How Do You Know if You Need BiPap?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
need2snooze
Posts: 60
Joined: Fri Apr 27, 2012 8:13 pm

How Do You Know if You Need BiPap?

Post by need2snooze » Tue Feb 12, 2013 9:49 pm

In my quest to find that elusive night of wonderful, rested sleep after over four years of CPAP, I'm wondering if I'm using the right machine - S9 Autoset. My dr. has never suggested it, but I'm curious about BiPap. Is it only for people who need high pressure settings b/c of high AHI numbers? Could BiPap possibly help with the aerophagia and sleeping better through the night?

sol
Posts: 457
Joined: Mon Dec 20, 2010 3:52 am

Re: How Do You Know if You Need BiPap?

Post by sol » Tue Feb 12, 2013 10:12 pm

need2snooze wrote:Is it only for people who need high pressure settings b/c of high AHI numbers? Could BiPap possibly help with the aerophagia and sleeping better through the night?
normally bi-level prescribe to people with OSA who non-responsive or non-tolerant of cpap
nothing to do with AHI numbers but people on high pressure might find breathing out on bi-level pressure support more tolerable than cpap
pressure support can help with aerophagia as can be set higher than cpap EPR 3

jweeks
Posts: 1474
Joined: Tue Feb 10, 2009 8:32 pm
Location: Minneapolis, MN
Contact:

Re: How Do You Know if You Need BiPap?

Post by jweeks » Tue Feb 12, 2013 10:25 pm

need2snooze wrote:...I'm curious about BiPap. Is it only for people who need high pressure settings b/c of high AHI numbers?
Hi,

BiPAP is typically needed by people who have issues with pressure causing central events to happen. They get into a catch-22 where the pressure needed for treating OSA events is higher than the pressure that causes central events. By using 2 different pressure levels, the machine can get past that catch-22 situation.

BiPAP is also used by people who need a high pressure level since using a lower exhale pressure can make it a lot easier to breathe using the machine.

When I first started, I needed BiPAP because I wasn't able to get used to exhaling under the high pressure level that I needed. BiPAP ended up being very comfortable, which solved the problem. Over time, due to weight loss, I now need far lower pressure levels. I could successfully use an APAP machine at the current time, but I am so used to BiPAP that I continue to use it. I already have 2 BiPAP machines, so I continue to use them.

What pressure levels are you on, and what issues are you having? BiPAP might or might or might not help with the air ingestion issue. If you want to give it a try, ask you doctor or DME if you could borrow or rent a BiPAP for a short period of time. It is always hard to tell...it could be a goose chase, or it could be the silver bullet that fixes everything.

-john-

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: How Do You Know if You Need BiPap?

Post by kaiasgram » Wed Feb 13, 2013 3:26 am

I thought I'd share here what happened with my recent switch to bipap: First screenshot is my flow limitation graph on auto CPAP with max pressure at 8.4. This is where my pressure has been set for most of the 8 months I've been on APAP:

Image

I tried to increase my max pressure in order to reduce that flow limitation -- below is a graph from when I increased my max APAP pressure to 9.4. Better, but I got bad aerophagia at this pressure and couldn't stay there.

Image

I started on bipap auto just four nights ago and the next screenshot is my flow limitation graph from last night. My max IPAP is 9.4, same as my max pressure on APAP in the screenshot above, and just 1 cm higher than the pressure in the first screenshot, but what a difference:

Image

The reason for the switch to bipap was not because of my AHI which has always been below 1. By switching to bipap I've been able to significantly decrease that flow limitation (which I suspected was disrupting my sleep and making me feel pretty crappy) without getting aerophagia. On APAP I couldn't stay consistently above 9 without getting bad aerophagia, but no problem on bipap auto with max IPAP 9.4.

As an aside -- while it is taking some getting used to the new machine, I feel like breathing is more comfortable for me on bipap -- it's like the machine seems more sensitive to my own breathing rhythm. Hard to describe.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

stage0
Posts: 169
Joined: Wed Jun 18, 2008 11:22 pm

Re: How Do You Know if You Need BiPap?

Post by stage0 » Wed Feb 13, 2013 4:42 am

In addition to cpap intolerance at higher pressures, some co-morbidities of COPD, Parkinsons's or loss of lung tissue can be helped by using the bi-pap. In some cases, COPD gas exchanges are better (exhaling carbon dioxide). With Parkinson's, the breathing mechanism within the brain is stimulated. With lung tissue loses, air is pushed further in.

Also, some morbidly obese ppl need a bipap to overcome the weight on the chest....ie much higher pressures are needed. I have personally got the pressure up to 27/21 for one patient.

User avatar
49er
Posts: 5624
Joined: Mon Jan 16, 2012 8:18 am

Re: How Do You Know if You Need BiPap?

Post by 49er » Wed Feb 13, 2013 5:16 am

stage0 wrote:In addition to cpap intolerance at higher pressures, some co-morbidities of COPD, Parkinsons's or loss of lung tissue can be helped by using the bi-pap. In some cases, COPD gas exchanges are better (exhaling carbon dioxide). With Parkinson's, the breathing mechanism within the brain is stimulated. With lung tissue loses, air is pushed further in.

Also, some morbidly obese ppl need a bipap to overcome the weight on the chest....ie much higher pressures are needed. I have personally got the pressure up to 27/21 for one patient.
Sorry to get off of the topic but I am floored that anyone could tolerate that high of a pressure, even on bipap. And where the heck could you find a mask that doesn't leak?

I guess you do what you have to do for your health but I still astonished that that could work at that high of a pressure.

49er

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: How Do You Know if You Need BiPap?

Post by Sheriff Buford » Wed Feb 13, 2013 6:15 am

To tag on to 49er's comment: Resmed offers the EPR mode. Isn't this what a Bipap machine does? What's the difference? Sheriff

User avatar
Pugsy
Posts: 65387
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How Do You Know if You Need BiPap?

Post by Pugsy » Wed Feb 13, 2013 6:37 am

Sheriff Buford wrote:To tag on to 49er's comment: Resmed offers the EPR mode. Isn't this what a Bipap machine does? What's the difference? Sheriff
EPR is slightly different than the bilevel pressures of a bilevel machine. It's close but the timing is a bit different.
EPR only goes to 3 cm difference. Bilevel machines allow for a much larger difference than 3 cm between inhale and exhale.
So while similar in function by varying the inhale (IPAP) and exhale (EPAP) pressures it is a bit different. Not exactly the same thing. Though I often tell S9 users to try EPR at 3 to get a feel for what Bilvel might feel like. It's close.
I have tried my VPAP in CPAP mode with EPR of 3 cm (only place EPR is available on ResMed's bilevel machine...called VPAP) and the VPAP with Pressure support of 3 and while it feels very similar the timing is a bit different. To me the VPAP with pressure support feels like it waits better for me to breathe. A timing thing. Not a huge difference but a difference. It's really hard to describe.

Most people using bilevel machine will need/want at least 4 cm difference and in some situations much higher differences. EPR only goes to 3 cm. Also, Respironics machines with the various Flex exhale relief. Totally different animals. The amount of reduction is based upon flow so a setting of 3 is extremely unlikely to produce a 3 cm reduction. The reductions are much less and the relief is afforded by the breathing rhythm and air flow based timing of the slight reduction that is offered.

Need2snooze...
See this link for my bilevel story. Never had a problem exhaling against my pressures, never had much of an aerophagia issue yet I do sleep better on a bilevel machine. Sleeping better equates into feeling better for me. Skip over the heated hose stuff.
viewtopic/t71442/viewtopic.php?f=1&t=71 ... te#p656179
Will bilevel be better for everyone? Of course not. It's just like everything else related to cpap ...YMMV.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Denial Dave
Posts: 1334
Joined: Wed Sep 19, 2012 6:45 am
Location: Connecticut

Re: How Do You Know if You Need BiPap?

Post by Denial Dave » Wed Feb 13, 2013 8:54 am

COPD and other health issues aside

my understanding from an article from the Journal of Clinical Medicine / Sleep Study.

During the sleep study / titration, the max level of CPAP is 15 cm H20. Unless the patient can has difficulties tolerating pressures above 10 cm H20

after that, BiPap is to be prescribed.

As Pugsy indicated, the minumum differential between IPAP and EPAP is 4....

If more than 2 obstructions are observed during a minimum set time period, pressures are slowly increased by 1cm.

The goal is no more than one obstruction per hour or if possible the complete elimination of them.

Dave

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started at VPAP pressure setting of 20/14.4. I've survived Stage 1 cancer and lost 80+ lbs. Pressure is now 14.5 / 11
if you don't know where you are going... any road will take you there.... George Harrison

User avatar
ChicagoGranny
Posts: 15528
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: How Do You Know if You Need BiPap?

Post by ChicagoGranny » Wed Feb 13, 2013 9:14 am

need2snooze wrote:In my quest to find that elusive night of wonderful, rested sleep after over four years of CPAP, I'm wondering if I'm using the right machine - S9 Autoset. My dr. has never suggested it, but I'm curious about BiPap. Is it only for people who need high pressure settings b/c of high AHI numbers? Could BiPap possibly help with the aerophagia and sleeping better through the night?
My doctor says BiPAP is indicated for patients who require pressure of 14 cm or higher. You will find many sleep doctors following this general guideline.

BTW, high pressure settings and high AHI numbers are not well correlated. You can have low AHI and require a high pressure or you could have AHI and only need a low pressure.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

need2snooze
Posts: 60
Joined: Fri Apr 27, 2012 8:13 pm

Re: How Do You Know if You Need BiPap?

Post by need2snooze » Wed Feb 13, 2013 1:35 pm

Thanks, everyone! Pugsy, I read your link. Very informative. I'll need to read it again when I've had more than two hours of sleep (rough night last night - don't know why). Per your suggestion, and that of others, I've been trying different setting changes over the past several months. I reset the EPR setting from 1 to 3. That seemed to help some. In looking over my Sleepyhead data over the past several months, when I had the EPR setting on 1, my Flow Limit number averaged at a higher number, 14, 18, 19. When I changed the EPR to 3, that Flow Limit number has steadily been below 10 (at the 95% level). Also, when the EPR was at 1, the Respiratory Rate number was higher, around 20; now it is 16-18. I noticed on your link that you talk about how easy it is to breathe. I just went and put my mask on to see how I really feel when I'm breathing through it, as I'm usually fumbling around getting ready to go to sleep and not thinking that much about how my breathing feels. What I noticed is that I feel more resistance trying to breathe in, as opposed to exhaling. I like the Mirage FX for Her mask, but I do feel "stuffy" with it, like I'm not getting as much air as I would like.

Here is a typical night for me, except that it appears I got several hours of sleep. Although the numbers may look decent, I always feel tired the next day.


Image

Image

jweeks
Posts: 1474
Joined: Tue Feb 10, 2009 8:32 pm
Location: Minneapolis, MN
Contact:

Re: How Do You Know if You Need BiPap?

Post by jweeks » Wed Feb 13, 2013 1:53 pm

Hi,

It is interesting that you are recording almost no events, yet are not sleeping well. I also see that your machine is not adjusting upwards except when you have leaks, so your machine is not seeing any issues other than the two events that were recorded. Does your sleep doctor have any ideas?

I went through that much of last year where RERA breathing (USARS) was disrupting my sleep. I would show no events, but still not sleep well. The solution was a higher fixed pressure. Due to body changes, that issue has vastly diminished. An indicator of this issue would be if you had a sleep study, and they computed an RDI index, and that RDI index was much higher than your AHI.

There are other sleep disorders out there that show up as poor sleep but few events. I wonder if it would be good to press your doctor (or a new doctor) for a follow-up sleep study to diagnose what exactly is messing up your sleep when you are using the CPAP machine.

-john-

need2snooze
Posts: 60
Joined: Fri Apr 27, 2012 8:13 pm

Re: How Do You Know if You Need BiPap?

Post by need2snooze » Thu Feb 14, 2013 10:33 pm

That's interesting about the RDI index. I'll have to see if it was done when I did my sleep study. Haven't heard of it before. My doctor is not willing to order me another sleep study b/c my AHI numbers are so good. I regularly wake up after having slept four hours, regardless how tired I was when I went to sleep. I may or may not be able to go right back to sleep. Sometimes, I wake up again after another hour or two. What I want to figure out is if I am waking up because of something insomnia related, or if a breathing event is causing me to wake up, although not necessarily an OSA event. I don't know much about the Sleepyhead data to know really what I'm looking at, though. Would that show up on Flow Limitation, for instance, or somewhere else? If someone can tell me that, and I can establish a pattern with my data, then I can show my dr. and hopefully get another sleep study, or maybe a different machine will do the trick.

I should have posted this data earlier, as this is a more typical night's sleep for me, although some nights I may have only an hypopnea, or an obstructive, or the variety plate of three different kinds of events. Even so, my AHI rarely goes over one. Again, I wake up after about 4 hours of sleep and may or may not be able to go back to sleep quickly. This night I did go back to sleep easily. Thanks for taking a look and any suggestions you can give me.

Image
Image

User avatar
Pugsy
Posts: 65387
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How Do You Know if You Need BiPap?

Post by Pugsy » Fri Feb 15, 2013 7:56 am

need2snooze wrote:What I want to figure out is if I am waking up because of something insomnia related, or if a breathing event is causing me to wake up, although not necessarily an OSA event.
You can try creating some sort of break in the therapy line when you wake up so you can look for something maybe happening just prior to the awakening. If you wake up then reach over and turn the machine off and then back on again. That way you have a break in the therapy line. Maybe be able to zoom in and see some sort of pattern.
Though you know it might just be sleep maintenance insomnia that is totally unrelated to sleep apnea.

Do you take any meds that might mess with sleep architecture?
Any pain issues?

Your leak line is so good I am green with envy. I don't know enough about what the Flow limitation graphs on the ResMed machine might mean in terms of FLs being present in such a fashion as to cause a wake up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.