New to CPAP and loosing the battle already

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
stevetaylor1234
Posts: 6
Joined: Fri Feb 01, 2013 1:40 pm
Location: Scotland/Germany

New to CPAP and loosing the battle already

Post by stevetaylor1234 » Fri Feb 01, 2013 4:28 pm

Hello, I went to see my doctor in relation to a snoring problem, who subsequently sent me for a sleep study, without telling me why.

I went for the study at the start of this week and was diagnosed with severe sleep apnia, I am 26 years old and quite healthy, the sleep lab stated that for someone in my category it was very rare. I was discharged from hospital a few nights ago with a new ResMed S9 elite with dehumidifier and the heated hose.

The first night after approximately 10 mins of me breathing in the pipe makes a popping noise with the water, which I persevered with for about 4 hours before turning the cpap off. Tonight as I am lay in bed the exact same thing is happening.

Does anyone have any idea how to stop this, I am not sure if I should turn the climate control up or down to eliminate the problem, I need help as this is drivivng me insane already all I want is a good nights sleep.

Thanks very much in advance

Steve

cosmo
Posts: 1371
Joined: Fri Nov 02, 2012 1:15 pm

Re: New to CPAP and loosing the battle already

Post by cosmo » Fri Feb 01, 2013 4:46 pm

No idea as to the humidifier problem, but troubleshoot to see if everything is connected properly. Turn humidifier/climateline settings to low and you can adjust from there.

Just be lucky that your sleep disorder was caught early on. Having severe sleep apnea is no joke.

Welcome to the hosehead clan

User avatar
LSAT
Posts: 13349
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: New to CPAP and loosing the battle already

Post by LSAT » Fri Feb 01, 2013 5:17 pm

I'm guessing, but it sounds like water in the hose. Be sure that you have your machine lower than the top of your mattress so any water accumulating in the hose runs back down to the humidifier. When you get the noise ,raise the hose to see if water runs out.

herefishy
Posts: 1012
Joined: Sun Oct 30, 2005 8:24 am

Re: New to CPAP and loosing the battle already

Post by herefishy » Fri Feb 01, 2013 5:20 pm

Turn the humidifier off for a night and see if that is the problem. Then put the humidifier lower than your head, then insulate your hose, then modify your heat.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: pressure 10-12

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: New to CPAP and loosing the battle already

Post by DreamStalker » Fri Feb 01, 2013 5:24 pm

I don't see how anyone could get enough rainout (condensation) in 10 minutes to get that kind of problem.

My "guess" is that the reservoir has been over filled.

Check the humidifier reservoir to see if it has been filled passed the full level.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New to CPAP and loosing the battle already

Post by Pugsy » Fri Feb 01, 2013 6:55 pm

DreamStalker wrote:My "guess" is that the reservoir has been over filled.
This would be my first suspect also.

To OP....when the reservoir is too full the air blowing over the water pushes water into the hose. Makes an awful noise and it can and will blow the water up the hose clear up to your face. I have done it. PITA because not only do you have to remove the reservoir and pour off some of the water...you also have to take the hose off and drain the water out of it and even out of the mask if it made it up that far and then you have to dry everything because even a few drops of water will make an awful racket.
When I did it the noise woke my husband and he was wondering if I "broke" something.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

stevetaylor1234
Posts: 6
Joined: Fri Feb 01, 2013 1:40 pm
Location: Scotland/Germany

Re: New to CPAP and loosing the battle already

Post by stevetaylor1234 » Sat Feb 02, 2013 5:29 am

Thanks very much for the quick replies, I managed 6.8 hours sleep with the mask on last night.

The problem is definitely water in the hose, as when I took the hose off the face mask water sprayed out.

The machine is raised above my head, and the reservoir was filled right to the maximum line, neither of these I will not do in the future.

I am slowly getting used to the machine, thank you very much.

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: New to CPAP and loosing the battle already

Post by Sheriff Buford » Sat Feb 02, 2013 6:47 am

You can use a regular or climateline hose for your machine. The climateline hose is a heated hose and will prevent the water from condensing as the water vapors travel thru the hose and into your mask. Try and get one. You can also try to sleep without the humidifier. If you don't need it, it can eliminate a lot of hassle. Glad to see you slept so well! You may feel bad that you have to have this therapy at such a young age, but sleep apnea will kill you by taking years off your life. You are blessed that you caught it early enough to eliminate long term damage to occur.

Sheriff

stevetaylor1234
Posts: 6
Joined: Fri Feb 01, 2013 1:40 pm
Location: Scotland/Germany

Re: New to CPAP and loosing the battle already

Post by stevetaylor1234 » Sat Feb 02, 2013 7:35 am

Sheriff Buford wrote:You can use a regular or climateline hose for your machine. The climateline hose is a heated hose and will prevent the water from condensing as the water vapors travel thru the hose and into your mask. Try and get one. You can also try to sleep without the humidifier. If you don't need it, it can eliminate a lot of hassle. Glad to see you slept so well! You may feel bad that you have to have this therapy at such a young age, but sleep apnea will kill you by taking years off your life. You are blessed that you caught it early enough to eliminate long term damage to occur.

Sheriff
Thanks for the reply. I do have the climateline hose which I have been using, I have to use the humidifier as I cannot breath through my nose at all and am a mouth breather, problem being I sleep with my mouth open. there was still some water getting in the hose, I have been to purchase a thermometer today so I can monitor my room temperature, I think I had the heat in the hose too low. (room temperature 23 degrees, Hose: 19 Degrees), and also the machine was above my head, so I have moved onto the floor until I can get a suitable solution for it to sit on a bedside table.

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: New to CPAP and loosing the battle already

Post by squid13 » Sat Feb 02, 2013 8:57 am

stevetaylor1234 wrote:(room temperature 23 degrees, Hose: 19 Degrees),
I assume that is Celsius.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Sleepy Pilot
Posts: 34
Joined: Thu Jan 17, 2013 12:03 pm

Re: New to CPAP and loosing the battle already

Post by Sleepy Pilot » Sat Feb 02, 2013 9:07 am

Steve,

If you don't have a table to rest the machine on, try this platform made to hold a cpap machine on the side of your bed. The board is placed between the box spring and mattress.

https://www.cpap.com/productpage/cpap-b ... older.html

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: New to CPAP and loosing the battle already

Post by Sheriff Buford » Sat Feb 02, 2013 9:31 am

Or you can use a milk crate, box or some phone books. Lower the machine and up the temperature on the humidifier. It's not unusual to have to "tinker" with the humidifier, mask adjustment, pressure, etc.., throughout your whole journey... I still "tinker".

Sheriff

User avatar
JamesW6175
Posts: 82
Joined: Thu Oct 18, 2012 7:14 pm
Location: Austin Texas

Re: New to CPAP and loosing the battle already

Post by JamesW6175 » Sat Feb 02, 2013 9:32 am

Have the same equipment, and had a popping sound one night & just turned the machine off and back on. It stopped the noise, also I never fill my tank above the line on the refill hole on the tank. Use distilled water only and not had any problem with any rain-out( water in the tube). I am a side and stomach sleeper with the CPAP machine on a bedside night stand, which is the same height as the bed. Have the hose up and over the headboard of the bed with no problem changing positions during the night. Use a CPAP pillow which has helped. Originally the machine was set at a temperature of 70 degrees, but that was too warm for me & I adjusted it down to 60 degrees which is much better.

_________________
MaskHumidifier
Additional Comments: Contour Pillow, Climate Control Hose, EPR turned off . Hose management - spring clamp on headboard suspend hose above head & pillow level.

stevetaylor1234
Posts: 6
Joined: Fri Feb 01, 2013 1:40 pm
Location: Scotland/Germany

Re: New to CPAP and loosing the battle already

Post by stevetaylor1234 » Sat Feb 02, 2013 9:35 am

squid13 wrote:
stevetaylor1234 wrote:(room temperature 23 degrees, Hose: 19 Degrees),
I assume that is Celsius.
Yes this is Celsius
Sleepy Pilot wrote:Steve,

If you don't have a table to rest the machine on, try this platform made to hold a cpap machine on the side of your bed. The board is placed between the box spring and mattress.

https://www.cpap.com/productpage/cpap-b ... older.html
I will have a look at one of these, however I am a frequent traveller.

Thanks to everyone for there help and advice, I will tinker again tonight and hopefully sleep right through the night.

User avatar
ughwhatname
Posts: 604
Joined: Sun Oct 14, 2012 8:23 am

Re: New to CPAP and loosing the battle already

Post by ughwhatname » Sat Feb 02, 2013 10:08 am

stevetaylor1234 wrote:
squid13 wrote:
stevetaylor1234 wrote:(room temperature 23 degrees, Hose: 19 Degrees),
I assume that is Celsius.
Yes this is Celsius
Sleepy Pilot wrote:Steve,

If you don't have a table to rest the machine on, try this platform made to hold a cpap machine on the side of your bed. The board is placed between the box spring and mattress.

https://www.cpap.com/productpage/cpap-b ... older.html
I will have a look at one of these, however I am a frequent traveller.

Thanks to everyone for there help and advice, I will tinker again tonight and hopefully sleep right through the night.
I traveled with mine, and the bed was too high to put the machine directly on the floor. I improvised and took the small trash can in the hotel room, turned it upside down, and set it next to the bed for my machine. Worked pretty well.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Mask is the TAP PAP Nasal Pillow CPAP Mask with Stability Mouthpiece