Zeo Headband - Think I've Got It!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Zeo Headband - Think I've Got It!

Post by Madalot » Tue Jan 22, 2013 3:25 pm

I've used my Zeo mobile two nights and noticed after the first night that my forehead hurt and was red and swollen in spots. I loosened it the second night but had more trouble keeping it in place with my FFM. And my forehead is even worse today. My hubby said it looks like somebody hit me and it's pretty sore.

I'd really like to get the recommended baseline.

Any suggestions?

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7
Last edited by Madalot on Tue Feb 05, 2013 6:04 am, edited 3 times in total.

User avatar
Pad A Cheek
Posts: 836
Joined: Sat Oct 22, 2005 3:40 pm
Location: Virginia, In the shadow of the Blue Ridge Mountains
Contact:

Re: Zeo Headband - Ouch

Post by Pad A Cheek » Tue Jan 22, 2013 4:27 pm

Hey Maddy,
I do not have one of those, but I bet if you put anything between your skin and the contacts it would not work as well. Is there any way to make it less tight but still have contact with your head where it needs to?

Sorry my fingers are itching to help make it comfortable and not leave a mark on your beautiful face.
Hugs,

Karen

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: This CPAP stuff at night makes the days so much more productive
And I think to myself...... What a Wonderful World

tetragon
Posts: 539
Joined: Wed Oct 19, 2011 4:06 am
Location: Toronto, ON
Contact:

Re: Zeo Headband - Ouch

Post by tetragon » Tue Jan 22, 2013 4:37 pm

Every time I use my Zeo it leaves a mark that lasts for a while. I keep it just tightly enough to not be easily moved, and put it on first. It works for me without being painful with my Swift FX, but I don't leave my apartment without a toque for at least an hour after I take off my Zeo.

It was painful, though, the times I used it while trying a nasal mask with a forehead brace. When I did that, I put on the sensor first, then the loose mask, and fit it with the brace on top of the sensor; my head is too small for both of them to each have their own patch of forehead.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Software: SleepyHead, loosely following HEAD

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Zeo Headband - Ouch

Post by Jay Aitchsee » Tue Jan 22, 2013 4:39 pm

Too tight the first night? It just needs to be snug. An allergy? Your not allergic to silver, are you? Something else? Zeo recommends washing the sensor pads in woolite, though I never have. Zeo also acknowledges that the sensor pad pads can be cleaned with alcohol. I have, by soaking, swishing, and allowing to air dry. Didn't seem to hurt it.
Maybe you need to let it rest one night before trying again.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

cosmo
Posts: 1371
Joined: Fri Nov 02, 2012 1:15 pm

Re: Zeo Headband - Ouch

Post by cosmo » Tue Jan 22, 2013 4:41 pm

Image

Secure it to you head with ace bandage?

I never had a problem with the zeo. I have it under my headgear and chinstrap.

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Zeo Headband - Ouch

Post by Madalot » Tue Jan 22, 2013 8:54 pm

I'm definitely not using the Zeo tonight and may have to go without the vent as well. I can't even touch my forehead. I don't know if I can handle the forehead support on my mask.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Zeo Headband - Ouch

Post by Madalot » Wed Jan 23, 2013 6:49 am

I WAS able to use my mask last night, so that's a relief.

I'm not sure what the heck happened. I'm guessing the Zeo was too tight that first night, but loosening it the second night made it slide when I tried to put on my mask.

I don't think I'm allergic to silver or anything. And I doubt that I can put anything between my forehead and the contacts.

I fear I'm going to need several days for my forehead to get back to normal. Even this morning it is still just as red, swollen and sore to the touch.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

cosmo
Posts: 1371
Joined: Fri Nov 02, 2012 1:15 pm

Re: Zeo Headband - Ouch

Post by cosmo » Wed Jan 23, 2013 7:08 am

I didn't notice you were using a mask with a forehead strap. How do you use both at the same time?

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Zeo Headband - Ouch

Post by Madalot » Wed Jan 23, 2013 7:18 am

cosmo wrote:I didn't notice you were using a mask with a forehead strap. How do you use both at the same time?
I don't have my mask forehead support very tight and there was plenty of room underneath it for the Zeo. I was concerned about that the first night but was able to get my mask on easily over it. When I loosened the Zeo the second night, the mask made the Zeo slide a bit, but I was still able to do it.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Zeo Headband - Ouch

Post by Jay Aitchsee » Wed Jan 23, 2013 7:32 am

And cosmo, I wear mine just above the forehead support. You need about two finger widths of space to fit the sensor between the forehead piece and the hairline. Another possibility for many masks is to remove the soft support part of the forehead piece and let the mask ride right on top of the transmitter as Madalot has done. The thickness of the soft support and the transmitter/sensor are about the same so it puts the mask at about the same distance horizontally from the face.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Zeo Headband - Ouch

Post by Madalot » Wed Jan 23, 2013 8:41 am

This is what my forehead looks like today, this after skipping one night of the Zeo. I didn't realize until I saw the picture that it has all these white bumps on it. Yuck!!

Image

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Zeo Headband - Ouch

Post by chunkyfrog » Wed Jan 23, 2013 10:27 am

Ouch!
(Cyber-kiss)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Zeo Headband - Ouch

Post by Madalot » Wed Jan 23, 2013 12:31 pm

I am going to try cleaning the Zeo sensors and see if there might be something on them that caused this reaction. But regardless, I am definitely waiting for this to clear up before trying again.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Zeo Headband - Ouch

Post by chunkyfrog » Wed Jan 23, 2013 12:40 pm

When I had a rash like that , but on my back, my doc had me use a cortisone cream (RX)
It worked. Might have to go OTC on the face, and keep it out of your eyes. (call your doc first)
Another idea: someone used alcohol to clean the pads, squeezing out the excess into a cotton cloth.
They claimed no problems, but all the skin oil was out. This might be something to try--once.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: Zeo Headband - Ouch

Post by Uncle_Bob » Wed Jan 23, 2013 4:25 pm

Madalot wrote:This is what my forehead looks like today, this after skipping one night of the Zeo. I didn't realize until I saw the picture that it has all these white bumps on it. Yuck!!
Thats exactly what happened to me. I had three of those marks right where the sensors are on the strap. I put it down to being too tight. Maybe the strap got under my papCap (which i wear at the same time).
Pad A Cheek wrote: Sorry my fingers are itching to help make it comfortable and not leave a mark on your beautiful face.
Hugs,

Karen
My beautiful face would apprciate it if you could come up with something too