HELP ! CAN'T EXHALE
Re: HELP ! CAN'T EXHALE
"Cokomo, can you zoom in your flow rate trace during one of these strange AA valve events? I'd like to see if your breathing pattern is unusual, or if the machine is able to track your breath at all."
I may have caused confusion by switching to CPAP. Tonight, I will leave it alone ---APAP 15-20. Will post graphs and try to blow up some events .
I may have caused confusion by switching to CPAP. Tonight, I will leave it alone ---APAP 15-20. Will post graphs and try to blow up some events .
- ChicagoGranny
- Posts: 15326
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: HELP ! CAN'T EXHALE
His leak rates (graph) are fairly consistent and sufficiently high enough. This is an indication that the system is delivering enough air.zoocrewphoto wrote:Any chance your hose has a hole or is being squished somehow? Could something have fallen into the hose and it is clogging it sometimes?
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: HELP ! CAN'T EXHALE
The graphs look like the pressure is staying on, but seeing and hearing that valve open and close sounds like something is causing the air flow to stop at the mask. Even if it is on and off quickly. He's being using this machine and style mask for over a year without problems, so something must have changed.ChicagoGranny wrote:His leak rates (graph) are fairly consistent and sufficiently high enough. This is an indication that the system is delivering enough air.zoocrewphoto wrote:Any chance your hose has a hole or is being squished somehow? Could something have fallen into the hose and it is clogging it sometimes?
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: HELP ! CAN'T EXHALE
If the hose got briefly pinched or kinked, would the PR1's total leak trace drop near zero, or is it an average that wouldn't show a brief interruption?ChicagoGranny wrote:His leak rates (graph) are fairly consistent and sufficiently high enough. This is an indication that the system is delivering enough air.zoocrewphoto wrote:Any chance your hose has a hole or is being squished somehow? Could something have fallen into the hose and it is clogging it sometimes?
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Software: SleepyHead 0.9 beta |
Download Sleepyhead here: https://sourceforge.net/projects/sleepyhead/
Re: HELP ! CAN'T EXHALE
Or it's the sound of the valve "opening."ChicagoGranny wrote:But he does say this,
The "audible" is indeed the valve closing.The asphyxiation flap in both swivels, shuts and is audible when it does and exhalation is greatly reduced .
It's a two position, three port valve. It has ports for the hose, mask, and the room room. There is no "open" and "closed." In one position, if you exhale, air goes back up the hose. In the other position, air goes out the hole and into the room.
When there is no air pressure, it's "open" to the room air and "closed" to the hose.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: HELP ! CAN'T EXHALE
The flap not being held open by the incoming air pressure is what really puzzles me.
Re: HELP ! CAN'T EXHALE
That's odd unless the pressure is going away from the machine. Try disconnecting from the hose and blowing into the mask while holding the mask against your hand like it was on your face. When I do that, I can feel the valve open and close and block the airflow.Cokomo wrote:The flap not being held open by the incoming air pressure is what really puzzles me.
Have you disassembled it at looked carefully? You can go all the way down to taking the silicone flap out. You can see the silicone tab that sticks out of the elbow. That's just in a little clip that snaps off of the elbow with very little effort without disassembling anything else. https://www.cpap.com/productpage/resmed ... -clip.html
disassembly guide..
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: HELP ! CAN'T EXHALE
So it's stuck in the position it would be in if the machine shut off? For safety's sake, those valves are designed so that if they fail, it will likely be in that position. This is beginning to sound like a mask problem after all, unless I have the current definitions of "open" and "closed" confused.Cokomo wrote:The flap not being held open by the incoming air pressure is what really puzzles me.
Edit: I went back and read your first post. Seems it's stuck in the other position, so that even with the machine off it doesn't open? That is both weird and mildly dangerous.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Software: SleepyHead 0.9 beta |
Download Sleepyhead here: https://sourceforge.net/projects/sleepyhead/
- ChicagoGranny
- Posts: 15326
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: HELP ! CAN'T EXHALE
If he hears it opening and closing, then it is not stuck in either position.Edit: I went back and read your first post. Seems it's stuck in the other position, so that even with the machine off it doesn't open? That is both weird and mildly dangerous.
There is no "open" and "closed."
Sure there is, otherwise there is no functionality to the valve.
Is terminology confusing this discussion?Cokomo wrote:The flap not being held open by the incoming air pressure is what really puzzles me.
Do you understand the valve is supposed to be closed when the mask is pressurized?
"Closed" means no air can leave the mask by way of this valve.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- The Choker
- Posts: 485
- Joined: Sat Dec 17, 2011 12:53 pm
Re: HELP ! CAN'T EXHALE
All the doctors I now have very little familiarity with masks. I am not so sure about many of the technicians either.MMPulmoSol wrote:Have you ever checked it with your technician or doctor in charge who recommend that for you? If not its a must to do it now.
Did you join to get free advertising in your signature? Free advertising by mooching off of one of your competitors?
T.C.
Re: HELP ! CAN'T EXHALE
Happy to report no issues last night! Have no idea why the bump in AHI the last few days, but I surely feel better today. No flap open/close problems. Would like to thank all for your help. I've included last night's graphs. Possibly there is an answer hidden, but I am pretty much clueless in reading between the lines. Thanks again. Cokomo












Re: HELP ! CAN'T EXHALE
Sometimes we just don't have the answers....that's when I like to blame it on the aliens. They do like to mess with us sometimes.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: HELP ! CAN'T EXHALE
Huh! Well, don't argue with success!!
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.
Re: HELP ! CAN'T EXHALE
The valve has two positions.ChicagoGranny wrote:There is no "open" and "closed."
Sure there is, otherwise there is no functionality to the valve.
In one position, the air path between the mask and the hose is "closed" and the path between the mask and room air is "open."
In the other position, the path between the mask and the room air is "closed" and the path between the mask and hose is "open."
The valve is always "open" to one side and "closed" to the other.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
- ChicagoGranny
- Posts: 15326
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: HELP ! CAN'T EXHALE
Yes it does. Open and closed.The valve has two positions.
The valve has only one raison d'etre. The valve's sole purpose is to allow the passage of air to and from the room and the patient in the case of failure of the system to provide mechanical air flow. No manufacturer would put a valve in this mask if this wasn't a desired, even required, feature.
So it logically follows the valve is correctly described as 1) being open when it allows air to flow to and from patient to room and 2) being closed when it does not allow air to flow to and from patient to room.
Attempts to define open and closed in any other manner are purely factious.
Back to you.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.