Sleepyhead glitch

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Saltydog
Posts: 38
Joined: Fri Jan 20, 2006 4:44 pm

Sleepyhead glitch

Post by Saltydog » Wed Dec 19, 2012 5:47 pm

I have been using the sleepyhead software with the patch for my PR 560 for about 5 nights. Two nights ago there was some sort of malfunction and the data from sleepyhead said I had an AHI of 2316.15. The software also said that I was now using a bi=level mode. The software has been ok since and I would like to eliminate the data from that one night. When I go to advanced purge cpap data the program shuts down and the data is not purged. What can I do?

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: Heated Hose

User avatar
Pugsy
Posts: 65293
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead glitch

Post by Pugsy » Wed Dec 19, 2012 6:01 pm

You can try manually removing the files for the affected dates.
Those files are found in MyDocuments/SleepyHeadDate/Profile/and your machine serial number.

The date modified is the file for the night before.
Like my last nights report will show today's date as modified date.
3 files for each date.
xxx.001
xxx.002
xxx.005 those are the wave form data files.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Sleepyhead glitch

Post by archangle » Wed Dec 19, 2012 7:19 pm

Try creating a new user under SleepyHead and then import the data from your card.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.