS wrote:Have you opened the Windows Control Panel and checked your Printer setup?

Are you able to print from other programs?
Yep, nothing has changed in the Printer setup. I can print from other programs okay. Just can't print from ResScan any more. Really has me stumped.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Bobby269
Posts: 803
Joined: Sat Nov 27, 2010 11:12 am
Location: San Antonio Texas

Re: ResScan reports now won't print

Post by Bobby269 » Sun Mar 24, 2013 8:25 am

If nothing else works. Save it as a PDF file and print that.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Set to Apap 14.00/20. EPR 3, Climate line hose, Oscar for Mac 1.1.0. Betta.1

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: ResScan reports now won't print

Post by wardmiller » Sun Mar 24, 2013 12:17 pm

Bobby269 wrote:If nothing else works. Save it as a PDF file and print that.
Bobby, while viewing the Detail data, I click on Report. Then Select Data, OK. I get the small screen saying "Creating the Report" and the bar starts filling from left to right. Then that small screen disappears, leaving a blank where the reports should be. There is no report to save or print. Previously, at that point the report would appear and I could page through it, save it, or print it.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ResScan reports now won't print

Post by NateS » Sun Mar 24, 2013 12:28 pm

wardmiller wrote:
Bobby269 wrote:If nothing else works. Save it as a PDF file and print that.
Bobby, while viewing the Detail data, I click on Report. Then Select Data, OK. I get the small screen saying "Creating the Report" and the bar starts filling from left to right. Then that small screen disappears, leaving a blank where the reports should be. There is no report to save or print. Previously, at that point the report would appear and I could page through it, save it, or print it.
What do you have selected by checkmark in the "Select Data" mini-window before you click OK?

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: ResScan reports now won't print

Post by wardmiller » Sun Mar 24, 2013 1:17 pm

NateS wrote:
wardmiller wrote:
Bobby269 wrote:If nothing else works. Save it as a PDF file and print that.
Bobby, while viewing the Detail data, I click on Report. Then Select Data, OK. I get the small screen saying "Creating the Report" and the bar starts filling from left to right. Then that small screen disappears, leaving a blank where the reports should be. There is no report to save or print. Previously, at that point the report would appear and I could page through it, save it, or print it.
What do you have selected by checkmark in the "Select Data" mini-window before you click OK?

Nate
There is only one session displayed in the Report mini-window. I've tried highlighting that session, then Create Report. I've tried with All checked and without it checked. And without the session highlighted. Each time it it brings up the small Creating Report window and begins filling in the progress bar. When the bar is about half full, those windows vanish and the window where the report used to appear is then blank. It still works fine on the laptop with version 3.1 and worked on this tabletop, until only the Report function stopped working last week. When I loaded the Version 4.2, the report feature still did not work.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

sheekom

Re: ResScan reports now won't print

Post by sheekom » Sat Nov 21, 2015 11:05 am

The problem of showing blank page is solved a sfollows:

Click on Tools on ResScan Software Menu Bar
Select Options
Select Reports
Use default for Temporary File

and you will see a report genereted that you can print to pdf or your printer