Going through water too fast on my S9

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
HOG_Daddy
Posts: 17
Joined: Sun Aug 29, 2010 1:41 pm
Location: Annapolis, MD

Going through water too fast on my S9

Post by HOG_Daddy » Wed Mar 06, 2013 12:18 pm

My S9 set has started going through water faster than ever before... and I don't know why. I'm now getting less than 7 hours from a full tank. I'm using the Mirage Quattro mask, with the Slimline hose that's supposed to manage the temp and humidity level.

Any ideas on what I should check first?

Thanks,

Steve

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure set to 9
Steve

First, exhaust the possible... then try the weird stuff

User avatar
Pugsy
Posts: 65136
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Going through water too fast on my S9

Post by Pugsy » Wed Mar 06, 2013 12:22 pm

Check your leaks on your software reports.
Excessive leaks will use up the water quickly.

Otherwise..anything change in regards to room temp or humidity in the ambient air...changed locations and now in a desert?

Did you mean the Climateline hose when you said....if so are you only seeing a temperature control adjustment being available?
HOG_Daddy wrote:the Slimline hose that's supposed to manage the temp and humidity level

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Going through water too fast on my S9

Post by chunkyfrog » Wed Mar 06, 2013 12:23 pm

When the humidity in the bedroom drops below 25%, there's a good chance mine will run dry.
If the tank is nice and clean, I might sleep right through it. (and wake up with a dry mouth)
If the tank is overdue for cleaning, the smell may wake me.
This leaves me with an ugly dilemma.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
HOG_Daddy
Posts: 17
Joined: Sun Aug 29, 2010 1:41 pm
Location: Annapolis, MD

Re: Going through water too fast on my S9

Post by HOG_Daddy » Wed Mar 06, 2013 12:29 pm

It's the Climateline... the one with the automatic feedback to auto control, which I believe is 80 degrees, 80% humidity (if memory serves). I'm still in the same location (waterfront in Maryland), so I can't blame a new desert setting.

I also have a room humidifier going, to keep the ambient humidity around/above 40%. I'm not really sure how to interpret the leak info.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure set to 9
Steve

First, exhaust the possible... then try the weird stuff

User avatar
Pugsy
Posts: 65136
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Going through water too fast on my S9

Post by Pugsy » Wed Mar 06, 2013 12:35 pm

When you look at your software reports do you see much time spent above 24 L/min on the graphs?
If so, leaks are excessive.

With a room humidifier going the ambient humidity should not have changed all that much to cause such an increase in water consumption.
If you can't blame leaks and can't blame ambient humidity and it persists...might want to bring it to the attention of your equipment supplier..just in case.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
HOG_Daddy
Posts: 17
Joined: Sun Aug 29, 2010 1:41 pm
Location: Annapolis, MD

Re: Going through water too fast on my S9

Post by HOG_Daddy » Wed Mar 06, 2013 12:42 pm

Thanks, Pugsy... I have my data open now...

A great deal of my time is over the 24 L/min line, but I'm getting a green symbol on mask fit.

It seemed to me that the Quattro had a much more free exhaust set up (more ports), which was the cause.

Do I simply tighten my mask straps a bit?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure set to 9
Steve

First, exhaust the possible... then try the weird stuff

User avatar
Pugsy
Posts: 65136
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Going through water too fast on my S9

Post by Pugsy » Wed Mar 06, 2013 12:51 pm

Large leak has to occur for more than 25% over 24 L/min of the night for Mr Red Frowny face to show up and that's really a lot of excess leak.
I have had night's with a just horrible leak overall...nearly 2 hours spent over 24 L/min and the rest pushing 24 L/min and out of 8 hours..that's not enough to trigger Mr Frowny face.

How's the age on your cushion? Think it might be getting tired and time for a new one?
Consider a mask liner? First reaction is to tighten things but sometimes tighter makes leaks worse...the mask needs that cushion of air to seal optimally and too tight just squishes it down more and it can't "float".

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
HOG_Daddy
Posts: 17
Joined: Sun Aug 29, 2010 1:41 pm
Location: Annapolis, MD

Re: Going through water too fast on my S9

Post by HOG_Daddy » Wed Mar 06, 2013 1:04 pm

Great info! Ummm.... looking back over the past month's detailed data... I have some good nights, and some lousy ones.

My mask is a just over a month old, so that's probably not it. I've been using the Quattro for about the past year. I have ZERO follow up with my provider. They threw this diagnosis and equipment at me, and it was kind of ... "good luck!".

What little I know about all this, I've learned from folks here on this site.

So... I'm not sure what a mask liner is, so I'll have to look that up.

I now recall that I tightened up the top straps recently, to try to stop some leaking at the bridge of my nose. I'll try softening those a bit, and see what happens.

Thanks again!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure set to 9
Steve

First, exhaust the possible... then try the weird stuff

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Going through water too fast on my S9

Post by LSAT » Wed Mar 06, 2013 1:13 pm

Mask Liners at....
Remzzzs.com or Padacheek.com

User avatar
Pugsy
Posts: 65136
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Going through water too fast on my S9

Post by Pugsy » Wed Mar 06, 2013 1:16 pm

Best mask liner that can be washed and used for months and months is from Padacheek
http://www.padacheek.com/PAC_Maskliners.html

You can try a homemade mask liner to get an idea if it would help or not.
Just cut up an old T Shirt.... Outline the mask about an inch or so around the mask and cut a whole in the middle for nose and mouth.
Padacheeks are nice because they can be anchored and won't crawl all over your face.

Disposable liners like Remzzzs are available...might give you an idea how to cut up the T Shirt.
https://www.cpap.com/productpage/remzzz ... iners.html

Don't know what pressure you are using but they are especially helpful at higher pressures.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: Going through water too fast on my S9

Post by RandyJ » Wed Mar 06, 2013 1:21 pm

I've been averaging over 3/4 tank per night (6-8 hours) w/ClimateLIne on Auto, so I imagine it would be easy to run out if you're having lots of leaks.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
HOG_Daddy
Posts: 17
Joined: Sun Aug 29, 2010 1:41 pm
Location: Annapolis, MD

Re: Going through water too fast on my S9

Post by HOG_Daddy » Wed Mar 06, 2013 1:36 pm

My pressure is 9. I just reprogrammed to enable manual adjustment of humidity. I have it set to 4 right now, just as a starting point.

I just adjusted my straps again, loosening the top ones, and it seems the place I get leaks is right under my eyes (bridge of the nose), and lower in my cheeks (just under my cheek bones). I'll try one of the liners, and see how that goes.

I see the nose pads, too... wondering if that might take up some of the volume around the bridge of my nose, and let me go a little looser on the straps.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure set to 9
Steve